humidifier issues

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
janhalcomb
Posts: 1
Joined: Sat Jun 24, 2023 8:07 pm

humidifier issues

Post by janhalcomb » Sat Jun 24, 2023 8:12 pm

I seem to have a problem no one else is experiencing. My water chamber runs out of water completely every night and I wake up several times with a really dry mouth. It is set at 3 and I do not use a heated hose. anyone have any suggestions. thank you

User avatar
LSAT
Posts: 13374
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: humidifier issues

Post by LSAT » Sun Jun 25, 2023 7:16 am

Mouth breathing will cause excessive water use.

User avatar
Pugsy
Posts: 65198
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: humidifier issues

Post by Pugsy » Sun Jun 25, 2023 7:27 am

These things affect water consumption.

1...ambient humidity...missed by a lot of people. If your ambient humidity in your house is low then more water will be consumed trying to maintain whatever humidity setting a person prefers
2...humidifier setting..obvious
3...pressures used..higher pressures will use up more water
4...leaks...either mouth breathing leaks that are excessive or just large amounts of mask movement leaks. The more leak you have the more water gets consumed.