Minimum pressure too low and causing microarousals?
-
funnyman26400
- Posts: 4
- Joined: Fri Sep 19, 2025 3:15 am
Minimum pressure too low and causing microarousals?
Started graph when i fell asleep see how normal tidal volume and resting respirations look until 7:15 and then it gets very choppy and seems to coincide with pressure increases you can also see it happen later in the night causing me to wake up to use the bathroom around 9:25am. Sleeping enviroment is totally dark cool and quiet, I sleep alone I'm wondering if 7 is too low of a minimum pressure. Wondering if anyone could chime in as to what is causing these transient spikes in my tidal volume and respirations I am fully asleep during those times
- Attachments
-
- screenshot-20250926-035625.png (185.92 KiB) Viewed 12577 times
-
- Screenshot from 2025-09-26 03-58-06.png (84.33 KiB) Viewed 12577 times
- ChicagoGranny
- Posts: 15532
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Minimum pressure too low and causing microarousals?
If you would get a free account at SleepHQ.com and post the link in this thread, members could easily zoom in to have a close look at the "event". SleepHQ is easy to use, and any personally identifying information in the link is redacted.
Do you do a double void before going to bed?
BTW, one pinkelpause per night is normal. Many people would be grateful to have this good fortune.
Do you do a double void before going to bed?
BTW, one pinkelpause per night is normal. Many people would be grateful to have this good fortune.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: Minimum pressure too low and causing microarousals?
I would not change anything...You are doing great
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
- Miss Emerita
- Posts: 3783
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Minimum pressure too low and causing microarousals?
The variability could reflect periods of REM sleep. Hard to know.
I basically agree with LSAT, though you might try some EPR to see whether that helps reduce flow limitations. I would leave your pressure settings right where they are and try EPR of 1 to start with. FLs may or may not affect the quality of your sleep, and EPR may or may not help. So it'd be an experiment.
And if you could do as ChicagoGranny suggests, that would be great.
I basically agree with LSAT, though you might try some EPR to see whether that helps reduce flow limitations. I would leave your pressure settings right where they are and try EPR of 1 to start with. FLs may or may not affect the quality of your sleep, and EPR may or may not help. So it'd be an experiment.
And if you could do as ChicagoGranny suggests, that would be great.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Minimum pressure too low and causing microarousals?
The volatility you are experiencing at 7:10 and again at 9:25 looks like REM.
The "choppiness" isn't a response to the pressure increase. The pressure increase is a response to REM and is a common characteristic of REM along with reduced patency.
I would consider increasing your minimum pressure so the pressure doesn't have to catch up during REM periods.
The "choppiness" isn't a response to the pressure increase. The pressure increase is a response to REM and is a common characteristic of REM along with reduced patency.
I would consider increasing your minimum pressure so the pressure doesn't have to catch up during REM periods.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2