New APAP User - Now with OSCAR Data

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
asian_string
Posts: 6
Joined: Wed Apr 02, 2025 5:44 pm

New APAP User - Now with OSCAR Data

Post by asian_string » Wed Apr 02, 2025 6:47 pm

Hi all,

After years of snoring and suspected sleep apnea, I finally completed a sleep study to find I have severe sleep apnea with an AHI of 42. Since then I've lurked here, ApneaBoard, and Reddit for a while; doing all the reading I can to fine tune my experience and give myself the best chance possible to maximise my therapy.

I have been on APAP for 4 weeks now, it's been a slow process to get used to the therapy and I have had some great nights where I wake up refreshed, though most nights I struggle and either take the mask off in my sleep, or wake up, remove the mask and fall back asleep without it.

I've likely been overzealous with my adjustments trying to find that perfect range, stemming from the sleep technician providing me with a range of 5-17 (low effort, range to wide), I had some good nights on that range, though most nights waking up feeling breathless. Since then I've been researching, reading forums, and started using OSCAR; the software has been great and I love having the data to track my progress, though I feel I've made too many adjustments and am now looking for a baseline to 're-start' my journey.

Starter Settings - ResMed AirSense 11 Auto
Mode: APAP
Pressure Min: 5.00 cmH2O
Pressure Max: 17.00 cmH2O
Climate Control: Auto
EPR: Full Time @ 3 cmH2O
Mask: Pillows
Ramp: On Auto
Response: Standard
Smart Start: On
Smart Stop: On

I was not using OSCAR at this stage, had a few good nights, though a lot of nights waking up breathless

Current Settings - ResMed AirSense 11 Auto
Mode: APAP
Pressure Min: 10.00 cmH2O
Pressure Max: 17.00 cmH2O
Climate Control: Auto
EPR: Full Time @ 3 cmH2O
Mask: Pillows (has been suggested I try Full Face)
Ramp: On Auto
Response: Standard
Smart Start: On
Smart Stop: On

OSCAR data shows a lot of RERA and CA events, leading me to believe the pressure settings to be incorrect, though less issues waking up breathless. From the little I know about interpreting OSCAR data, I am leaning towards a min. pressure of 9 and a max pressure of 13 - with my Median @ 9.92 and 95% @ 11.02 from recent OSCAR data of a successful night, though I wonder if a min. of 9 is too high. As I have changed settings up and down this will also affect my stats so I understand this is essentially flying blind currently.

I would be most grateful if someone is able to provide some input on a back to basics starting point, and I am prepared to undertake an actual self titration to determine my best settings for the best sleep. Now knowing the long term affects of sleep apnea (Alzheimer's) I am dedicated to optimising my sleep from today. I do have OSCAR data should anyone wish to review, finally, if you got this far - thank you.

EDIT: Grammar, added OSCAR screenshots.

Cheers,
AS

OSCAR 250323.jpg
OSCAR 250323.jpg (589.86 KiB) Viewed 21609 times
OSCAR 250322.jpg
OSCAR 250322.jpg (601.36 KiB) Viewed 21609 times
OSCAR 250321.jpg
OSCAR 250321.jpg (576.58 KiB) Viewed 21609 times
Last edited by asian_string on Wed Apr 02, 2025 9:43 pm, edited 3 times in total.

User avatar
zonker
Posts: 11514
Joined: Fri Jun 19, 2015 4:36 pm

Re: New APAP User - time to acclimatise?

Post by zonker » Wed Apr 02, 2025 8:10 pm

asian_string wrote: ↑
Wed Apr 02, 2025 6:47 pm


I was not using OSCAR at this stage,
asian_string wrote: ↑
Wed Apr 02, 2025 6:47 pm
OSCAR data shows a lot of RERA and CA events
so, either you're using oscar or not. pls DO use it and post some charts.

viewtopic/t172378/StickyIf-you-want-CPA ... Gquot.html
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

asian_string
Posts: 6
Joined: Wed Apr 02, 2025 5:44 pm

Re: New APAP User - time to acclimatise?

Post by asian_string » Wed Apr 02, 2025 8:44 pm

zonker wrote: ↑
Wed Apr 02, 2025 8:10 pm

so, either you're using oscar or not. pls DO use it and post some charts.

viewtopic/t172378/StickyIf-you-want-CPA ... Gquot.html
You might have missed it, though I mentioned I was not using OSCAR at that stage, or, in that point in time of my therapy. I have since been using an SD card in the machine and have OSCAR data on hand.
I do have OSCAR data should anyone wish to review, finally, if you got this far - thank you.
I have attached a few nights worth of data, I'm only new and there are many variables to account for, I can also adjust the visible data or adjust the panels to include more data should anyone prefer a different layout. Any assistance from the community would be greatly appreciated.

Thanks,
AS

OSCAR 250327.jpg
OSCAR 250327.jpg (616.35 KiB) Viewed 21620 times
OSCAR 250325.jpg
OSCAR 250325.jpg (601.8 KiB) Viewed 21620 times
OSCAR 250322.jpg
OSCAR 250322.jpg (601.36 KiB) Viewed 21620 times

User avatar
zonker
Posts: 11514
Joined: Fri Jun 19, 2015 4:36 pm

Re: New APAP User - Now with OSCAR Data

Post by zonker » Thu Apr 03, 2025 5:22 pm

this is weird. I just spent twenty minutes typing a reply which the forum didn't take.

won't type it again now as that's really aggravating. suffice to say I didn't read your original post carefully. sorry about that!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

asian_string
Posts: 6
Joined: Wed Apr 02, 2025 5:44 pm

Re: New APAP User - Now with OSCAR Data

Post by asian_string » Thu Apr 03, 2025 6:23 pm

zonker wrote: ↑
Thu Apr 03, 2025 5:22 pm
this is weird. I just spent twenty minutes typing a reply which the forum didn't take.

won't type it again now as that's really aggravating. suffice to say I didn't read your original post carefully. sorry about that!
Thanks Z, appreciate the attempt, unfortunate it didn't post! As a substitute for retyping your reply, do you confirm my suspicion that I have indeed adjusted my settings too often and I do need to find a good baseline and start again?

I am thinking of the following:

Min: 10
Max: 17
EPR: On, 3
Ramp: Off

Following that can then review some new data with a good baseline and go from there.

Cheers,
AS

User avatar
zonker
Posts: 11514
Joined: Fri Jun 19, 2015 4:36 pm

Re: New APAP User - Now with OSCAR Data

Post by zonker » Thu Apr 03, 2025 6:44 pm

asian_string wrote: ↑
Thu Apr 03, 2025 6:23 pm
zonker wrote: ↑
Thu Apr 03, 2025 5:22 pm
this is weird. I just spent twenty minutes typing a reply which the forum didn't take.

won't type it again now as that's really aggravating. suffice to say I didn't read your original post carefully. sorry about that!
Thanks Z, appreciate the attempt, unfortunate it didn't post! As a substitute for retyping your reply, do you confirm my suspicion that I have indeed adjusted my settings too often and I do need to find a good baseline and start again?

I am thinking of the following:

Min: 10
Max: 17
EPR: On, 3
Ramp: Off

Following that can then review some new data with a good baseline and go from there.

Cheers,
AS
basically, yes. otherwise you'd be what we call "dial winging". you can see on your charts that your pressure needs go up from 9 and don't come back down. it's accepted that one's minimum pressure does the bulk of the work.

if you want, you can set your max at 20. it won't hurt anything and could help if your pressure needs increase.

now stand back while I attempt to post this shorter message! :wink:

eta-it woiked!

also, stick with the changes you make for 3 or 4 nights as it may take a while for any changes to become effective. patience, young grasshopper, that's the key.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65444
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New APAP User - Now with OSCAR Data

Post by Pugsy » Thu Apr 03, 2025 8:19 pm

Using OSCAR can you provide screen shots of a couple of the statistics pages.
First one is the CPAP Statistics page and please make sure to include the section below called Changes to Device settings.
Second one is the OverView page.
Also when you share reports please include how you slept and how you feel......like did you sleep soundly or wake often...that kind of thing.
I may have to RISE but I refuse to SHINE.

asian_string
Posts: 6
Joined: Wed Apr 02, 2025 5:44 pm

Re: New APAP User - Now with OSCAR Data

Post by asian_string » Sun Apr 06, 2025 5:52 pm

Pugsy wrote: ↑
Thu Apr 03, 2025 8:19 pm
Using OSCAR can you provide screen shots of a couple of the statistics pages.
First one is the CPAP Statistics page and please make sure to include the section below called Changes to Device settings.
Second one is the OverView page.
Also when you share reports please include how you slept and how you feel......like did you sleep soundly or wake often...that kind of thing.
Hi Pugsey,

Thanks for your feedback, here is the Overview and Statistics Page. Essentially, except for one or two nights of actual restful sleep I find for the majority of the time I not getting proper deep sleep/aware of the mask/waking up breathless or like I need more air.
OSCAR - Overview 1.jpg
OSCAR - Overview 1.jpg (502.6 KiB) Viewed 21001 times
OSCAR - Overview 2.jpg
OSCAR - Overview 2.jpg (482.36 KiB) Viewed 21001 times
OSCAR - Stats.jpg
OSCAR - Stats.jpg (406.34 KiB) Viewed 21001 times
Cheers,
AS

asian_string
Posts: 6
Joined: Wed Apr 02, 2025 5:44 pm

Re: New APAP User - Now with OSCAR Data

Post by asian_string » Sun Apr 06, 2025 5:54 pm

zonker wrote: ↑
Thu Apr 03, 2025 6:44 pm

basically, yes. otherwise you'd be what we call "dial winging". you can see on your charts that your pressure needs go up from 9 and don't come back down. it's accepted that one's minimum pressure does the bulk of the work.

if you want, you can set your max at 20. it won't hurt anything and could help if your pressure needs increase.

now stand back while I attempt to post this shorter message! :wink:

eta-it woiked!

also, stick with the changes you make for 3 or 4 nights as it may take a while for any changes to become effective. patience, young grasshopper, that's the key.
Yeah I have been winging it, though now with the help of this forum I would like to begin fine-tuning my CPAP therapy.

Thanks for you help so far.

Cheers,
AS

asian_string
Posts: 6
Joined: Wed Apr 02, 2025 5:44 pm

Re: New APAP User - Now with OSCAR Data

Post by asian_string » Sun Apr 13, 2025 9:08 pm

Hi all,

Here is an update following up on the changes I've made. I still seem to be struggling to get a full night though my events are down which is good. I am lasting anywhere from 2 to 4 hours before I wake up, previously it was an out of breath feeling but now I'm just woken up from sleep, any thoughts?
250414.jpg
250414.jpg (864 KiB) Viewed 19966 times