New to CPAP – Requesting Some Guidance

Post by Sleepy864 » Sun Aug 27, 2023 2:31 pm

Hi all! I'm hoping to get some guidance from you to improve my treatment with CPAP.

I am a 30 year old male with a normal BMI that has been extremely tired for over a decade. I had a sleep study back in 2016, and though the doctor didn’t mention it, the report shows a potential UARS diagnosis (RDI > 5).

As I kept getting more tired with each passing year, I decided to try another sleep study using a WatchPAT ONE in 2023. My diagnosis came back as mild OSA based on a pRDI of 9.4 and O2 nadir of 91%. My overall AHI was 3.4. Both of these are a little bit higher than my sleep study in 2016.

I was prescribed CPAP therapy and received a ResMed AirSense 11 Autoset a little while back. It arrived with a minimum pressure of 4 cmH20, maximum of 20 cmH20, and EPR of 1. It is hard for me to see much of a difference yet, though I see some encouraging signs starting to emerge (like not feeling terrible following a nap after work).

I recently put an SD card in my CPAP machine and was hoping to get some recommendations based on my OSCAR data. I’ve been reading a lot of threads on this forum, and am not really sure which category I fall into (mild OSA, UARS, something else) and what adjustments I should make to improve my fatigue.

I attached my chart output over from OSCAR below from a recent night in addition to a zoomed in view that seemed to be a calmer time and one where a clear airway was detected (I also uploaded the screenshots at https://imgur.com/a/M40r2Gy. This seems to be one of the “smoother” looking nights. It has been normal over the past year for me to get up a couple of times per night (and it started prior to CPAP therapy), though I didn’t always wake up quite as much during the night.

Are there any setting adjustments you might recommend going forward? The only adjustment I have made thus far is adjusting the EPR to 3 a couple of nights ago. I can’t tell whether that made much of a difference yet.

Thank you!
Attachments
Sleep_Data_08_23_Zoomed_in_2.png
Sleep_Data_08_23_Zoomed_in_2.png (106 KiB) Viewed 1142 times
Sleep_Data_08_23_Zoomed_in_1.png
Sleep_Data_08_23_Zoomed_in_1.png (104.29 KiB) Viewed 1142 times
Sleep_Data_08_23.png
Sleep_Data_08_23.png (115.91 KiB) Viewed 1142 times

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: New to CPAP – Requesting Some Guidance

Post by Julie » Sun Aug 27, 2023 4:58 pm

Hi, not bad for a newbie, but I'd raise your min. pressure to e.g. 6 or 7 as 4, the default low, is hard to breathe at.

Is there a reason for the EPR? It actually makes your min. setting even lower and unless you were told to use it for a good reason, I wouldn't bother, or at least use e.g. 1 or 2, rather than 3.

And having "a" clear airway means zip, and having even a handful at bedtime or before waking up isn't a big deal, only if there are lots throughout the chart would it be worth looking into.

User avatar
LSAT
Posts: 13374
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: New to CPAP – Requesting Some Guidance

Post by LSAT » Sun Aug 27, 2023 5:02 pm

What was your AHI when you had your sleep study.

Sleepy864
Posts: 2
Joined: Sun Aug 27, 2023 2:27 pm

Re: New to CPAP – Requesting Some Guidance

Post by Sleepy864 » Sun Aug 27, 2023 7:21 pm

Thank you for your replies! I've adjusted the minimum pressure to 6 and set the EPR to 1. I'll report back on how these differences go.

In the sleep study, my AHI was 3.4 and my RDI was 9.4.