F&P Evora Full Face Mask
F&P Evora Full Face Mask
I got the mask last week and had zero leaks each day. I washed it this weekend and now it is leaking badly. I have read the manual and watched videos on it. I will try to contact the medical device store tomorrow but often they don't answer the phone. I will use the nasal mask which leaks terribly and keeps me awake. Any suggestions? I find this mask extremely hard to put on and take off - and now leaks!
- ChicagoGranny
- Posts: 15470
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: F&P Evora Full Face Mask
Have you been able to identify the position of the leak?
When you washed the mask, did you damage the Anti-Asphyxia Valve?
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- vandownbytheriver
- Posts: 542
- Joined: Sat Feb 03, 2024 11:42 pm
Re: F&P Evora Full Face Mask
The Evora FF is what's known as a 'hybrid' mask, nasal with mouth covering. I find if I undo both side hooks, put the mask under my nose, then pull the headgear over, I can hook the side pieces and be going.
Just fit the mask without the headgear, you should be able to find the leak while holding the cushion to your face.
Just fit the mask without the headgear, you should be able to find the leak while holding the cushion to your face.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: F&P Evora Full Face Mask
Tell me more about this valve, please. It might be the problem.ChicagoGranny wrote: ↑Sun Jun 23, 2024 6:12 pmHave you been able to identify the position of the leak? No
When you washed the mask, did you damage the Anti-Asphyxia Valve?
- vandownbytheriver
- Posts: 542
- Joined: Sat Feb 03, 2024 11:42 pm
Re: F&P Evora Full Face Mask
I don't believe the Evora FF has this valve. The Vitera has one.
https://www.fphcare.com/us/homecare/sle ... vora-full/
Examine where the small tube snaps onto the mask spigot.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: F&P Evora Full Face Mask
There is flexible flap located in the mask frame right above where the hose connects. This is the antiaphyxiation valve.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Evora Full Face Mask - Fitpack |
| Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
- ChicagoGranny
- Posts: 15470
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: F&P Evora Full Face Mask
Yep.Okie bipap wrote: ↑Mon Jun 24, 2024 1:58 pmThere is flexible flap located in the mask frame right above where the hose connects. This is the antiaphyxiation valve.
The flap needs to close when the circuit is pressurized. When the machine is not running, the flap opens to prevent CO2 buildup in the mask.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- ChicagoGranny
- Posts: 15470
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: F&P Evora Full Face Mask
Members don't have to comment when they don't know what they are talking about.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- vandownbytheriver
- Posts: 542
- Joined: Sat Feb 03, 2024 11:42 pm
Re: F&P Evora Full Face Mask
Well, made me look! Yep, mine has one. It's hidden behind the oval hose connection.ChicagoGranny wrote: ↑Mon Jun 24, 2024 2:59 pmMembers don't have to comment when they don't know what they are talking about.
Thanks for your gentle correction.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: F&P Evora Full Face Mask
I fixed it. It was the little flexible anti-asphyxiation valve that is located between the mask and the tube coming out of the mask. Somehow it had gotten misaligned. No leaks now but, for me, an uncomfortable mask.