Help titrating BIPAP for sensitive sleeper

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Nocibur
Posts: 503
Joined: Sat Jan 25, 2025 6:08 am

Re: Help titrating BIPAP for sensitive sleeper

Post by Nocibur » Thu Dec 18, 2025 3:45 am

Miss Emerita wrote: ↑
Wed Dec 17, 2025 12:39 pm
...the bumps or lumps in the inhalation trace.
That's actually part of the expiratory limb, so IMO probably cardioballistic artifact.

User avatar
Nocibur
Posts: 503
Joined: Sat Jan 25, 2025 6:08 am

Re: Help titrating BIPAP for sensitive sleeper

Post by Nocibur » Thu Dec 18, 2025 4:04 am

semblables wrote: ↑
Wed Dec 17, 2025 4:19 pm
... the data...

When I switched from CPAP to BIPAP I felt 10x better and had deeper more restful sleep.
Well, looking at "the data":
semb.jpg
semb.jpg (85.56 KiB) Viewed 6996 times
there does appear to be a very slight decrease in the presence of naps, so I'd say maybe 0.3x better.

However, what jumps out is the amount of "sleep" (and IDK if xPAP time equates to actual "sleep" time) is a LOT (starting to move into pathological range); the need (or at least presence) of naps despite an abundance of sleep; and a sleep period that some people might consider to be "socially unacceptable" (lunchtime and you're still in the rack?)(but an xPAP clock is not necessarily a "real" clock).

Respiratory events essentially -0- for the period regardless of mode.

Anyway, that's MO of "the data".

User avatar
Nocibur
Posts: 503
Joined: Sat Jan 25, 2025 6:08 am

Re: Help titrating BIPAP for sensitive sleeper

Post by Nocibur » Thu Dec 18, 2025 4:10 am

Nocibur wrote: ↑
Thu Dec 18, 2025 4:04 am

there does appear to be a very slight decrease in the presence of naps
And even that doesn't look like it's going to be sustained.

User avatar
Nocibur
Posts: 503
Joined: Sat Jan 25, 2025 6:08 am

Re: Help titrating BIPAP for sensitive sleeper

Post by Nocibur » Thu Dec 18, 2025 6:01 am

semblables wrote: ↑
Sun Dec 07, 2025 4:32 pm
On CPAP in an in-lab sleep test I was having 11-12 non-movement arousals per sleep cycle, so apparently that's enough for my sleep quality to tank.
Arousals/sleep cycle is not what you'd call a "traditional parameter", but if a sleep cycle is ~90 minutes, then the arousal index would be ~8.0.

Would that cause sleep quality to "tank"?

Sincerely doubt it, but if that were the case, IIWY I'd be looking for the cause of AI 8.0.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Help titrating BIPAP for sensitive sleeper

Post by Miss Emerita » Thu Dec 18, 2025 12:39 pm

Nocibur wrote: ↑
Thu Dec 18, 2025 3:45 am
Miss Emerita wrote: ↑
Wed Dec 17, 2025 12:39 pm
...the bumps or lumps in the inhalation trace.
That's actually part of the expiratory limb, so IMO probably cardioballistic artifact.
I'd love to learn why the "bump" in the first part of inhalation is part of the "expiratory limb." Probably I don't know what "limb" means here. I googled the term but didn't find anything I actually understood. Sigh. Hoping you can help!
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Nocibur
Posts: 503
Joined: Sat Jan 25, 2025 6:08 am

Re: Help titrating BIPAP for sensitive sleeper

Post by Nocibur » Thu Dec 18, 2025 2:29 pm

Miss Emerita wrote: ↑
Thu Dec 18, 2025 12:39 pm

I'd love to learn why the "bump" in the first part of inhalation is part of the "expiratory limb."
It's not the first part of inhalation-- it's the last part of exhalation.

Inspiratory phase is only when there is actual inspiration. Expiratory phase is everything else.
expiratory-flow-pattern-n.jpg
expiratory-flow-pattern-n.jpg (30.05 KiB) Viewed 6953 times

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Help titrating BIPAP for sensitive sleeper

Post by Miss Emerita » Thu Dec 18, 2025 4:57 pm

OK, good; that's what I thought. But when the OP starts inhalation, there isn't smooth acceleration at the outset; instead there's a pointy bump or lump. Maybe it doesn't have any significance, but I thought it was a bit unusual. I've pasted in the screenshot again so you can take a look.
Screenshot 2025-12-17 at 10-36-07 SleepHQ.png
Screenshot 2025-12-17 at 10-36-07 SleepHQ.png (119.09 KiB) Viewed 6939 times
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Nocibur
Posts: 503
Joined: Sat Jan 25, 2025 6:08 am

Re: Help titrating BIPAP for sensitive sleeper

Post by Nocibur » Fri Dec 19, 2025 3:49 am

Miss Emerita wrote: ↑
Thu Dec 18, 2025 4:57 pm
...when the OP starts inhalation, there isn't smooth acceleration at the outset; instead there's a pointy bump or lump.
Oh right! I should have paid attention to your original point:
One is that many of your inhalation curves have a more pointed shape than we usually see.
Anyway that's probably BiPAP Engineering. Would need the Oscar file uploaded to a third party server to show why that is (the one advantage Oscar has over SleepHQ).
semb2.jpg
semb2.jpg (174.38 KiB) Viewed 6901 times

semblables
Posts: 9
Joined: Sun Dec 07, 2025 4:29 pm

Re: Help titrating BIPAP for sensitive sleeper

Post by semblables » Sun Dec 21, 2025 3:20 pm

Hi! I've got a bit more data for the higher EPAP of 7.6:

https://sleephq.com/public/08276559-6b5 ... da8ae5864e

Any interpretation / settings recommendations would be welcome!

User avatar
ejbpesca
Posts: 346
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Re: Help titrating BIPAP for sensitive sleeper

Post by ejbpesca » Mon Dec 22, 2025 3:36 pm

One is that many of your inhalation curves have a more pointed shape than we usually see.
Nocibur Wrote:
Anyway that's probably BiPAP Engineering. Would need the Oscar file uploaded to a third party server to show why that is (the one advantage Oscar has over SleepHQ).

semb2.jpg
[/quote]

The pointed wave shape in question is what most of mine look like with APAP therapy. Does anyone know why some have such a breath wave shape?

_________________
MachineMask

User avatar
Nocibur
Posts: 503
Joined: Sat Jan 25, 2025 6:08 am

Re: Help titrating BIPAP for sensitive sleeper

Post by Nocibur » Sun Dec 28, 2025 7:27 am

ejbpesca wrote: ↑
Mon Dec 22, 2025 3:36 pm
Does anyone know why some have such a breath wave shape?
Yes.

User avatar
ejbpesca
Posts: 346
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Re: Help titrating BIPAP for sensitive sleeper

Post by ejbpesca » Sun Dec 28, 2025 3:59 pm

Nocibur wrote: ↑
Sun Dec 28, 2025 7:27 am
ejbpesca wrote: ↑
Mon Dec 22, 2025 3:36 pm
Does anyone know why some have such a breath wave shape?
Yes.
Using APAP not BiLevel so what do you think causes the steep to a point waves?

_________________
MachineMask

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Re: Help titrating BIPAP for sensitive sleeper

Post by ozij » Sun Dec 28, 2025 10:41 pm

ejbpesca wrote: ↑
Sun Dec 28, 2025 3:59 pm
Nocibur wrote: ↑
Sun Dec 28, 2025 7:27 am
ejbpesca wrote: ↑
Mon Dec 22, 2025 3:36 pm
Does anyone know why some have such a breath wave shape?
Yes.
Using APAP not BiLevel so what do you think causes the steep to a point waves?
I too would love to hear Nocibur's explanation of that.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Nocibur
Posts: 503
Joined: Sat Jan 25, 2025 6:08 am

Re: Help titrating BIPAP for sensitive sleeper

Post by Nocibur » Mon Dec 29, 2025 3:37 am

ejbpesca wrote: ↑
Sun Dec 28, 2025 3:59 pm
Nocibur wrote: ↑
Sun Dec 28, 2025 7:27 am
ejbpesca wrote: ↑
Mon Dec 22, 2025 3:36 pm
Does anyone know why some have such a breath wave shape?
Yes.
Using APAP not BiLevel so what do you think causes the steep to a point waves?
Did you do that favor about Albercook that I asked you to do for me?

User avatar
ejbpesca
Posts: 346
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Re: Help titrating BIPAP for sensitive sleeper

Post by ejbpesca » Mon Dec 29, 2025 10:10 am

Nocibur wrote: ↑
Mon Dec 29, 2025 3:37 am
ejbpesca wrote: ↑
Sun Dec 28, 2025 3:59 pm
Nocibur wrote: ↑
Sun Dec 28, 2025 7:27 am
ejbpesca wrote: ↑
Mon Dec 22, 2025 3:36 pm
Does anyone know why some have such a breath wave shape?
Yes.
Using APAP not BiLevel so what do you think causes the steep to a point waves?
Did you do that favor about Albercook that I asked you to do for me?
I do not have an Albercook favor asked of me. I am curious as to why breath waves show as a steep rise to a sharp peak as opposed to a rounded shape.

_________________
MachineMask