Settings Suggestion

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Cedric35
Posts: 4
Joined: Wed Jul 12, 2023 8:31 am

Settings Suggestion

Post by Cedric35 » Wed Jul 12, 2023 9:18 am

]I'm an experienced CPAP user, but recently got a new machine and my AHI has gone from being controlled (< 5; my definition ) to over 5 most nights. I'm using the same mask as before so only the machine changed.

I've attached a summary report from OSCAR to show my history. My equipment profile info is up to date.

I'd appreciate any suggestions on settings.

Thanks for a great resource!

John

Image

User avatar
zonker
Posts: 11339
Joined: Fri Jun 19, 2015 4:36 pm

Re: Settings Suggestion

Post by zonker » Wed Jul 12, 2023 10:55 am

Cedric35 wrote:
Wed Jul 12, 2023 9:18 am
]I'm an experienced CPAP user, but recently got a new machine and my AHI has gone from being controlled (< 5; my definition ) to over 5 most nights. I'm using the same mask as before so only the machine changed.

I've attached a summary report from OSCAR to show my history. My equipment profile info is up to date.

I'd appreciate any suggestions on settings.

Thanks for a great resource!

John

Image
welcome to the zoo!

please read the following so you know what the forum expects to see in the way of oscar data-

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

there's none of the data needed to judge your situation in your image.

good luck!

btw, to those who see only an image icon, click it to open in a new tab. it takes you to imgur.com and shows the image.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ChicagoGranny
Posts: 15327
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Settings Suggestion

Post by ChicagoGranny » Wed Jul 12, 2023 2:38 pm

Cedric35 wrote:
Wed Jul 12, 2023 9:18 am
so only the machine changed
You switched from a ResMed machine to a Philips machine. The ResMed algorithm gets higher ratings for effectiveness. If you will follow zonker's instructions for posting charts, members can help you get your Philips machine set up for better results.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Settings Suggestion

Post by palerider » Wed Jul 12, 2023 4:36 pm

Cedric35 wrote:
Wed Jul 12, 2023 9:18 am
]I'm an experienced CPAP user, but recently got a new machine and my AHI has gone from being controlled (< 5; my definition )
That's not actually an acceptable level, unless you're a doctor/insurance company

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Cedric35
Posts: 4
Joined: Wed Jul 12, 2023 8:31 am

Re: Settings Suggestion

Post by Cedric35 » Thu Jul 13, 2023 8:34 am

Here's the daily view:

Image

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Settings Suggestion

Post by Julie » Thu Jul 13, 2023 10:47 am

Something isn't working - nothing's been posted of any use... look again at the Oscar info to see how to do it and look at others' posts to see what's been shown and how.

User avatar
zonker
Posts: 11339
Joined: Fri Jun 19, 2015 4:36 pm

Re: Settings Suggestion

Post by zonker » Thu Jul 13, 2023 10:54 am

Julie wrote:
Thu Jul 13, 2023 10:47 am
Something isn't working - nothing's been posted of any use... look again at the Oscar info to see how to do it and look at others' posts to see what's been shown and how.
it's there. i admit it's not exactly smooth, but it's there.

you have to open the image icon in a new tab and it'll load the imgur.com hosted image.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
zonker
Posts: 11339
Joined: Fri Jun 19, 2015 4:36 pm

Re: Settings Suggestion

Post by zonker » Thu Jul 13, 2023 11:03 am

Cedric35 wrote:
Thu Jul 13, 2023 8:34 am
Here's the daily view:

Image
i'd suggest raising your minimum pressure to 9. if this is too uncomfortable to do at one fell swoop, i'd suggest raising it slightly until you get used to it.

see in your pressure chart how you climb up from your min of 7 and then never really get back to 7? looks like your machine wants to get there, but it's a slow process. try it and see what happens.

also, you're going to want to work on those leaks.

others, with more knowledgeable suggestions, will be along soon.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65122
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Settings Suggestion

Post by Pugsy » Thu Jul 13, 2023 11:12 am


_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Miss Emerita
Posts: 3757
Joined: Sun Nov 04, 2018 8:07 pm

Re: Settings Suggestion

Post by Miss Emerita » Thu Jul 13, 2023 11:44 am

I agree with Zonker that you need to raise your minimum. PR machines are by design pretty conservative; they don't raise the pressure much after an event. With a higher minimum, you may be able to head off more obstructive apneas and hypopneas before they have a chance to happen.

Your median pressure is already 9, so you may barely notice the change. But if you do, and it bothers you, as Zonker says you can raise your minimum by increments (1 or even .5).

I wouldn't be surprised if you'd benefit from an even higher minimum, but let's see some data before trying to figure that out.

About your leaks: they're actually OK. PR machines reports include intentional leaks (venting). On the left panel, your data indicate that you didn't have any "large leaks," i.e., leaks your machine couldn't compensate for. Of course, if you have leaks that are waking you up, you do need to deal with those.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

Cedric35
Posts: 4
Joined: Wed Jul 12, 2023 8:31 am

Re: Settings Suggestion

Post by Cedric35 » Thu Jul 13, 2023 12:11 pm

Thanks for the suggestions. I’ll give a minimum of 9 a try.

User avatar
zonker
Posts: 11339
Joined: Fri Jun 19, 2015 4:36 pm

Re: Settings Suggestion

Post by zonker » Thu Jul 13, 2023 5:03 pm

Cedric35 wrote:
Thu Jul 13, 2023 12:11 pm
Thanks for the suggestions. I’ll give a minimum of 9 a try.
please come back to this thread and let us know how things went!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Cedric35
Posts: 4
Joined: Wed Jul 12, 2023 8:31 am

Re: Settings Suggestion

Post by Cedric35 » Fri Jul 14, 2023 6:34 am

Better results one night later. AHI: 3.1

https://imgur.com/CiObn5p

Thanks all!

User avatar
Miss Emerita
Posts: 3757
Joined: Sun Nov 04, 2018 8:07 pm

Re: Settings Suggestion

Post by Miss Emerita » Fri Jul 14, 2023 11:02 am

Definitely better! Settle in with this new setting for maybe 3 or 4 more days, then see what happens if you raise the minimum to 9.5.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zonker
Posts: 11339
Joined: Fri Jun 19, 2015 4:36 pm

Re: Settings Suggestion

Post by zonker » Fri Jul 14, 2023 11:38 am

Cedric35 wrote:
Fri Jul 14, 2023 6:34 am
Better results one night later. AHI: 3.1

https://imgur.com/CiObn5p

Thanks all!
definitely a move in the right direction!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe