High heartrate and capped upper pressure
-
bbridge_GA
- Posts: 50
- Joined: Sun Sep 22, 2013 5:12 am
High heartrate and capped upper pressure
I wear a Fitbit and like to use it to track my sleep quality and heartrate. My resting heartrate ranges from 53-80 and up to 145 max when exercising which I think is pretty healthy. My Fitbit has recorded 3 dates recently when my heartrate was between 160-180 while asleep. I didn't know if it was a malfunction of the Fitbit. I haven't looked at my CPAP data recently. This made me look at my data on Oscar to see what was going on. One of these events (heartrate of 180) correlated with a 30 second apnea that was preceded earlier by some strange waveforms earlier in the night. The others were not as clear of a relationship to a long apnea or anything unusual. I did notice that my pressure setting was capped at 10. My average pressure is typically around 6-7 with my 95% pressure around 9. I don't know why the cap was reduced to 10, but apparently there are times when I need more pressure. I have adjusted the top setting to 18 and plan to see if there are any more of these high hearrate events. I have also been having a lot of daytime sleepyness recently and hope that gets better with this adjustment. Any ideas about the funny waveforms? Here is the whole night.
https://prnt.sc/qiW9hGH-W9kf
https://prnt.sc/qiW9hGH-W9kf
Last edited by bbridge_GA on Sat May 25, 2024 5:54 pm, edited 2 times in total.
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15
Airsense 10 APAP
Wisp mask
Pressure 7-15
- ChicagoGranny
- Posts: 15527
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: High heartrate and capped upper pressure
How old are you?bbridge_GA wrote: ↑Sat May 25, 2024 3:07 pmup to 145 max when exercising which I think is pretty healthy.
Many of us, even those with low-pressure needs, set the max pressure to 20. The machine will not go any higher than it needs to.
Use imgur.com to host the pics. Or easier, get a free account at sleephq.com and post your data link in this thread.bbridge_GA wrote: ↑Sat May 25, 2024 3:07 pmAny ideas about the funny waveforms? I couldn't upload the pictures.
Data is critical for members trying to help you.
Re: High heartrate and capped upper pressure
Can you post pix of those wave forms?
- Miss Emerita
- Posts: 3783
- Joined: Sun Nov 04, 2018 8:07 pm
Re: High heartrate and capped upper pressure
I’d recommend that you ask your doctor for a Holter monitor to see what’s going on.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
bbridge_GA
- Posts: 50
- Joined: Sun Sep 22, 2013 5:12 am
Re: High heartrate and capped upper pressure
I added a link to the image of the whole night to the original post.
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15
Airsense 10 APAP
Wisp mask
Pressure 7-15
-
bbridge_GA
- Posts: 50
- Joined: Sun Sep 22, 2013 5:12 am
Re: High heartrate and capped upper pressure
Here is the 2am apnea that corresponds to the heartrate being really high and the wave form earlier in the night zoomed in.bbridge_GA wrote: ↑Sat May 25, 2024 5:52 pmI added a link to the image of the whole night to the original post.
https://prnt.sc/eR7rcC6dy0Dm
https://prnt.sc/qiW9hGH-W9kf
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15
Airsense 10 APAP
Wisp mask
Pressure 7-15
-
bbridge_GA
- Posts: 50
- Joined: Sun Sep 22, 2013 5:12 am
Re: High heartrate and capped upper pressure
I am 63 with a normal ekg and normal heart ultrasound. My ex was a holter tech and put one on me in 1985. He said my heart rate went crazy in the middle of the night. In hindsight, I have always wondered if that wasn't when I was first diagnosed with sleep apnea...via holter and 20 years before my formal diagnosis. A follow-up with a cardiologist showed a slight murmur... that no one has mentioned since then.Miss Emerita wrote: ↑Sat May 25, 2024 5:42 pmI’d recommend that you ask your doctor for a Holter monitor to see what’s going on.
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15
Airsense 10 APAP
Wisp mask
Pressure 7-15
-
bbridge_GA
- Posts: 50
- Joined: Sun Sep 22, 2013 5:12 am
Re: High heartrate and capped upper pressure
thanks for looking at these!bbridge_GA wrote: ↑Sat May 25, 2024 5:58 pmHere is the 2am apnea that corresponds to the heartrate being really high https://prnt.sc/daOkcjDZIUJm and the wave form earlier in the night zoomed in.bbridge_GA wrote: ↑Sat May 25, 2024 5:52 pmI added a link to the image of the whole night to the original post and here https://prnt.sc/qiW9hGH-W9kf
https://prnt.sc/eR7rcC6dy0Dm
Here is this past week with the upper limit raised. https://prnt.sc/v8S7U3XRwA0V
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15
Airsense 10 APAP
Wisp mask
Pressure 7-15
Re: High heartrate and capped upper pressure
Sure wish rubicon would take a look at that zoomed in area.bbridge_GA wrote: ↑Sat May 25, 2024 5:58 pmHere is the 2am apnea that corresponds to the heartrate being really high and the wave form earlier in the night zoomed in.
https://prnt.sc/eR7rcC6dy0Dm
https://prnt.sc/qiW9hGH-W9kf
Also, could you zoom in further on just the area between 0:20 and 0:30---in other words, zoom in close enough where we can see the individual breaths.
And, do you remember being restless at all around this time? Do you remember any wakes during the night?
Finally, do you remember waking up to deal with any leaks during the night?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Re: High heartrate and capped upper pressure
Lots of us have "slight murmurs" that are nothing to worry about in themselves without other accompanying problems that may or may not matter either.
- Miss Emerita
- Posts: 3783
- Joined: Sun Nov 04, 2018 8:07 pm
Re: High heartrate and capped upper pressure
1985? Time for another one. All kinds of reasons you might be having a high heart rate.bbridge_GA wrote: ↑Sat May 25, 2024 6:05 pmI am 63 with a normal ekg and normal heart ultrasound. My ex was a holter tech and put one on me in 1985. He said my heart rate went crazy in the middle of the night. In hindsight, I have always wondered if that wasn't when I was first diagnosed with sleep apnea...via holter and 20 years before my formal diagnosis. A follow-up with a cardiologist showed a slight murmur... that no one has mentioned since then.Miss Emerita wrote: ↑Sat May 25, 2024 5:42 pmI’d recommend that you ask your doctor for a Holter monitor to see what’s going on.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
- ChicagoGranny
- Posts: 15527
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: High heartrate and capped upper pressure
In OSCAR, click on the Events tab and tell us how long that apnea is.
If you upload your data to a free account at SleepHQ.com and post the link in this thread, we can zoom in to take a close look at your data.
Re: High heartrate and capped upper pressure
Agree, look particularly to rule out Proximal AFIB.Miss Emerita wrote: ↑Sat May 25, 2024 5:42 pmI’d recommend that you ask your doctor for a Holter monitor to see what’s going on.
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
Re: High heartrate and capped upper pressure
I would like to nominate this post as the initial entry in the sticky "The Best of dataq1".
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
-
bbridge_GA
- Posts: 50
- Joined: Sun Sep 22, 2013 5:12 am
Re: High heartrate and capped upper pressure
https://prnt.sc/IbNom3O4KVeW is zoomed in :20 to :30.robysue1 wrote: ↑Sat May 25, 2024 8:18 pmSure wish rubicon would take a look at that zoomed in area.bbridge_GA wrote: ↑Sat May 25, 2024 5:58 pmHere is the 2am apnea that corresponds to the heartrate being really high and the wave form earlier in the night zoomed in.
https://prnt.sc/eR7rcC6dy0Dm
https://prnt.sc/qiW9hGH-W9kf
Also, could you zoom in further on just the area between 0:20 and 0:30---in other words, zoom in close enough where we can see the individual breaths.
And, do you remember being restless at all around this time? Do you remember any wakes during the night?
Finally, do you remember waking up to deal with any leaks during the night?
I don't remember or have fitbit sleep data on my waking on that night. No leaks during that :20-:30 time period. Generally, I don't move much at all during the night. I have leaks some nights, but usually not enough for the red leak flag to show up in the morning. I usually just re-seat my mask if it ever leaks. I have a mouth strap, but I have quit using it lately. I just ordered some mouth tape because i just got a mouthguard from the dentist.
I have a period of similar wave data more recently with the pressure cap set higher. https://prnt.sc/lZVQ8VPw3ouG
Last edited by bbridge_GA on Sat Jun 01, 2024 7:15 pm, edited 4 times in total.
AHI 30 in 2004, UPP surgery in 2006 and 60 in 2012 with OA and CA
Airsense 10 APAP
Wisp mask
Pressure 7-15
Airsense 10 APAP
Wisp mask
Pressure 7-15