F30 elbow

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Hettering80
Posts: 1
Joined: Sat Nov 04, 2023 7:29 am

F30 elbow

Post by Hettering80 » Sat Nov 04, 2023 8:16 am

I've just replaced the elbow but there seems to be a lot of air and noise coming from it, so I think it may be faulty. Previous one was quiet.
Question is when I do the mask test it says its fine and doesn't make a noise, but the minute the pressure drops to 4 its very very loud.

Is this normal? Hospital not open till Monday so any help would be appreciated

User avatar
ChicagoGranny
Posts: 15538
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: F30 elbow

Post by ChicagoGranny » Sat Nov 04, 2023 10:16 am

Hettering80 wrote: ↑
Sat Nov 04, 2023 8:16 am
Is this normal?
No.

This is hard to diagnosis without seeing the mask.

What are your pressure settings?

When does the pressure run at 4.0 cm?

Most people cannot tolerate a pressure that low. You might raise the pressure until you can have your supplier look at the elbow.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: F30 elbow

Post by dataq1 » Sat Nov 04, 2023 11:06 am

Hettering80 wrote: ↑
Sat Nov 04, 2023 8:16 am
I've just replaced the elbow but there seems to be a lot of air and noise coming from it, so I think it may be faulty. Previous one was quiet.
Can we assume that you replaced the F30 QuietAir elbow because it was broken or because it didn't latch securely?
The Quietair elbow should be quiet at all pressures when properly seated in the frame.

The intentional leak (vent in the elbow) is 19 l/m @ 3cmH20, barely observable.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
ChicagoGranny
Posts: 15538
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: F30 elbow

Post by ChicagoGranny » Sat Nov 04, 2023 4:52 pm

dataq1 wrote: ↑
Sat Nov 04, 2023 11:06 am
The intentional leak (vent in the elbow) is 19 l/m @ 3cmH20
No, that's the total passive vent rate including the vent in the mask itself.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: F30 elbow

Post by dataq1 » Sat Nov 04, 2023 9:32 pm

ChicagoGranny wrote: ↑
Sat Nov 04, 2023 4:52 pm
dataq1 wrote: ↑
Sat Nov 04, 2023 11:06 am
The intentional leak (vent in the elbow) is 19 l/m @ 3cmH20
No, that's the total passive vent rate including the vent in the mask itself.
Everything is vented in the elbow, the air passing thru the slots in the cushion support go into the elbow. As far as I see there are no other "vents" except in the elbow.

But the real point is: Why would the "elbow noise" be "very very loud" at 4 cm pressure but with higher pressures much quieter?
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
BlueDragon
Posts: 572
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: F30 elbow

Post by BlueDragon » Sat Nov 04, 2023 10:41 pm

The elbow has a clip on each side that attach to the mask. Are both sides clipped in properly? If you have just one clip seated and the other one is loose, you will get a lot of noise.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

Garethep
Posts: 14
Joined: Sat Jul 29, 2023 8:18 am

Re: F30 elbow

Post by Garethep » Sun Nov 05, 2023 3:46 am

I'm not familiar with this mask but I assume that, like other full face masks, it has a non return valve so you can breathe if the machine shuts down.....it's possible that this is leaking at the lower pressure?

_________________
Machine

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Re: F30 elbow

Post by ozij » Sun Nov 05, 2023 7:34 am

Garethep wrote: ↑
Sun Nov 05, 2023 3:46 am
I'm not familiar with this mask but I assume that, like other full face masks, it has a non return valve so you can breathe if the machine shuts down.....it's possible that this is leaking at the lower pressure?
Correct.

The F30 Elbow has two flaps on the sides, they drop open when there's no pressure, and pressurize air flaps them shut.

On the very rare occasions the flap don't shut, I find that blowing out firmly shuts them.

So there are two possible explanations:
The elbow itself isn't fixed correctly to the mask -
The anti-asphyxia valve remain open when the pressure is too low.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: F30 elbow

Post by dataq1 » Sun Nov 05, 2023 7:36 am

Garethep wrote: ↑
Sun Nov 05, 2023 3:46 am
I'm not familiar with this mask but I assume that, like other full face masks, it has a non return valve so you can breathe if the machine shuts down.....it's possible that this is leaking at the lower pressure?
The valve that you reference is the anti-asphyxiation valve. It relies on flow, not pressure, to close off the external ports. But what you said otherwise, that a mal-function of either of these two flaps that causes the causes the external ports to not seal would result in blower flow to leak out these ports.
Regardless, the elbow is either not seated and secured properly or the elbow is defective.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: F30 elbow

Post by Okie bipap » Sun Nov 05, 2023 1:56 pm

Did you replace the elbow with one of the older type of elbows? These elbows vented directly without the baffle over the vent holes. They were sometimes a little noisy if something was in front of the vent holes.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

si17
Posts: 1
Joined: Thu Feb 15, 2024 6:04 am

Re: F30 elbow

Post by si17 » Thu Feb 15, 2024 6:07 am

Hettering80 wrote: ↑
Sat Nov 04, 2023 8:16 am
I've just replaced the elbow but there seems to be a lot of air and noise coming from it, so I think it may be faulty. Previous one was quiet.
Question is when I do the mask test it says its fine and doesn't make a noise, but the minute the pressure drops to 4 its very very loud.

Is this normal? Hospital not open till Monday so any help would be appreciated
Hi,

Did you ever fix this issue? I’m having the exact same problem with a new F20 mask.

User avatar
zonker
Posts: 11499
Joined: Fri Jun 19, 2015 4:36 pm

Re: F30 elbow

Post by zonker » Thu Feb 15, 2024 12:28 pm

si17 wrote: ↑
Thu Feb 15, 2024 6:07 am
Hettering80 wrote: ↑
Sat Nov 04, 2023 8:16 am
I've just replaced the elbow but there seems to be a lot of air and noise coming from it, so I think it may be faulty. Previous one was quiet.
Question is when I do the mask test it says its fine and doesn't make a noise, but the minute the pressure drops to 4 its very very loud.

Is this normal? Hospital not open till Monday so any help would be appreciated
Hi,

Did you ever fix this issue? I’m having the exact same problem with a new F20 mask.
hettering80 hasn't been in since nov 25th 2023, so unlikely s/he will see your question.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg