Airsense 11 leaking air from around humidifier chamber
-
- Posts: 17
- Joined: Wed Dec 06, 2023 12:37 pm
Airsense 11 leaking air from around humidifier chamber
My AirSense 11 has developed an intermittent problem. Sometimes after I reinstall the humidifier chamber after refilling it, an air leak develops from somewhere around it when I next start the CPAP machine. The screen reports a large leak and the machine shuts off. I'll try reseating the chamber repeatedly, making sure it is completely seated, but no luck. I tried this fix here:
https://www.youtube.com/watch?v=kh-6lsiFY0E
but also no luck. The only thing that does solve it is swapping out the chamber for another I have on hand. This new chamber will run fine for awhile, but then it develops the issue. When I swap it out with the first chamber, and it runs okay for a while, but then the same problem develops again. It seems odd to me that simply swapping chambers back and forth fixes the problem temporarily. The only thing I can think of is that the silicone fittings decompress when the chamber isn't being used for a while, just enough to solve the air leakage.
Has anyone else had this issue? I'm still in the rental phase and I like to get this fixed before I own the machine.
https://www.youtube.com/watch?v=kh-6lsiFY0E
but also no luck. The only thing that does solve it is swapping out the chamber for another I have on hand. This new chamber will run fine for awhile, but then it develops the issue. When I swap it out with the first chamber, and it runs okay for a while, but then the same problem develops again. It seems odd to me that simply swapping chambers back and forth fixes the problem temporarily. The only thing I can think of is that the silicone fittings decompress when the chamber isn't being used for a while, just enough to solve the air leakage.
Has anyone else had this issue? I'm still in the rental phase and I like to get this fixed before I own the machine.
Re: Airsense 11 leaking air from around humidifier chamber
I had a very similar problem and ended up getting a new humidifier chamber which did solve the problem. When you close the lid to your water tank, check to see if the lid latch is going past the point where it should click into place. If so, try unlatching and lifting the lid till you meet resistance after the chamber is inside of the machine. I was able to get mine working again by doing that. I just had to be careful to not bump the lid where it would closer further and re-open a gap at the top for a leak.
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
Additional Comments: heated tube, unsure which one |
-
- Posts: 17
- Joined: Wed Dec 06, 2023 12:37 pm
Re: Airsense 11 leaking air from around humidifier chamber
Thanks! I'll give it a try. It does explain why swapping the chamber solves my problem: I have to open and fill that chamber before I swap it in and it probably latches correctly. It seems like my AirSense 11 has very narrow tolerances when it comes to how the chamber fits into the machines.
Re: Airsense 11 leaking air from around humidifier chamber
Yes, the humidifier can sometimes be quirky in fitting on the AS/AC 11. I've developed a technique of closing the chamber completely, then angling it downward when inserting into the machine. By angling downward, it seems to make sure that the connection is tight. I've had many times where it didn't connect correctly to the seal on the machine, and it won't work properly when I start the machine.
Eventually it becomes second nature...
Eventually it becomes second nature...
_________________
Machine: AirCurve 10 Vauto USA C2C CO |
Mask: AirFit™ F40 System - M/STD |
Additional Comments: AirMini Travel Machine |
Re: Airsense 11 leaking air from around humidifier chamber
I specifically created an account just to say THANK YOU! I kept having this happen and couldn't figure out the problem, but this works perfectly! Thank you so much!pisco21 wrote: ↑Wed Sep 18, 2024 12:26 pmYes, the humidifier can sometimes be quirky in fitting on the AS/AC 11. I've developed a technique of closing the chamber completely, then angling it downward when inserting into the machine. By angling downward, it seems to make sure that the connection is tight. I've had many times where it didn't connect correctly to the seal on the machine, and it won't work properly when I start the machine.
Eventually it becomes second nature...

Re: Airsense 11 leaking air from around humidifier chamber
@Kaden, you're welcome! Sorry I didn't see this until now, but it's been pretty crazy at work... glad I could help! 

_________________
Machine: AirCurve 10 Vauto USA C2C CO |
Mask: AirFit™ F40 System - M/STD |
Additional Comments: AirMini Travel Machine |
Re: Airsense 11 leaking air from around humidifier chamber
THANKS SO MUCH!!!!!!! Since I took your advice, I haven’t had any issues—IT WORKS EVERY TIME!!!!pisco21 wrote: ↑Wed Sep 18, 2024 12:26 pmYes, the humidifier can sometimes be quirky in fitting on the AS/AC 11. I've developed a technique of closing the chamber completely, then angling it downward when inserting into the machine. By angling downward, it seems to make sure that the connection is tight. I've had many times where it didn't connect correctly to the seal on the machine, and it won't work properly when I start the machine.
Eventually it becomes second nature...
I had been struggling with this issue ever since my insurance company “upgraded” my machine from an AirSense 10 (problem free for years) to an 11. It was so frustrating! THIS HAS MADE MY LIFE SO MUCH EASIER!
Thanks again! I really appreciate your post!
Re: Airsense 11 leaking air from around humidifier chamber
My unit started misbehaving this way. After several frustrating attempts to firmly place it back in, I found that using a gentler touch and not trying to push it in too firmly has worked well.
n a similar vein, when refilling the water chamber, if I tried to tilt it all the way open, it would always fall back; if I opened it more gently and not all the way, there was a sweet6 spot where it stayed put while I poured in some water.
Now if only the markings were a little easier to read while standing above and pouring in the water...
n a similar vein, when refilling the water chamber, if I tried to tilt it all the way open, it would always fall back; if I opened it more gently and not all the way, there was a sweet6 spot where it stayed put while I poured in some water.
Now if only the markings were a little easier to read while standing above and pouring in the water...
-
- Posts: 1
- Joined: Tue Jun 17, 2025 8:46 am
Re: Airsense 11 leaking air from around humidifier chamber
pisco21 wrote: ↑Wed Sep 18, 2024 12:26 pmYes, the humidifier can sometimes be quirky in fitting on the AS/AC 11. I've developed a technique of closing the chamber completely, then angling it downward when inserting into the machine. By angling downward, it seems to make sure that the connection is tight. I've had many times where it didn't connect correctly to the seal on the machine, and it won't work properly when I start the machine.
Eventually it becomes second nature...
I want to thank you for your suggestion. The tub on my Airsense11 has not been seating properly for quite some time now. Then, last night it took me almost 20 minutes to get it to work. Your tip works!! Thanks again, Pete
Re: Airsense 11 leaking air from around humidifier chamber
Here is another thing that may help.
When filling your chamber, after cleaning and rinsing the chamber, close it first before filling with water.
Then, holding it with the gasket and holes facking upward,fill it through the hold with the silicone gasket just to the level where the lip on the front stops to allow the chamber to enter the machine. Then rub a small amount of water on the gasket, tilt the unit slightly downward as it enters the machine.
You can find videos showing this method on YouTube. Once I started doing it this way, I can always fill the chamber just to the fill line, and it always seals properly with no leaks.
When filling your chamber, after cleaning and rinsing the chamber, close it first before filling with water.
Then, holding it with the gasket and holes facking upward,fill it through the hold with the silicone gasket just to the level where the lip on the front stops to allow the chamber to enter the machine. Then rub a small amount of water on the gasket, tilt the unit slightly downward as it enters the machine.
You can find videos showing this method on YouTube. Once I started doing it this way, I can always fill the chamber just to the fill line, and it always seals properly with no leaks.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset |
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto