Resmed airsense 10 takes a break at end of breath

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
BlackBags
Posts: 2
Joined: Wed Feb 28, 2024 6:11 pm

Resmed airsense 10 takes a break at end of breath

Post by BlackBags » Wed Feb 28, 2024 6:40 pm

Hi everyone!

Trying to understand how stuff works... The device is supposed to keep a constant pressure (adjusting through the night but more or less constant during a breath) BUT we inhale, exhale, so is it true that it needs to adjust, otherwise the pressure would go up at every exhale?

Not sure what makes sense, but I do hear the motor working harder to keep up during inhales.

The second question, the airsense 10 seems to stop the pressure at the end of each exhale. In the sense that I turn it on, before breathing, pressure is there (can be felt and measured). Inhale, pressure is there. Exhale, pressure is there. Stop the exhale slowly, pressure drops to zero for a second or so before returning. If I try to inhale during that break I would have no pressure at all. EPR is turned off.

Is there a way to remove that break? It feels like a lag/delay between my breath and the device reacting.
Maybe some settings?

User avatar
Pugsy
Posts: 65114
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed airsense 10 takes a break at end of breath

Post by Pugsy » Wed Feb 28, 2024 7:07 pm

BlackBags wrote:
Wed Feb 28, 2024 6:40 pm
Stop the exhale slowly, pressure drops to zero for a second or so before returning. If I try to inhale during that break I would have no pressure at all. EPR is turned off.
Impossible. The machine can't EVER drop the pressure below 4 cm. Simply won't/can't do it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Respirator99
Posts: 389
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: Resmed airsense 10 takes a break at end of breath

Post by Respirator99 » Wed Feb 28, 2024 7:55 pm

G'day BlackBags, welcome to the forum.

It sounds like the EPR (expiratory pressure relief) isn't properly synching with your breathing. If you have EPR turned on, it reduces pressure slightly as you exhale. It then increases it again as you inhale. The available settings are 0 (off), 1, 2 or 3 - where the number corresponds to the number of cm reduction. As Pugsy said, the machine won't go below 4 cm, even if EPR is set to the maximum.

To overcome your problem, try turning EPR off. You'll find it takes a little more effort to exhale, but you may eliminate that synchronisation problem.
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

BlackBags
Posts: 2
Joined: Wed Feb 28, 2024 6:11 pm

Re: Resmed airsense 10 takes a break at end of breath

Post by BlackBags » Thu Feb 29, 2024 3:37 am

Thank you for your replies Pugsy and Respirator99!

Very strange... I did buy a calibrated manometer (pressure sensor) - mostly to check if the motor was still working properly, as the DME would have done (they switched from patient care and device maintenance to remote compliance checks only, so trying to learn how to do it myself, but that's a different story).

That sensor is measuring 0 pressure (in the mask) during that short time.

Will check EPR again.