Newbie here! Staying positive, but I keep waking up right as I doze off
Newbie here! Staying positive, but I keep waking up right as I doze off
UPDATED!
I didn't realize how much data was involved in all this, it's very cool. I now have OSCAR charts and metrics. I've posted the imgur links below.
Background info:
I'm 37 years old, male, 5'9, 170lbs. My snoring started out of no where about 5 years ago. I sleep with my mouth closed, breath through my nose. I snore SO loud and am very lethargic during the day after my morning coffee buzz wears off. My wife has to wear ear plugs while we sleep and she is still bothered by my snoring. So, I usually move to the couch. I really want to sleep in the same bed with my wife again and also want to get some good sleep so I can function better during the day. So, I did an at home sleep study using a WatchPAT device.
Here's the summary of results from my at home sleep study:
https://i.imgur.com/5p5y6ga.png
The at home sleep study indicates I have moderate sleep apnea. So, I do an overnight sleep study in a sleep lab. They started me on CPAP but after an hour they adjusted to BiPAP because they noticed I kept waking up due to struggling to exhale. Once the switches to BiPAP I was able to fall asleep and sleep pretty good, but it did confirm Obstructive Sleep Apnea.
Here's a summary and conclusion of my in-lab overnight sleep study:
https://i.imgur.com/ES3SuQl.png
https://i.imgur.com/1U7FO8O.png
My current situation:
I now have the AirCurve 10 with a AirFit N30i mask (I also have a DreamWisp that I like). They started me out with an IPAP of 9 and an EPAP of 5. I've tried this and found that I was waking up as soon as I doze off (I fall asleep very easily). Like, as soon as I would usually fall asleep and start snoring, I am being awaken by the fact that I am not inhaling air/breathing. It's very apparent to me, I can actually keep count of them. Also, I can hear it in the machine as well, almost like it's idling, waiting for me to take a breath.
I told my doctor this and they said they would change the pressure settings to IPAP of 15 and EPAP of 7. They dragged their feet, so I just adjusted the settings myself. I tried those settings and had the same results, being awakened by not breathing as soon as I doze off. I can only wear the mask for about 2hours until I need to take it off, so I can get some sleep.
Any suggestions or advice? Is it possible the BiPAP is causing more Central Sleep Apnea events? If you need any more charts or data, please let me know. Thank you all!
OSCAR Chart: https://i.imgur.com/eH9ncEn.png
I didn't realize how much data was involved in all this, it's very cool. I now have OSCAR charts and metrics. I've posted the imgur links below.
Background info:
I'm 37 years old, male, 5'9, 170lbs. My snoring started out of no where about 5 years ago. I sleep with my mouth closed, breath through my nose. I snore SO loud and am very lethargic during the day after my morning coffee buzz wears off. My wife has to wear ear plugs while we sleep and she is still bothered by my snoring. So, I usually move to the couch. I really want to sleep in the same bed with my wife again and also want to get some good sleep so I can function better during the day. So, I did an at home sleep study using a WatchPAT device.
Here's the summary of results from my at home sleep study:
https://i.imgur.com/5p5y6ga.png
The at home sleep study indicates I have moderate sleep apnea. So, I do an overnight sleep study in a sleep lab. They started me on CPAP but after an hour they adjusted to BiPAP because they noticed I kept waking up due to struggling to exhale. Once the switches to BiPAP I was able to fall asleep and sleep pretty good, but it did confirm Obstructive Sleep Apnea.
Here's a summary and conclusion of my in-lab overnight sleep study:
https://i.imgur.com/ES3SuQl.png
https://i.imgur.com/1U7FO8O.png
My current situation:
I now have the AirCurve 10 with a AirFit N30i mask (I also have a DreamWisp that I like). They started me out with an IPAP of 9 and an EPAP of 5. I've tried this and found that I was waking up as soon as I doze off (I fall asleep very easily). Like, as soon as I would usually fall asleep and start snoring, I am being awaken by the fact that I am not inhaling air/breathing. It's very apparent to me, I can actually keep count of them. Also, I can hear it in the machine as well, almost like it's idling, waiting for me to take a breath.
I told my doctor this and they said they would change the pressure settings to IPAP of 15 and EPAP of 7. They dragged their feet, so I just adjusted the settings myself. I tried those settings and had the same results, being awakened by not breathing as soon as I doze off. I can only wear the mask for about 2hours until I need to take it off, so I can get some sleep.
Any suggestions or advice? Is it possible the BiPAP is causing more Central Sleep Apnea events? If you need any more charts or data, please let me know. Thank you all!
OSCAR Chart: https://i.imgur.com/eH9ncEn.png
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Last edited by Brysleeps on Sat Jul 08, 2023 12:15 pm, edited 1 time in total.
- ChicagoGranny
- Posts: 15447
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
If you will post an OSCAR chart (or SleepHq.com for the less technically inclined) in this thread, some members will give you some informed advice. OSCAR instructions are pinned on the index page.
Re: Newbie here! Staying positive, but I keep waking up right as I doze off

to elaborate on what granny has said, the forum needs details of how your particular sleep goes. oscar can provide the forum with that info. this is important because we all don't sleep the same and what might work for me probably won't work for you.
also, when you get a moment, please fill out your equipment profile. that way, your mask and machine will appear under each of your posts as mine does below.
and yes, machine and mask make a difference.
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
Thank you! I will fill out all my info and upload my studies and OSCAR data. Thanks again!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
i see you have!
please post your oscar charts right here in this thread. see the stickies at the top of the forum for more info.
if you have any trouble, feel free to ask about it. we've all been there.
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
Thank you! I just reposted with all my info. I hope it's helpful. I really appreciated everyone's times and expertise.ChicagoGranny wrote: ↑Fri Jul 07, 2023 9:33 amIf you will post an OSCAR chart (or SleepHq.com for the less technically inclined) in this thread, some members will give you some informed advice. OSCAR instructions are pinned on the index page.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
I just updated my original post. I hope that's ok. If not, please let me know!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
Highly possible...
With bilevel fixed mode...that's the 7 EPAP and 15 IPAP....equals PS (pressure support) of 8 and pretty much guarantees central apneas popping up.
How did you come up with that much PS??? It's actually not a good thing to be doing that much PS without a damn good reason...and plain old OSA isn't a good reason.
Have you done a trial in auto mode with a much more reasonable PS? That's what I would recommend.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
Thank you for your input! Can you explain to me what PS is? Why 8? I assume it’s the difference between the numbers, but why is it significant?Pugsy wrote: ↑Sat Jul 08, 2023 1:01 pmHighly possible...
With bilevel fixed mode...that's the 7 EPAP and 15 IPAP....equals PS (pressure support) of 8 and pretty much guarantees central apneas popping up.
How did you come up with that much PS??? It's actually not a good thing to be doing that much PS without a damn good reason...and plain old OSA isn't a good reason.
Have you done a trial in auto mode with a much more reasonable PS? That's what I would recommend.
It was originally set to 9 - 5 and I told them about my issue with waking up, so they changed the numbers to 15 -7.
How do I try auto-mode? And what is a more reasonable PS?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
PS is pressure support or simply the difference between inhale and exhale.
Too much (too high) PS can cause a person to inhale and exhale to a point where they wash out too much carbon dioxide in their breaths. Centrals are caused when a person's carbon dioxide levels aren't high enough to for the brain to know it needs to send the signal to breathe. So too much PS causes carbon dioxide wash out which in turn causes the brain to not know it should breathe and a central apnea happens because a person simply isn't breathing.
I don't understand your problem with 5/9 and PS of 4 but changing it to 7 EPAP and 15 IPAP ends up with PS of 8 and it is a prime recipe for central apnea production because of carbon dioxide wash out.
A better choice IMHO would have been maybe try that EPAP of 7 but keep PS of 4.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
and you learn how to get into the clinical menu setup area and make changes to mode and other stuff.
As for a more reasonable PS....I don't ever suggest going with more than 5 PS unless special circumstances are present or a doctor is involved.
I use the AirCurve 10 VAuto....I use PS of 4 and I am in auto mode and not fixed mode.
My current settings are minimum EPAP at 5....PS at 4...maximum IPAP I just leave at the default of 25 but it never goes anywhere near the max. It works for me. So on inhale I get the minimum EPAP of 5 plus the PS of 4 which makes my inhale at 9 and then it can auto adjust as needed.
If nothing else...reduce your PS to 4...and to do that in fixed mode you would have EPAP at 7 and IPAP at 11.
Or try auto bilevel mode with minimum 7 Epap and PS of 4 (in auto mode you actually get a choice for PS instead of having to do the math) and see what happens.
I am sorry but I don't understand what the problem was with the original 5/9 settings except maybe they simply were sub optimal settings and not doing a good job holding the airway open.
But I do know that a PS of 8 is a really, really dumb idea for plain old OSA.
Too much (too high) PS can cause a person to inhale and exhale to a point where they wash out too much carbon dioxide in their breaths. Centrals are caused when a person's carbon dioxide levels aren't high enough to for the brain to know it needs to send the signal to breathe. So too much PS causes carbon dioxide wash out which in turn causes the brain to not know it should breathe and a central apnea happens because a person simply isn't breathing.
That's 5 EPAP (exhale) and 9 IPAP (inhale) which makes PS at 4 which is a much more normal PS.
I don't understand your problem with 5/9 and PS of 4 but changing it to 7 EPAP and 15 IPAP ends up with PS of 8 and it is a prime recipe for central apnea production because of carbon dioxide wash out.
A better choice IMHO would have been maybe try that EPAP of 7 but keep PS of 4.
You go here and request the provider/clinical manual for your machine
https://www.apneaboard.com/adjust-cpap- ... tup-manual
and you learn how to get into the clinical menu setup area and make changes to mode and other stuff.
As for a more reasonable PS....I don't ever suggest going with more than 5 PS unless special circumstances are present or a doctor is involved.
I use the AirCurve 10 VAuto....I use PS of 4 and I am in auto mode and not fixed mode.
My current settings are minimum EPAP at 5....PS at 4...maximum IPAP I just leave at the default of 25 but it never goes anywhere near the max. It works for me. So on inhale I get the minimum EPAP of 5 plus the PS of 4 which makes my inhale at 9 and then it can auto adjust as needed.
If nothing else...reduce your PS to 4...and to do that in fixed mode you would have EPAP at 7 and IPAP at 11.
Or try auto bilevel mode with minimum 7 Epap and PS of 4 (in auto mode you actually get a choice for PS instead of having to do the math) and see what happens.
I am sorry but I don't understand what the problem was with the original 5/9 settings except maybe they simply were sub optimal settings and not doing a good job holding the airway open.
But I do know that a PS of 8 is a really, really dumb idea for plain old OSA.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
Not sure if this is helpful but I have an AirSense 11 and I'm able to turn off the ramp time. It's still on auto but instead of waiting til I go to sleep it starts upping the air as I need it right away. I was having the same issue of waking up all the time when I quit breathing. I am starting to actually sleep with it now that I turned the ramp time off. Good luck.
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
| Additional Comments: i don't know what the software and humidifier are |
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
Thank you! I am learning so much. Please excuse any stupid questions.
Thank you so much for your input! I will take the advice/suggestions!
My Problem is that when my setting were on 5/9, I was waking up as I dozed off, due to forgetting to breathe. My doctor suggested the increase to 15/7.I don't understand your problem with 5/9 and PS of 4 but changing it to 7 EPAP and 15 IPAP ends up with PS of 8 and it is a prime recipe for central apnea production because of carbon dioxide wash out.
A better choice IMHO would have been maybe try that EPAP of 7 but keep PS of 4.
This helped a lot! I was wondering how in the heck you got a PS of 4 from 5 - 25, but I see when I switch to Vauto mode, I can now choose my PS. Very cool, Thank you!If nothing else...reduce your PS to 4...and to do that in fixed mode you would have EPAP at 7 and IPAP at 11.
Or try auto bilevel mode with minimum 7 Epap and PS of 4 (in auto mode you actually get a choice for PS instead of having to do the math) and see what happens.
Thank you so much for your input! I will take the advice/suggestions!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
Thank you! I've turned off ramp mode.mp6617 wrote: ↑Sat Jul 08, 2023 3:12 pmNot sure if this is helpful but I have an AirSense 11 and I'm able to turn off the ramp time. It's still on auto but instead of waiting til I go to sleep it starts upping the air as I need it right away. I was having the same issue of waking up all the time when I quit breathing. I am starting to actually sleep with it now that I turned the ramp time off. Good luck.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
In fixed mode there is no separate PS setting to control. Instead you simply do the math and PS/Pressure support is simply the difference between EPAP and IPAP. So you regulate PS by changing EPAP and/or IPAP.
In Auto mode it gives us a separate choice for whatever PS we want to use.
End result is still the same...just a different way of getting there.
How do you know you were forgetting to breathe? Maybe you were having OAs and they were causing the wake ups.
Forgetting to breathe is central apneas....airway is open but no air is moving because you aren't breathing.
Holding your breath...airway is open but no air is moving...that's a central apnea (Clear airway on OSCAR).
Now there is a thing called sleep onset central which can sometimes happen to some people when they transition from awake to asleep but we don't know for sure that is what is going on with you.
I don't have time to sort through all the potential stuff...lets see what happens tonight with more normal settings and go from there.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Newbie here! Staying positive, but I keep waking up right as I doze off
Gotcha, thank you for the clarification.End result is still the same...just a different way of getting there.
How do you know you were forgetting to breathe?
Well, I could be mistaking it for something else, but it happens every time i try to sleep with the mask on. I doze-off...then I gently wake up, noticing I'm not taking a breath. I notice the whir of the machine also seem to idle, then I take a breath and all is good until I begin to doze-off again. It seems as if I am holding my breath. This has happened every night for 5 nights. I try the mask for over an hour and can't fall asleep due to this.
Thank you so much for you time! I can't thank you enough. I will try tonight and report back!I don't have time to sort through all the potential stuff...lets see what happens tonight with more normal settings and go from there.