Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Thu Apr 13, 2023 4:06 pm

Here is my situation:
For the last 10 years, I've been using the System One series 50 with great success. Due to foamgate, I bought a Resmed 10.
I'm using the same mask (Swift FX Medium) and I'm using the tubing that came with the Resmed with the heating function of the tube turned off.
Every since the switch I wake up at either 3 or 430am.
When I switch back to the System One, everything works fine.

For the System one I have the following settings:
Pressure 9.0
Cflex 3

For the resmed I have pressure 9.0 EPR 3

I have tried using it an APA mode from 9.0 to 11.0
I tried turning off the EPR after ramping.

Any suggestions?
I can provide Oscar settings if needed. Please let me know.

Any help would be appreciated.

_________________
MachineMask
Additional Comments: Also use System one Series 50

jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Fri Apr 14, 2023 5:06 am

Bump

_________________
MachineMask
Additional Comments: Also use System one Series 50

User avatar
ChicagoGranny
Posts: 15085
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by ChicagoGranny » Fri Apr 14, 2023 6:44 am

jojopuppyfish wrote:
Thu Apr 13, 2023 4:06 pm
I can provide Oscar settings if needed.
Post 'em.
jojopuppyfish wrote:
Thu Apr 13, 2023 4:06 pm
Every since the switch I wake up at either 3 or 430am.
You don't say whether you are able to go back to sleep.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by chunkyfrog » Fri Apr 14, 2023 8:54 am

How long has this been going on?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Fri Apr 14, 2023 2:43 pm

It's happening every time I use the CPAP
Here is a JPG I uploaded to FLickr
https://www.flickr.com/photos/jojopuppy ... datetaken/
I was able to get back to sleep. But this doesn't happen with the System One

ImageOSCAR Resmed 10 by , on Flickr

_________________
MachineMask
Additional Comments: Also use System one Series 50

User avatar
robysue1
Posts: 1298
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by robysue1 » Fri Apr 14, 2023 9:27 pm

jojopuppyfish wrote:
Fri Apr 14, 2023 2:43 pm
It's happening every time I use the CPAP
Here is a JPG I uploaded to FLickr
https://www.flickr.com/photos/jojopuppy ... datetaken/
I was able to get back to sleep. But this doesn't happen with the System One

ImageOSCAR Resmed 10 by , on Flickr
Did the wake for this night happen at 3:00? Or at 4:00?

In general, if the wake is short and you get back to sleep quickly, you don't need to worry about it.

But out of curiosity: How does the leak line when you are using the Resmed 10 compare to the leak line on the System One? Keep in mind that in Oscar, you have two leak lines for the System One; the bottom one is the one that is equivalent to the single leak line for the Resmed 10.
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls

jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Sat Apr 15, 2023 5:03 am

The particular night wakeup that I posted was at 4:30am.
Previous night on the Resmed was 3:30.

As for the Leak Rate, on the Resmed its been all over the place with a range of anywhere from 16 to 26. Doesn't look consistent at all . On the system one it is 30ish

Here is my system one from last night
ImageSystemOne 4-14-2023 by , on Flickr

_________________
MachineMask
Additional Comments: Also use System one Series 50

jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Sat Apr 15, 2023 5:07 am

Here is my posting of my system one last night with the leak rate data to the left
ImageSystemOne(part2 with Leak Rate) 4-14-2023

_________________
MachineMask
Additional Comments: Also use System one Series 50

jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Mon Apr 17, 2023 4:47 am

So I figured out a solution to my problem. Someone mentioned above about leak rate. Well I've been using my Swift FX medium mask for 10 years with great success....with the System One.
On a hunch I bought the P10 Resmed mask which is similar but some people said was better than the Swift FX.
I tried it for two nights in a row on both the resmed 10 and my Dreamstation 2 with great success. It leaks much less than the Swift FX
I hope this helps anyone with making the transaction to a new machine. It might be time to think about getting a new mask too.

_________________
MachineMask
Additional Comments: Also use System one Series 50

jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Wed May 10, 2023 4:36 am

Just an update.
The p10 was a little tight and causing me headaches.
So I switched to the Brevida mask and I now have had 2 successful nights where I don't wake up in the middle of the night.
This mask does a great job of reducing air leaks.
Conclusion:
The Swift FX was great for the System one. But the Resmed 10 is not very tolerant with high air flows. The Brevida definitely has lower air flow leaks.
Hope that helps someone out there.

_________________
MachineMask
Additional Comments: Also use System one Series 50

jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Sat May 27, 2023 5:33 am

Just want to give an update for future reference.
Even the Brevida mask was causing me to wake up sometimes at 3am or 4:30.
Then I came up with an idea:
I would go back to the P10 but I would look for other headgear that wasn't as tight.
And I found it.
The headgear made by Medihealer isn't nearly as tight at the Resmed headgear that comes with the P10.
https://www.amazon.com/dp/B09ZXVM2PV?ps ... ct_details

I have now had 2 good nights of sleep without walking up.
Again I am writing this for anyone else who runs into the same problems I have had with the move from a successful old CPAP to the latest CPAP machines.
The adjustment is not as easy as it would seem on paper.

_________________
MachineMask
Additional Comments: Also use System one Series 50

User avatar
zonker
Posts: 11236
Joined: Fri Jun 19, 2015 4:36 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by zonker » Sat May 27, 2023 9:36 pm

jojopuppyfish wrote:
Sat May 27, 2023 5:33 am


I have now had 2 good nights of sleep without walking up.
Again I am writing this for anyone else who runs into the same problems I have had with the move from a successful old CPAP to the latest CPAP machines.
The adjustment is not as easy as it would seem on paper.
Image

thanks for the update. always glad to hear it was a simple fix!
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

jojopuppyfish
Posts: 14
Joined: Thu Apr 13, 2023 3:38 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by jojopuppyfish » Sat Aug 05, 2023 5:59 am

Just wanted to update my situation:
After watching this video, I changed my Resmed 10 to full face mask and went back to using my Swift FX medium mask.
https://www.youtube.com/watch?v=WpQC7TgCoe8
It made an immediate difference.
It's been a few months now and I'd say the Redmed 10 is getting me the same results as my old system one.
But it certainly took some time.

_________________
MachineMask
Additional Comments: Also use System one Series 50

User avatar
ozij
Posts: 10435
Joined: Fri Mar 18, 2005 11:52 pm

Re: Switching from System One to Resmed 10 and now I wake up at 3 or 4am every night

Post by ozij » Sat Aug 05, 2023 8:13 am

Thanks for the update!

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023