One issue I've had since I can remember is that I can't stay asleep for longer than maybe 45 minutes. Every night I go to sleep, I am constantly waking up and falling asleep over and over until it's time to get out of bed. This is especially difficult dealing with when I have this mask on my face and I'm having to check and make sure I didn't pull it off. I'm disappointed to see that the apneas being reduced didn't stop my issue of constantly waking up, and am curious as to what you guys might see from the data. Today I only have one day as I just discovered how to use oscar and such, so pardon the lack of data. Would love to get some feedback on what you guys are seeing, it just looks like a bunch of nothing to me as of right now.
That aside, I've been using cpap semi-consistently now for a little over a month. There have been maybe five or so nights in the past month where I'll fall asleep without it, so maybe that has slowed down my progress. Overall, I can't say I've noticed many benefits of using the cpap yet. Hoping to feel more clearerminded and less daytime drowsiness soon, as I know everybody is different!
I use a RedMed AIrsense 11 and the P10 mask.
Struggling to stay asleep
Re: Struggling to stay asleep
So you have a long history of sleep maintenance insomnia. And starting CPAP didn't fix it all by itself. That most likely means that prior to your sleep apnea diagnosis you had two separate sleep issues: The sleep apnea and the insomnia problem.
Now, it was not an unreasonable hypothesis that the apnea may have been causing or aggravating the insomnia, but since you are still waking up frequently for long enough periods of time to remember the wakes, it may be time to accept that you will need to focus on each problem both at the same time and through separate treatment methods. More on this in a bit.
Sounds like you are a night-time worrier. Now, in addition to worrying about waking up all the time, worrying about how little you've slept since the last wake, and worrying about how little time there is before morning to get some good sleep, you are also worrying about whether you've pulled the mask off. All this worrying is quite likely aggravating the sleep maintenance insomnia.This is especially difficult dealing with when I have this mask on my face and I'm having to check and make sure I didn't pull it off.
Out of curiosity: When you wake up in the night, do you often look at a clock (including a cell phone or smart watch) to see what time it is and then start "calculating" how much sleep you got since the last wake up?
The worry/fear of pulling the mask off is a common one for newbies. There is a lot of sensory stuff to get used to when learning how to sleep (and sleep well) with a CPAP mask on your nose. Sometimes the fix is just to allow yourself the freedom to not worry quite so much. In other words, don't worry about pulling the mask off when you are asleep: If you happen to wake up and the mask is not on your face, just calmly put the mask back on and turn the machine back on if you also turned it off in your sleep or if it turned itself off. Don't start fretting about why it happened. Just accept it. And accept that as your body genuinely acclimates to this new way of sleeping, the number of times that you pull the mask off while asleep will go down.
It's also worth checking to see if AutoStop is turned ON in your settings. If AutoStop is set to ON, the I'd suggest turning it OFF. That way if you do pull the mask off while you are actually asleep, the noise of the machine blowing air into the mask might wake you up enough for you to realize that you need to put the mask back on.
Now having said all that, it's also important to add this: If you are awake and are tempted to pull the mask off to make it easier to get back to sleep, you need to resist that urge. Consciously pulling your mask off while you are awake and know that you are awake simply reenforces your subconscious mind's resistance to the idea that you need to learn to sleep with the mask on your nose.
Here's what I'm seeing in your data, even at this "whole night scale"I'm disappointed to see that the apneas being reduced didn't stop my issue of constantly waking up, and am curious as to what you guys might see from the data. Today I only have one day as I just discovered how to use oscar and such, so pardon the lack of data. Would love to get some feedback on what you guys are seeing, it just looks like a bunch of nothing to me as of right now.
1) Lots of evidence of arousal breathing, backing up your own feeling that you wake up frequently during the night. My guess is that if you posted the data at SleepHQ and I were to scroll through it using a horizontal scale where we can see the individual breaths, all or most of those CAs being scored would have a large inhalation or two just before they are scored. In that case, those CAs are post-arousal "false positives"---in other words, they would not be scored on an in-lab sleep test because you were awake when they happened. They're probably normal sleep transitional events---it's common, even for people with normal sleep to have a sleep transitional "CA" or two when they are drifting off to sleep and control of the respiration is being transferred to the autonomic nervous system.
2) You are using a ramp that starts at 4cm, but when it starts up, the pressure ramps up pretty quickly. It's worth asking: Do you really need the ramp? When you first turn the machine on and it's blowing at 4cm, do you feel comfortable breathing? If not, describe what's uncomfortable about the breathing. It's also worth asking whether you are sensitive enough to notice (and/or worry about) the fact that the machine starts to increase the pressure pretty soon after you turn it on.
3) Is your ramp set to AUTO? If so, then your data indicates that you are initially falling asleep within about 10-15 minutes of starting the machine. That indicates you don't have any problems with sleep initiation---i.e. falling asleep at the beginning of the night.
4) You are not using EPR. EPR is a setting that reduces the pressure on exhalation, which makes it easier to exhale. For many people, breathing with EPR on feels more natural. The EPR setting corresponds to the drop in pressure. But the pressure will never drop below 4cm. So with a ramp set to start at 4cm, you won't notice any difference in how the machine feels until the pressure increases.
5) You have your minimum pressure set to 6cm. You might do better turning the ramp OFF and turning EPR to FULL TIME with a setting of 2 or 3. That way you would have an inhalation pressure of 6cm and an exhalation pressure of 4cm when you turn the machine on and you'd have the exhale relief on every exhalation you take during the night. If any of your arousals are being caused by difficulty with exhaling against the pressure, turning on EPR may help.
6) You may have a leak problem. While the time spend in Large Leak territory (5.77%) is so small that Mr. Green Smiley Face is going to show up on the machine's LCD and the "Leak" part of the MyAir report will say things look pretty good, the fact is you've got some level of leaks going on all night long. Now, if you were sleeping well and had plenty of energy during the day, I'd say just ignore these leaks. But you are NOT sleeping well. And so it begs the questions: Are the leaks potentially feeding the sleep maintenance insomnia problem? Do you wake up with a dry mouth? Do you ever wake up feeling like air is blowing into your eyes? And since you are using a P10 mask, it's also worth asking: Do you find yourself breathing through your mouth during the daytime? What to do about the leaks is something for another whole post, but it really depends on the answers to those questions.
7) You have a lot of flow limitations being scored. That may indicate a bit more pressure might be useful. Or it could be that turning EPR on will help smooth them out. Or they may be related to something like nasal congestion or a deviated septum and there's not much to do about them other than not worry about them. Or it could just be that your normal wake breathing looks sufficiently ragged enough to the machine that it scores flow limitations during periods when you are not really asleep. In other words, it's far, far too early to try to tackle the question, "Are the flow limitations are real problem?"
Yep. Using the cpap semi-consistently is a good way to slow your progress down. Every time you allow yourself to consciously sleep without the machine, you are teaching your body and the unconscious part of your brain that you don't really have to learn how to sleep with this dang mask on your nose.That aside, I've been using cpap semi-consistently now for a little over a month. There have been maybe five or so nights in the past month where I'll fall asleep without it, so maybe that has slowed down my progress.
Patience, grasshopper. Patience.Overall, I can't say I've noticed many benefits of using the cpap yet. Hoping to feel more clearerminded and less daytime drowsiness soon, as I know everybody is different!
We all start out CPAP therapy hoping that we'll start feeling substantially better almost right away. And, yes, there are a few lucky folks who do.
But for most of us? It takes several weeks to several months of consistently using the machine all night long before we feel much better. And that's only if there are no other sleep problems to deal with.
CPAP fixes one thing---sleep disordered breathing, aka obstructive sleep apnea. And if OSA is the only health problem a person has, then yes the person will start feeling better after they learn to sleep well with the machine.
But if there are other sleep problems, like sleep maintenance insomnia, as well as OSA? Then CPAP ain't going to fix those other problems.
In other words, you need to work on the insomnia issues at the same time you are teaching your body to sleep with the mask on your nose.
So that brings up a whole lot of other insomnia related questions:
1) What does your sleep hygiene look like? (If you don't know what sleep hygiene is, google it.)
2) What behavior patterns do you have that are feeding the sleep maintenance insomnia? In particular, are you a "clock watcher" when you find yourself awake in the middle of the night?
3) What are your expectations of a "good night's sleep?"
4) Is there a disconnect between how much time you think you are awake in the middle of the night versus how much time you actually are awake in the middle of the night? (Most people with sleep maintenance insomnia over estimate both the number of times they wake up and the length of time they are awake. Consequently they underestimate how much sleep they actually got.)
5) What kind of a sleep test did you have that led to the OSA diagnosis? And did it say anything about spontaneous arousals and WASO (Wake after sleep onset)? Did the sleep report include a hypnogram so you can see when the wakes occurred?
6) What kinds of things have you tried in the past to reduce the number of night time wakes?
7) Do you use something like a FitBit or SmartWatch (including an Apple Watch) to try to track your sleep? If so, what does that data say about your sleep and the number of wakes you have each night.
Finally, I'll offer this advice: See if you can find a copy of Sound Sleep, Sound Mind by Dr. Barry Krakow. It is a well written guide to self-treat insomnia, and it's full of information about what things insomniacs do that aggravate the insomnia as well as practical tips on how to replace those insomnia-inducing behaviors with behaviors that encourage better sleep.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Re: Struggling to stay asleep
Argh... I had a giant reply typed up and accidentally closed it. Replies won't be as detailed as they originally were and apologies if I miss anything on my second way through. I've also attached images of my sleep study in case they are helpful.
Sleep maintenance insomnia... Interesting. Also just wanted to thank you for taking the time to parse through my data and giving such a thoughtful response. That means a lot, thank you so much.
1. Makes sense. I'll keep sleeping with it on and report back with more data. One day isn't a whole lot to work off of.
2/3 I'll turn the ramp off. Usually what I feel is a slight pushback when I start the machine upo, like my airways are overwhelmed. I get used to it enough to fall asleep though.
4. I don't have the ability to configure EPR on my own. I will have to see the clinic about this.
5. I'll bring that up with the clinic as well. Ramp off, EPR on full time, and look at the pressure. Got it!
6. This was my main suspect before posting. I am a natural mouth breather who has been shamed to nasal breathing. I now breathe through my nose anytime I'm not doing heavy exercise naturally. My head anatomy definitely is one of a mouth breather though - recessed jaw/chin and smaller nasal airway. Should I consider doing a full face mask? I found them extremely uncomfortable and hard to sleep with while having facial ahir and such.
7. WIll keep this in mind and report back when I have a lot more data.
------ Insomnia ------
I'll look towards tackling the insomnia. They put me on something like Zolpedim? I can't remember if that's it or what the name was but it had a z in it. They put me on a dose when I took my sleep test but I didn't see any benefits in helping me stay asleep at least.
1. My sleep hygiene is OK. Could be better. I fall asleep and go to bed around tghe same time everyday, give or take 30 mins to an hour. Sometimes I do an extra few sleep periods on the weekends. I don't smoke, drink, or take caffeine. I try not to stay in bed for doing anything other than sleeping. My room's cool. Maybe too much stress on the mind doesn't help, I work in software development and I 1000% sleep worse when deadlines are approaching or work is stressful. I sometimes eat close to bedtime but not too frequently. Anything I'm missing?
2. No clock watching at all. Maybe just the anxiety of having to get up soon on my office days. Blackout curtains have helped a lot - seeing daylight and knowing I have to wake up in an hour or two is the absolute worst feeling ever and definitely does not help. If I wake up and can guess that it's an hour or two within I have to wake up, I'll probably wake up every fifteen minutes at that point. Reflecting on this and writing it down, maybe this is stress related...
3/4. Waking up feeling ready to take the day. Getting at least eight hours in (which I need to do more). Not waking up every 30 minutes haha. Seriously, I'll have to count the number of times I wake up one of these days. It's bad!
5/ ncluded my sleep test in the attachments. Had an at home one and a in office one which I included.
6. Good question. I've tried vitamin D supplementation, working out before bed, tiring myself out completely as well. Not sure if they've worked.
7. Yes, I do have an Apple WAtch. I haven't gotten any sleep data recently though.
I'll check that book out! Again, thank you so much for taking the time to reply. Wish I had my oriignal message but oh well, hopefully I have addressed everything!
Sleep maintenance insomnia... Interesting. Also just wanted to thank you for taking the time to parse through my data and giving such a thoughtful response. That means a lot, thank you so much.
1. Makes sense. I'll keep sleeping with it on and report back with more data. One day isn't a whole lot to work off of.
2/3 I'll turn the ramp off. Usually what I feel is a slight pushback when I start the machine upo, like my airways are overwhelmed. I get used to it enough to fall asleep though.
4. I don't have the ability to configure EPR on my own. I will have to see the clinic about this.
5. I'll bring that up with the clinic as well. Ramp off, EPR on full time, and look at the pressure. Got it!
6. This was my main suspect before posting. I am a natural mouth breather who has been shamed to nasal breathing. I now breathe through my nose anytime I'm not doing heavy exercise naturally. My head anatomy definitely is one of a mouth breather though - recessed jaw/chin and smaller nasal airway. Should I consider doing a full face mask? I found them extremely uncomfortable and hard to sleep with while having facial ahir and such.
7. WIll keep this in mind and report back when I have a lot more data.
------ Insomnia ------
I'll look towards tackling the insomnia. They put me on something like Zolpedim? I can't remember if that's it or what the name was but it had a z in it. They put me on a dose when I took my sleep test but I didn't see any benefits in helping me stay asleep at least.
1. My sleep hygiene is OK. Could be better. I fall asleep and go to bed around tghe same time everyday, give or take 30 mins to an hour. Sometimes I do an extra few sleep periods on the weekends. I don't smoke, drink, or take caffeine. I try not to stay in bed for doing anything other than sleeping. My room's cool. Maybe too much stress on the mind doesn't help, I work in software development and I 1000% sleep worse when deadlines are approaching or work is stressful. I sometimes eat close to bedtime but not too frequently. Anything I'm missing?
2. No clock watching at all. Maybe just the anxiety of having to get up soon on my office days. Blackout curtains have helped a lot - seeing daylight and knowing I have to wake up in an hour or two is the absolute worst feeling ever and definitely does not help. If I wake up and can guess that it's an hour or two within I have to wake up, I'll probably wake up every fifteen minutes at that point. Reflecting on this and writing it down, maybe this is stress related...
3/4. Waking up feeling ready to take the day. Getting at least eight hours in (which I need to do more). Not waking up every 30 minutes haha. Seriously, I'll have to count the number of times I wake up one of these days. It's bad!
5/ ncluded my sleep test in the attachments. Had an at home one and a in office one which I included.
6. Good question. I've tried vitamin D supplementation, working out before bed, tiring myself out completely as well. Not sure if they've worked.
7. Yes, I do have an Apple WAtch. I haven't gotten any sleep data recently though.
I'll check that book out! Again, thank you so much for taking the time to reply. Wish I had my oriignal message but oh well, hopefully I have addressed everything!
- Attachments
-
- d2.png (345.68 KiB) Viewed 4786 times
-
- mspaint_dE367BXpCj.png (268.75 KiB) Viewed 4786 times
- Miss Emerita
- Posts: 3765
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Struggling to stay asleep
You can change your settings yourself, if you like.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
Scroll down until you find your machine.
Your in-lab sleep study is somewhat unsatisfying because you had only 100 minutes of sleep time during the diagnostic part of your sleep test. You had a lot more spontaneous arousals per hour of sleep time during the therapeutic part of the study than during the diagnostic part, though the very short sleep time for the diagnostic part means it's hard to put a lot of weight on that, and some of that increase in spontaneous arousals may just have been due to the disruptive experience of wearing a mask.
I also notice that your AHI was almost entirely composed of hypopneas, and that almost all of them happened while you were on your back. This leads me to wonder whether you usually sleep on your back.
One small thing: it's recommend that people *not* exercise close to bedtime. During the hour leading up to bedtime, it's ideal if you can do calming things that don't involve using screens.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
Scroll down until you find your machine.
Your in-lab sleep study is somewhat unsatisfying because you had only 100 minutes of sleep time during the diagnostic part of your sleep test. You had a lot more spontaneous arousals per hour of sleep time during the therapeutic part of the study than during the diagnostic part, though the very short sleep time for the diagnostic part means it's hard to put a lot of weight on that, and some of that increase in spontaneous arousals may just have been due to the disruptive experience of wearing a mask.
I also notice that your AHI was almost entirely composed of hypopneas, and that almost all of them happened while you were on your back. This leads me to wonder whether you usually sleep on your back.
One small thing: it's recommend that people *not* exercise close to bedtime. During the hour leading up to bedtime, it's ideal if you can do calming things that don't involve using screens.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: Struggling to stay asleep
Thank you so much for the tip on modifying the settings on my own. Saved me the money of a visit and the time of the drive. I was able to get it working. I just turned the EPR on full time at level 2 and also got rid of the ramp time. I'll give it a shot for a few days. That ought to cover the suggestions I think...
I agree on the sleep test, it was very short. Hypopneas can kind of be false positives too, is that correct? It's not necessarily that I stopped breathing? I think what I found when I first started cpap was that I was experiencing a few central apneas, but I don't know if that is still present. I typically sleep on my side. They made me sleep on my back at the sleep exam. I do need to be careful about the screens, I stare at screens for pretty much every moment I'm not sleeping. Bad habit.
I do have a home study that I took around two years ago. The difference here is that this is prior to me getting a tonsillectomy and a turbinate reduction. As you can see my ahi did go down but the sleep apnea is still here, probably due to my forementioned facial anatomy and I've been cursed with having a big neck as well!
I went ahead and attached my older sleep exam but again, not sure if it'll be too helpful now. Thanks again for taking the time to reply guys.
I agree on the sleep test, it was very short. Hypopneas can kind of be false positives too, is that correct? It's not necessarily that I stopped breathing? I think what I found when I first started cpap was that I was experiencing a few central apneas, but I don't know if that is still present. I typically sleep on my side. They made me sleep on my back at the sleep exam. I do need to be careful about the screens, I stare at screens for pretty much every moment I'm not sleeping. Bad habit.
I do have a home study that I took around two years ago. The difference here is that this is prior to me getting a tonsillectomy and a turbinate reduction. As you can see my ahi did go down but the sleep apnea is still here, probably due to my forementioned facial anatomy and I've been cursed with having a big neck as well!
I went ahead and attached my older sleep exam but again, not sure if it'll be too helpful now. Thanks again for taking the time to reply guys.
- Attachments
-
- chrome_PkwY688GdN.png (170.72 KiB) Viewed 4579 times
- Miss Emerita
- Posts: 3765
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Struggling to stay asleep
Hypopneas involve shallow breathing with a drop in O2 levels. They can be central, obstructive, or mixed. More here:
https://www.sleepfoundation.org/sleep-apnea/hypopnea
I really wish they hadn't made you sleep on your back, given that you are a side-sleeper.
At any rate, in a day or two, please post a chart with your revised settings so we can see what kind of difference, if any, they made. And do also include information about how rested or otherwise you felt the next day.
https://www.sleepfoundation.org/sleep-apnea/hypopnea
I really wish they hadn't made you sleep on your back, given that you are a side-sleeper.
At any rate, in a day or two, please post a chart with your revised settings so we can see what kind of difference, if any, they made. And do also include information about how rested or otherwise you felt the next day.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: Struggling to stay asleep
Hi guys, here is a few nights with the customizations recommended. Let me know what y'all think! I tacked on an extra night without configurations that I managed to get recorded.
How do I feel? I can't say I felt anything particularly new when waking up after changing my configs. I do feel the difference in pressure while breathing, however. It is much, much more comfortable when exhaling and I greatly appreciate the EPR feature. There have been two or three times in one of my night time awakenings where my breaths feel shallow, but I just need a slight adjustment in my mask until it feels normal again. When waking up for the past few nights, I definitely feel a little tired but I'm not dragging myself out of bed by any means. I've had a little bit of daytime drowsiness over the past few days as well, which isn't necessarily abnormal. Just paying my sleep debt still from all the years of sleep apnea I imagine.
Thanks again guys.




How do I feel? I can't say I felt anything particularly new when waking up after changing my configs. I do feel the difference in pressure while breathing, however. It is much, much more comfortable when exhaling and I greatly appreciate the EPR feature. There have been two or three times in one of my night time awakenings where my breaths feel shallow, but I just need a slight adjustment in my mask until it feels normal again. When waking up for the past few nights, I definitely feel a little tired but I'm not dragging myself out of bed by any means. I've had a little bit of daytime drowsiness over the past few days as well, which isn't necessarily abnormal. Just paying my sleep debt still from all the years of sleep apnea I imagine.
Thanks again guys.