Resmed 10 Heat Plate
Resmed 10 Heat Plate
Hi my heat plate does not heat but everything else works -no alert on screen-if the fuse is blown should a fault be shown on screen-first l will test fuse easy bit first-if fuse is ok what should l check next and how-one last thing-if its the heat plate and l change it-is it plug and play or is there some calibration needed.Any help would be greatly appreciated-thanks
Re: Resmed 10 Heat Plate
Is the humidity turned on?
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is a new AS10. |
- Dog Slobber
- Posts: 4215
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Resmed 10 Heat Plate
I don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.
Remove all water from the the tank, put the tank back in, set the humidity to 8, and then run it for 10-15 minutes. Turn it off, remove tank and feel heater plate. Put a thermometer on it, and/or feel the plate.
If it's not hot, then yes, somethings wrong. might be the fuse.
Remove all water from the the tank, put the tank back in, set the humidity to 8, and then run it for 10-15 minutes. Turn it off, remove tank and feel heater plate. Put a thermometer on it, and/or feel the plate.
If it's not hot, then yes, somethings wrong. might be the fuse.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Re: Resmed 10 Heat Plate
Just what we need, a check engine light on the CPAP!Dog Slobber wrote: ↑Mon Apr 17, 2023 8:26 amI don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.



Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Resmed 10 Heat Plate
maybe with an a-oo-gah! horn as well?palerider wrote: ↑Mon Apr 17, 2023 4:53 pmJust what we need, a check engine light on the CPAP!Dog Slobber wrote: ↑Mon Apr 17, 2023 8:26 amI don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.![]()
![]()
![]()
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Resmed 10 Heat Plate
Hi fuse checked=ok - humidity was on one put to 4 plate heats up thanks for your adviceDog Slobber wrote: ↑Mon Apr 17, 2023 8:26 amI don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.
Remove all water from the the tank, put the tank back in, set the humidity to 8, and then run it for 10-15 minutes. Turn it off, remove tank and feel heater plate. Put a thermometer on it, and/or feel the plate.
If it's not hot, then yes, somethings wrong. might be the fuse.
Re: Resmed 10 Heat Plate
Well had fuse checked=ok- humidity was on 1 put it on 4 plate heats up-thanks to all for your suggestions- one more thing if l need to replace the heat plate can l unplug the old plate and plug new plate in- without disturbing the calibration-thanks
- Dog Slobber
- Posts: 4215
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Resmed 10 Heat Plate
Unfortunately for the OP, the Check Engine Light feature wasn't introduced until the Air 11.palerider wrote: ↑Mon Apr 17, 2023 4:53 pmJust what we need, a check engine light on the CPAP!Dog Slobber wrote: ↑Mon Apr 17, 2023 8:26 amI don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.![]()
![]()
![]()

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Re: Resmed 10 Heat Plate
The Air 11 need a code scanner as well?
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Started 2005 with a Remstar machine. Then a 2020 Dreamstation 1. Now a Resmed Aircurve ST |
Re: Resmed 10 Heat Plate
Dog Slobber wrote: ↑Tue Apr 18, 2023 9:13 amUnfortunately for the OP, the Check Engine Light feature wasn't introduced until the Air 11.palerider wrote: ↑Mon Apr 17, 2023 4:53 pmJust what we need, a check engine light on the CPAP!Dog Slobber wrote: ↑Mon Apr 17, 2023 8:26 amI don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.![]()
![]()
![]()
![]()



Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Resmed 10 Heat Plate
Sure beats pulling out the planchet and Oija Board.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |