[EDIT] Problem resolved somehow with fiddling. Heating normally now. Thanks for the ideas.[/EDIT]
I have an AirSense 10 which is no longer warming the water in the humidifier tank. I have confirmed that after hours of use the warming plate is cold. I think the heating hasn't worked since a morning I woke to find the machine had shut itself off and the power brick was unusually warm. Otherwise the machine works as it should.
The fuse doesn't seem to be blown as it passes a continuity test. Can anyone suggest 1) a source for a replacement humidifier warming plate (ebay and general web searches have not been fruitful) or 2) a reliable repair service? Thanks
[RESOLVED] AirSense 10 Humidifier Repair
[RESOLVED] AirSense 10 Humidifier Repair
Last edited by sfwood on Mon Oct 07, 2024 8:27 pm, edited 1 time in total.
- ChicagoGranny
- Posts: 15462
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: AirSense 10 Humidifier Repair
Have you been using it without heat? If so, did you get dry nasal passages or any other unwanted side effects?
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- Dog Slobber
- Posts: 4310
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AirSense 10 Humidifier Repair
Consider buying a used machine of FB market place for parts.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: AirSense 10 Humidifier Repair
You said you did a continuity test of the fuse. Did you measure the heater resistance?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Re: AirSense 10 Humidifier Repair
Thanks. I didn't measure heater resistance, as I didn't see an explanation of how to or what it should be.. This would confirm the heater plate is bad and not the control board? I would have just put in a new hearer plate but they don't seem to be available to mortals.
Re: [RESOLVED] AirSense 10 Humidifier Repair
thank you for updating!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg