F30 elbow
-
Hettering80
- Posts: 1
- Joined: Sat Nov 04, 2023 7:29 am
F30 elbow
I've just replaced the elbow but there seems to be a lot of air and noise coming from it, so I think it may be faulty. Previous one was quiet.
Question is when I do the mask test it says its fine and doesn't make a noise, but the minute the pressure drops to 4 its very very loud.
Is this normal? Hospital not open till Monday so any help would be appreciated
Question is when I do the mask test it says its fine and doesn't make a noise, but the minute the pressure drops to 4 its very very loud.
Is this normal? Hospital not open till Monday so any help would be appreciated
- ChicagoGranny
- Posts: 15541
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: F30 elbow
No.
This is hard to diagnosis without seeing the mask.
What are your pressure settings?
When does the pressure run at 4.0 cm?
Most people cannot tolerate a pressure that low. You might raise the pressure until you can have your supplier look at the elbow.
Re: F30 elbow
Can we assume that you replaced the F30 QuietAir elbow because it was broken or because it didn't latch securely?Hettering80 wrote: ↑Sat Nov 04, 2023 8:16 amI've just replaced the elbow but there seems to be a lot of air and noise coming from it, so I think it may be faulty. Previous one was quiet.
The Quietair elbow should be quiet at all pressures when properly seated in the frame.
The intentional leak (vent in the elbow) is 19 l/m @ 3cmH20, barely observable.
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
- ChicagoGranny
- Posts: 15541
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: F30 elbow
Everything is vented in the elbow, the air passing thru the slots in the cushion support go into the elbow. As far as I see there are no other "vents" except in the elbow.ChicagoGranny wrote: ↑Sat Nov 04, 2023 4:52 pmNo, that's the total passive vent rate including the vent in the mask itself.
But the real point is: Why would the "elbow noise" be "very very loud" at 4 cm pressure but with higher pressures much quieter?
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
- BlueDragon
- Posts: 575
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: F30 elbow
The elbow has a clip on each side that attach to the mask. Are both sides clipped in properly? If you have just one clip seated and the other one is loose, you will get a lot of noise.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
Re: F30 elbow
I'm not familiar with this mask but I assume that, like other full face masks, it has a non return valve so you can breathe if the machine shuts down.....it's possible that this is leaking at the lower pressure?
Re: F30 elbow
Correct.
The F30 Elbow has two flaps on the sides, they drop open when there's no pressure, and pressurize air flaps them shut.
On the very rare occasions the flap don't shut, I find that blowing out firmly shuts them.
So there are two possible explanations:
The elbow itself isn't fixed correctly to the mask -
The anti-asphyxia valve remain open when the pressure is too low.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: F30 elbow
The valve that you reference is the anti-asphyxiation valve. It relies on flow, not pressure, to close off the external ports. But what you said otherwise, that a mal-function of either of these two flaps that causes the causes the external ports to not seal would result in blower flow to leak out these ports.
Regardless, the elbow is either not seated and secured properly or the elbow is defective.
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: F30 elbow
Did you replace the elbow with one of the older type of elbows? These elbows vented directly without the baffle over the vent holes. They were sometimes a little noisy if something was in front of the vent holes.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Evora Full Face Mask - Fitpack |
| Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: F30 elbow
Hi,Hettering80 wrote: ↑Sat Nov 04, 2023 8:16 amI've just replaced the elbow but there seems to be a lot of air and noise coming from it, so I think it may be faulty. Previous one was quiet.
Question is when I do the mask test it says its fine and doesn't make a noise, but the minute the pressure drops to 4 its very very loud.
Is this normal? Hospital not open till Monday so any help would be appreciated
Did you ever fix this issue? I’m having the exact same problem with a new F20 mask.
Re: F30 elbow
hettering80 hasn't been in since nov 25th 2023, so unlikely s/he will see your question.si17 wrote: ↑Thu Feb 15, 2024 6:07 amHi,Hettering80 wrote: ↑Sat Nov 04, 2023 8:16 amI've just replaced the elbow but there seems to be a lot of air and noise coming from it, so I think it may be faulty. Previous one was quiet.
Question is when I do the mask test it says its fine and doesn't make a noise, but the minute the pressure drops to 4 its very very loud.
Is this normal? Hospital not open till Monday so any help would be appreciated
Did you ever fix this issue? I’m having the exact same problem with a new F20 mask.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg