High heartrate and capped upper pressure

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
bbridge_GA
Posts: 50
Joined: Sun Sep 22, 2013 5:12 am

High heartrate and capped upper pressure

Post by bbridge_GA » Sat May 25, 2024 3:07 pm

I wear a Fitbit and like to use it to track my sleep quality and heartrate. My resting heartrate ranges from 53-80 and up to 145 max when exercising which I think is pretty healthy. My Fitbit has recorded 3 dates recently when my heartrate was between 160-180 while asleep. I didn't know if it was a malfunction of the Fitbit. I haven't looked at my CPAP data recently. This made me look at my data on Oscar to see what was going on. One of these events (heartrate of 180) correlated with a 30 second apnea that was preceded earlier by some strange waveforms earlier in the night. The others were not as clear of a relationship to a long apnea or anything unusual. I did notice that my pressure setting was capped at 10. My average pressure is typically around 6-7 with my 95% pressure around 9. I don't know why the cap was reduced to 10, but apparently there are times when I need more pressure. I have adjusted the top setting to 18 and plan to see if there are any more of these high hearrate events. I have also been having a lot of daytime sleepyness recently and hope that gets better with this adjustment. Any ideas about the funny waveforms? Here is the whole night.
https://prnt.sc/qiW9hGH-W9kf

_________________
MachineMask
Additional Comments: changed from cpap to apap April 2014
Last edited by bbridge_GA on Sat May 25, 2024 5:54 pm, edited 2 times in total.
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15

User avatar
ChicagoGranny
Posts: 15485
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: High heartrate and capped upper pressure

Post by ChicagoGranny » Sat May 25, 2024 3:44 pm

bbridge_GA wrote:
Sat May 25, 2024 3:07 pm
up to 145 max when exercising which I think is pretty healthy.
How old are you?
bbridge_GA wrote:
Sat May 25, 2024 3:07 pm
I have adjusted the top setting to 18
Many of us, even those with low-pressure needs, set the max pressure to 20. The machine will not go any higher than it needs to.
bbridge_GA wrote:
Sat May 25, 2024 3:07 pm
Any ideas about the funny waveforms? I couldn't upload the pictures.
Use imgur.com to host the pics. Or easier, get a free account at sleephq.com and post your data link in this thread.

Data is critical for members trying to help you.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: High heartrate and capped upper pressure

Post by Julie » Sat May 25, 2024 3:44 pm

Can you post pix of those wave forms?

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: High heartrate and capped upper pressure

Post by Miss Emerita » Sat May 25, 2024 5:42 pm

I’d recommend that you ask your doctor for a Holter monitor to see what’s going on.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

bbridge_GA
Posts: 50
Joined: Sun Sep 22, 2013 5:12 am

Re: High heartrate and capped upper pressure

Post by bbridge_GA » Sat May 25, 2024 5:52 pm

Julie wrote:
Sat May 25, 2024 3:44 pm
Can you post pix of those wave forms?
I added a link to the image of the whole night to the original post.

_________________
MachineMask
Additional Comments: changed from cpap to apap April 2014
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15

bbridge_GA
Posts: 50
Joined: Sun Sep 22, 2013 5:12 am

Re: High heartrate and capped upper pressure

Post by bbridge_GA » Sat May 25, 2024 5:58 pm

bbridge_GA wrote:
Sat May 25, 2024 5:52 pm
Julie wrote:
Sat May 25, 2024 3:44 pm
Can you post pix of those wave forms?
I added a link to the image of the whole night to the original post.
Here is the 2am apnea that corresponds to the heartrate being really high and the wave form earlier in the night zoomed in.
https://prnt.sc/eR7rcC6dy0Dm
https://prnt.sc/qiW9hGH-W9kf

_________________
MachineMask
Additional Comments: changed from cpap to apap April 2014
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15

bbridge_GA
Posts: 50
Joined: Sun Sep 22, 2013 5:12 am

Re: High heartrate and capped upper pressure

Post by bbridge_GA » Sat May 25, 2024 6:05 pm

Miss Emerita wrote:
Sat May 25, 2024 5:42 pm
I’d recommend that you ask your doctor for a Holter monitor to see what’s going on.
I am 63 with a normal ekg and normal heart ultrasound. My ex was a holter tech and put one on me in 1985. He said my heart rate went crazy in the middle of the night. In hindsight, I have always wondered if that wasn't when I was first diagnosed with sleep apnea...via holter and 20 years before my formal diagnosis. A follow-up with a cardiologist showed a slight murmur... that no one has mentioned since then.

_________________
MachineMask
Additional Comments: changed from cpap to apap April 2014
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15

bbridge_GA
Posts: 50
Joined: Sun Sep 22, 2013 5:12 am

Re: High heartrate and capped upper pressure

Post by bbridge_GA » Sat May 25, 2024 6:14 pm

bbridge_GA wrote:
Sat May 25, 2024 5:58 pm
bbridge_GA wrote:
Sat May 25, 2024 5:52 pm
Julie wrote:
Sat May 25, 2024 3:44 pm
Can you post pix of those wave forms?
I added a link to the image of the whole night to the original post and here https://prnt.sc/qiW9hGH-W9kf
Here is the 2am apnea that corresponds to the heartrate being really high https://prnt.sc/daOkcjDZIUJm and the wave form earlier in the night zoomed in.
https://prnt.sc/eR7rcC6dy0Dm
Here is this past week with the upper limit raised. https://prnt.sc/v8S7U3XRwA0V
thanks for looking at these!

_________________
MachineMask
Additional Comments: changed from cpap to apap April 2014
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15

User avatar
robysue1
Posts: 1311
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: High heartrate and capped upper pressure

Post by robysue1 » Sat May 25, 2024 8:18 pm

bbridge_GA wrote:
Sat May 25, 2024 5:58 pm
Here is the 2am apnea that corresponds to the heartrate being really high and the wave form earlier in the night zoomed in.
https://prnt.sc/eR7rcC6dy0Dm
https://prnt.sc/qiW9hGH-W9kf
Sure wish rubicon would take a look at that zoomed in area.

Also, could you zoom in further on just the area between 0:20 and 0:30---in other words, zoom in close enough where we can see the individual breaths.

And, do you remember being restless at all around this time? Do you remember any wakes during the night?

Finally, do you remember waking up to deal with any leaks during the night?
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: High heartrate and capped upper pressure

Post by Julie » Sat May 25, 2024 8:32 pm

Lots of us have "slight murmurs" that are nothing to worry about in themselves without other accompanying problems that may or may not matter either.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: High heartrate and capped upper pressure

Post by Miss Emerita » Sat May 25, 2024 10:47 pm

bbridge_GA wrote:
Sat May 25, 2024 6:05 pm
Miss Emerita wrote:
Sat May 25, 2024 5:42 pm
I’d recommend that you ask your doctor for a Holter monitor to see what’s going on.
I am 63 with a normal ekg and normal heart ultrasound. My ex was a holter tech and put one on me in 1985. He said my heart rate went crazy in the middle of the night. In hindsight, I have always wondered if that wasn't when I was first diagnosed with sleep apnea...via holter and 20 years before my formal diagnosis. A follow-up with a cardiologist showed a slight murmur... that no one has mentioned since then.
1985? Time for another one. All kinds of reasons you might be having a high heart rate.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
ChicagoGranny
Posts: 15485
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: High heartrate and capped upper pressure

Post by ChicagoGranny » Sun May 26, 2024 7:23 am

bbridge_GA wrote:
Sat May 25, 2024 5:58 pm
Here is the 2am apnea
In OSCAR, click on the Events tab and tell us how long that apnea is.

If you upload your data to a free account at SleepHQ.com and post the link in this thread, we can zoom in to take a close look at your data.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: High heartrate and capped upper pressure

Post by dataq1 » Sun May 26, 2024 9:42 am

Miss Emerita wrote:
Sat May 25, 2024 5:42 pm
I’d recommend that you ask your doctor for a Holter monitor to see what’s going on.
Agree, look particularly to rule out Proximal AFIB.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: High heartrate and capped upper pressure

Post by Rubicon » Sun May 26, 2024 1:53 pm

dataq1 wrote:
Sun May 26, 2024 9:42 am
Agree, look particularly to rule out Proximal AFIB.
I would like to nominate this post as the initial entry in the sticky "The Best of dataq1".
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

bbridge_GA
Posts: 50
Joined: Sun Sep 22, 2013 5:12 am

Re: High heartrate and capped upper pressure

Post by bbridge_GA » Sun May 26, 2024 2:39 pm

robysue1 wrote:
Sat May 25, 2024 8:18 pm
bbridge_GA wrote:
Sat May 25, 2024 5:58 pm
Here is the 2am apnea that corresponds to the heartrate being really high and the wave form earlier in the night zoomed in.
https://prnt.sc/eR7rcC6dy0Dm
https://prnt.sc/qiW9hGH-W9kf
Sure wish rubicon would take a look at that zoomed in area.

Also, could you zoom in further on just the area between 0:20 and 0:30---in other words, zoom in close enough where we can see the individual breaths.

And, do you remember being restless at all around this time? Do you remember any wakes during the night?

Finally, do you remember waking up to deal with any leaks during the night?
https://prnt.sc/IbNom3O4KVeW is zoomed in :20 to :30.
I don't remember or have fitbit sleep data on my waking on that night. No leaks during that :20-:30 time period. Generally, I don't move much at all during the night. I have leaks some nights, but usually not enough for the red leak flag to show up in the morning. I usually just re-seat my mask if it ever leaks. I have a mouth strap, but I have quit using it lately. I just ordered some mouth tape because i just got a mouthguard from the dentist.

I have a period of similar wave data more recently with the pressure cap set higher. https://prnt.sc/lZVQ8VPw3ouG

_________________
MachineMask
Additional Comments: changed from cpap to apap April 2014
Last edited by bbridge_GA on Sat Jun 01, 2024 7:15 pm, edited 4 times in total.
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15