New to CPAP and struggling (sort of)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
2young2bawake
Posts: 16
Joined: Sat Aug 05, 2023 8:32 am
Location: San Marcos Texas

New to CPAP and struggling (sort of)

Post by 2young2bawake » Sat Aug 05, 2023 9:35 am

I was diagnosed with sleep apnea as a teenager. Couldn't sleep on the CPAP then and quit. I had the Uvulopalatopharyngoplasty surgery and never went back to see if it was fixed.
Currently, 45 yoa, 205 lbs, 5'11" very active with 41 VO2max. Borderline hypertensive and borderline hyperlipidemia.
Tired of being tired......
Dr checkup says healthy and numbers are good (blood test)
I got a Fenix 6 pro health tracking watch. I noticed my sleeping O2 stats were in the 80's. Noticed I would wake up 8 times a night to pee and it was when my O2 stats were low (under 90).
At home sleep test says AHI of 12 and O2 in 80's. Same data the watch was recording so watch is close enough even thou it is not medical device.
Prescribed CPAP, Resmed Airsense 11 auto set with full face mask. pressures set auto from 6-12.
CPAP does not have SD card and transmits with cellular data.
I wake up multiple times a night and feel like I'm suffocating and have to remove the CPAP mask. Watch shows O2 at 84 and such.
The CPAP shows average 6-12 events per hour. I spoke with DME and they said my numbers look good????
Waking up feeling like I am suffocating takes me time to calm down and making me even more tired. The opposite of what I was hoping.
At one month of trying to adjust to CPAP. I am a mouth breather (nose with deviated septum) and can drool at night.
It feels like I am trying to suck air thru a wet paper straw.
I am so tired now that am about to just say throw it all away. Not to mention there is this compliance thing with insurance that I have to show its use or I get to pay for this device.
Sorry I did try to read the new user and apologize if I didn't follow all the rules. And if Pugsy asks three questions, I will try to answer all three.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Integrated heated humidifier with heated tube

User avatar
Pugsy
Posts: 65412
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New to CPAP and struggling (sort of)

Post by Pugsy » Sat Aug 05, 2023 9:53 am

Go buy an SD card and use it.
Get the OSCAR software program and use it.
2young2bawake wrote: ↑
Sat Aug 05, 2023 9:35 am
The CPAP shows average 6-12 events per hour. I spoke with DME and they said my numbers look good????
I disagree....those are horrible numbers but before we can offer ideas on improvement we have to know what that AHI is composed of....that's where the software come in to play.

Only a couple of questions at this time.

How many hours of sleep (not just laying in bed) are you getting and are those hours fragmented with a lot of wake ups that you remember?

Do you take any medications at this time (even OTC) and if you do...what?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65412
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New to CPAP and struggling (sort of)

Post by Pugsy » Sat Aug 05, 2023 9:56 am

BTW any generic SD card will work (and won't mess with the cellular transmission of data).

Any SD card....doesn't have to be a huge capacity SD Card but it's getting hard to find those smaller size cards.
Last I looked at walmart the 16 GB was the smallest available.
No need for one of those large capacity fancy SD cards....save your money.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

2young2bawake
Posts: 16
Joined: Sat Aug 05, 2023 8:32 am
Location: San Marcos Texas

Re: New to CPAP and struggling (sort of)

Post by 2young2bawake » Sat Aug 05, 2023 10:13 am

So about 7 hours in bed. I can remember waking up all night long. Sometimes 4 times and others 8 plus.

OTC ... vitamin D3, vitamin K2, fish oil, b complex, sometimes zyrtec
Prescription - Dilantin 400mgs

I will pick up an SD card.

Is there any device that would pair with the CPAP that would show O2, so my Dr would be able to evaluate the therapy?
I have not had a titration study just the in home test.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Integrated heated humidifier with heated tube

User avatar
Pugsy
Posts: 65412
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New to CPAP and struggling (sort of)

Post by Pugsy » Sat Aug 05, 2023 10:24 am

2young2bawake wrote: ↑
Sat Aug 05, 2023 10:13 am
Is there any device that would pair with the CPAP that would show O2
Yep there is but very, very pricey at about a grand.

Most forum members just use one of those overnight recording pulse ox devices. Compatible with OSCAR but not tied to the cpap machine at all. Priced between 50 and 120 bucks depending on how fast you want it and what features you might want.

OSCAR shows a list of compatible devices in its help section.

Frequent wake ups for any reason will seriously impact how you feel with or without adequate/optimal therapy.
At this point you therapy doesn't appear to be adequate or optimal...so lets start there and see what can be done or needs to be done.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: New to CPAP and struggling (sort of)

Post by Julie » Sat Aug 05, 2023 11:40 am

Does your MD have anything to say about using Dilantin with Cpap? I checked it out and it doesn't look directly problematic, but there were a couple of things mentioned which you're probably aware of and I know you probably have no choice re taking it.

2young2bawake
Posts: 16
Joined: Sat Aug 05, 2023 8:32 am
Location: San Marcos Texas

Re: New to CPAP and struggling (sort of)

Post by 2young2bawake » Sat Aug 05, 2023 1:16 pm

Julie wrote: ↑
Sat Aug 05, 2023 11:40 am
Does your MD have anything to say about using Dilantin with Cpap? I checked it out and it doesn't look directly problematic, but there were a couple of things mentioned which you're probably aware of and I know you probably have no choice re taking it.
I have been on Dilantin since I was 16... On the medication I don't have seizures. I don't think the Dr's would chance changing meds when seizures are controlled.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Integrated heated humidifier with heated tube

2young2bawake
Posts: 16
Joined: Sat Aug 05, 2023 8:32 am
Location: San Marcos Texas

Re: New to CPAP and struggling (sort of)

Post by 2young2bawake » Sat Aug 05, 2023 5:29 pm

https://imgur.com/a/UIjcsL0

from oscar of the last month

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Integrated heated humidifier with heated tube

User avatar
zonker
Posts: 11494
Joined: Fri Jun 19, 2015 4:36 pm

Re: New to CPAP and struggling (sort of)

Post by zonker » Sat Aug 05, 2023 5:42 pm

2young2bawake wrote: ↑
Sat Aug 05, 2023 5:29 pm
https://imgur.com/a/UIjcsL0

from oscar of the last month
that's not useful as it doesn't show what makes up the ahi. and the smart folk here need to see that breakdown in order to help.

please read here-
viewtopic.php?t=158560
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

2young2bawake
Posts: 16
Joined: Sat Aug 05, 2023 8:32 am
Location: San Marcos Texas

Re: New to CPAP and struggling (sort of)

Post by 2young2bawake » Sat Aug 05, 2023 6:15 pm

zonker wrote: ↑
Sat Aug 05, 2023 5:42 pm
that's not useful as it doesn't show what makes up the ahi. and the smart folk here need to see that breakdown in order to help.
I don't see any of the data like that. I even tried to use sleepyhead but it did not support the Resmed air sense 11.

So on 8/4/23 ... AHI 6.10, H 2.40, OA .40, CA 3.4

8/3/23 ... AHI 12.40, UA 2.10, H 1, OA 1.O CA 7.3

8/2/23 ... AHI 5.80, UA .2, H .4 , OA .4, CA 3.9

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Integrated heated humidifier with heated tube
Last edited by 2young2bawake on Sat Aug 05, 2023 6:47 pm, edited 1 time in total.

User avatar
Pugsy
Posts: 65412
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New to CPAP and struggling (sort of)

Post by Pugsy » Sat Aug 05, 2023 6:37 pm

The SD card MUST be in the machine for someone to see/get their detailed data.
Right now your old data is just the summary data that is stored on the machine itself.
Any of the good detailed stuff has to be written directly to the SD card while in use.
So that's why you don't have any detailed data or more useful graphs....nothing to see....if the machine doesn't write to the SD card all you get are summary numbers.
Doesn't matter what software you use....if the machine doesn't put the data on the card then the software can't make it up.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

2young2bawake
Posts: 16
Joined: Sat Aug 05, 2023 8:32 am
Location: San Marcos Texas

Re: New to CPAP and struggling (sort of)

Post by 2young2bawake » Sat Aug 05, 2023 6:49 pm

OK, that was my mistake, thank you. It will be in the machine tonight and from now on. :D

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Integrated heated humidifier with heated tube

User avatar
zonker
Posts: 11494
Joined: Fri Jun 19, 2015 4:36 pm

Re: New to CPAP and struggling (sort of)

Post by zonker » Sat Aug 05, 2023 9:57 pm

2young2bawake wrote: ↑
Sat Aug 05, 2023 6:49 pm
OK, that was my mistake, thank you. It will be in the machine tonight and from now on. :D
Image
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Re: New to CPAP and struggling (sort of)

Post by ozij » Sun Aug 06, 2023 7:57 pm

zonker wrote: ↑
Sat Aug 05, 2023 5:42 pm
2young2bawake wrote: ↑
Sat Aug 05, 2023 5:29 pm
https://imgur.com/a/UIjcsL0

from oscar of the last month
that's not useful as it doesn't show what makes up the ahi. and the smart folk here need to see that breakdown in order to help.

please read here-
viewtopic.php?t=158560
Of course it shows the breakdown of the AHI - but that breakdown isn't informative enough. Because we need to know when and at which pressure each type of event happens.
What we know is that you still have obstructive events every night, that you have Clear Airvway breathing stops every night - the reason for which we don't know at this point, and that you seem to be hitting your maximum pressure every night, and for long periods. You can check each of my statements by placing the curser on the relevant colored column.

No wonder you feel bad.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

2young2bawake
Posts: 16
Joined: Sat Aug 05, 2023 8:32 am
Location: San Marcos Texas

Re: New to CPAP and struggling (sort of)

Post by 2young2bawake » Mon Aug 07, 2023 5:34 pm

https://imgur.com/a/zwspG93

Data from OSCAR

I hope that works this time

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear
Additional Comments: Integrated heated humidifier with heated tube