High respiratory rate and weird breathing
-
- Posts: 16
- Joined: Fri Jul 14, 2023 7:47 am
High respiratory rate and weird breathing
Imgur link with OSCAR data screenshot: https://imgur.com/a/6jEujGQ
Hi,
I keep seeing these weird breaths in my OSCAR reports. Someone on reddit suggested cardiogenic oscillations. Are they something to worry about? They happen pretty much every single night in episodes.
I'm absolutely going to talk to the sleep specialist about it in october when the appointment is booked. Just wanted to ask about it here until then, it worries me a bit.
And also, my respiratory rate seems a bit high right? Or is it normal?
Sorry if this is written a bit messy.
Hi,
I keep seeing these weird breaths in my OSCAR reports. Someone on reddit suggested cardiogenic oscillations. Are they something to worry about? They happen pretty much every single night in episodes.
I'm absolutely going to talk to the sleep specialist about it in october when the appointment is booked. Just wanted to ask about it here until then, it worries me a bit.
And also, my respiratory rate seems a bit high right? Or is it normal?
Sorry if this is written a bit messy.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I don't use the humudifier. |
- ChicagoGranny
- Posts: 15085
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: High respiratory rate and weird breathing
Your median respiratory rate is 13.6. This is normal.
If you want to post the full-night RR line, we could possibly make some more comments. In the meantime, there is nothing that should alarm you about the RR.
If you want to post the full-night RR line, we could possibly make some more comments. In the meantime, there is nothing that should alarm you about the RR.
Re: High respiratory rate and weird breathing
Why don't you post a couple of Oscar reports and circle whatever seems weird, then experts here can see them and comment.
-
- Posts: 16
- Joined: Fri Jul 14, 2023 7:47 am
Re: High respiratory rate and weird breathing
Thank you. It's good to hear that the breathing rate is okay. I still don't quite get what the 95% mean. Does it mean that 95% of the time the breathing rate is at that, or below that value?ChicagoGranny wrote: ↑Sun Aug 20, 2023 3:37 pmYour median respiratory rate is 13.6. This is normal.
If you want to post the full-night RR line, we could possibly make some more comments. In the meantime, there is nothing that should alarm you about the RR.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I don't use the humudifier. |
-
- Posts: 16
- Joined: Fri Jul 14, 2023 7:47 am
Re: High respiratory rate and weird breathing
I honestly think I'm going to let it be as it is until the appointment. I'm getting a bit too stressed out about the data.
The breaths just looks a bit jagged, like waves on top of each inhalation.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I don't use the humudifier. |
- ChicagoGranny
- Posts: 15085
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: High respiratory rate and weird breathing
Yes. If you woke up and went to the restroom, your breathing rate is going to be in the other 5%. Same if you rolled around in bed a bit.SlothInAMoth wrote: ↑Mon Aug 21, 2023 2:19 pmDoes it mean that 95% of the time the breathing rate is at that, or below that value?
Re: High respiratory rate and weird breathing
95 % numbers of anything (pressure, leak, respiration rate, whatever) just means you were AT OR BELOW that number for 95% of the night.
The OR BELOW part is really important.
It does NOT mean your number means you were at that rate for 95% of the night.
The median average number is more indicative of how things went for the bulk of the night.
95% numbers of anything are too easily skewed to the high side by just a small amount at the higher rate to be all that important in the grand scheme of things.
While I haven't had much of a higher respiration rate I have had 95% leak numbers that make things out a whole lot worse than they are.
I have examples as to leak numbers in this thread and you can see what I mean.
viewtopic.php?f=1&t=108724&p=1036669&hi ... t#p1036669
The OR BELOW part is really important.
It does NOT mean your number means you were at that rate for 95% of the night.
The median average number is more indicative of how things went for the bulk of the night.
95% numbers of anything are too easily skewed to the high side by just a small amount at the higher rate to be all that important in the grand scheme of things.
While I haven't had much of a higher respiration rate I have had 95% leak numbers that make things out a whole lot worse than they are.
I have examples as to leak numbers in this thread and you can see what I mean.
viewtopic.php?f=1&t=108724&p=1036669&hi ... t#p1036669
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: High respiratory rate and weird breathing
http://adventures-in-hosehead-land.blog ... de-to.htmlSlothInAMoth wrote: ↑Mon Aug 21, 2023 2:19 pmI still don't quite get what the 95% mean. Does it mean that 95% of the time the breathing rate is at that, or below that value?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Miss Emerita
- Posts: 3732
- Joined: Sun Nov 04, 2018 8:07 pm
Re: High respiratory rate and weird breathing
I am puzzled by the flow rate in the snippet you posted, and I'm hoping some experts will take a look. I do see what appears to be cardioballistic (cardiogenic) effects (artifacts) in the small oscillations around the zero line. But I'm skeptical whether the "double inhalations" reflect your heartbeat. They might reflect flow limitations, though they're quite exaggerated if that's what they are. How much of the night do they go on?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: High respiratory rate and weird breathing
I think you need to load up the entire night to SleepHQ to get a better idea of the source of the oscillations.
However, IMO that's palatal interference, based on how the inspiratory limb gets pretty much destroyed (airway FL) and anyway CBA is usually much less (if at all) apparent on inspiration.

That said, the oscillations occur at a rate of ~78/minute-- about the same as a resting heart rate (of someone who should be doing more aerobic exercise) so there could be a little CBA in there.
However, IMO that's palatal interference, based on how the inspiratory limb gets pretty much destroyed (airway FL) and anyway CBA is usually much less (if at all) apparent on inspiration.

That said, the oscillations occur at a rate of ~78/minute-- about the same as a resting heart rate (of someone who should be doing more aerobic exercise) so there could be a little CBA in there.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
-
- Posts: 16
- Joined: Fri Jul 14, 2023 7:47 am
Re: High respiratory rate and weird breathing
Thank you, you explained it really well. I get it now!
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I don't use the humudifier. |
-
- Posts: 16
- Joined: Fri Jul 14, 2023 7:47 am
Re: High respiratory rate and weird breathing
I'm a bit confused by them too. My average heart rate during the night seem to be quite high. Between 77-78 bpm. I'm a tiny lady, work out quite a few days a week, have a bmi at the lower end of normal. So no weight-related breathing blockage.Miss Emerita wrote: ↑Tue Aug 22, 2023 10:34 amI am puzzled by the flow rate in the snippet you posted, and I'm hoping some experts will take a look. I do see what appears to be cardioballistic (cardiogenic) effects (artifacts) in the small oscillations around the zero line. But I'm skeptical whether the "double inhalations" reflect your heartbeat. They might reflect flow limitations, though they're quite exaggerated if that's what they are. How much of the night do they go on?
Anxiety is something I've struggled with for years. Maybe that could be the reason for a high bpm?
I'll see if I can post a better overview of that night later.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I don't use the humudifier. |
-
- Posts: 16
- Joined: Fri Jul 14, 2023 7:47 am
Re: High respiratory rate and weird breathing
Is sleepHQ easier to interpret than OSCAR?Rubicon wrote: ↑Tue Aug 22, 2023 11:17 amI think you need to load up the entire night to SleepHQ to get a better idea of the source of the oscillations.
However, IMO that's palatal interference, based on how the inspiratory limb gets pretty much destroyed (airway FL) and anyway CBA is usually much less (if at all) apparent on inspiration.
That said, the oscillations occur at a rate of ~78/minute-- about the same as a resting heart rate (of someone who should be doing more aerobic exercise) so there could be a little CBA in there.
I'll go ahead and post a screenshot with a better overview later.
77-78 bpm seem to be my average heart rate during the night. Might have to do with anxiety which I have.
I'm actually really physically active, do running and weights and my bmi is at the lower end of normal. So no weight-related sleep issues.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I don't use the humudifier. |
Re: High respiratory rate and weird breathing
from what i can gather (and i may be wrong) the advantage of sleephq is that after you post it, someone else looking at the chart can zoom in to see certain aspects. with oscar, you are the only one that can zoom in.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 16
- Joined: Fri Jul 14, 2023 7:47 am
Re: High respiratory rate and weird breathing
That's great! I wish OSCAR had the same function. I'll check out SleepHQ and see if I can post it here.
Here is a link to one night:
https://sleephq.com/public/b80ef2fd-e60 ... 8d97b07c55
I had to click "show advanced charts" in order to view the data!
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Additional Comments: I don't use the humudifier. |