My same newbie post. Continuing to update
-
Trainer4346
- Posts: 40
- Joined: Thu Feb 16, 2023 7:29 am
My same newbie post. Continuing to update
This is my first night and it's a fail. Airsense 11 shuts off because of leak. Says check tubing. The same thing happened with the test drive and when I went back to show them it of course didn't happen at the office! Now here I am at home and it isnt working again. Kinda PO'd. UPDATE: So the leak is at the tub. You have to really jam it in. For a $1000 machine that's BS. The other question I have is how do you know it's on? There's no power light and it's so quiet I'm not even sure. MORNING UPDATE: Thanks for your help. The mask was blowing and the machine tried. I just was wide awake all night. I think as soon as I open my mouth I wake up from the hiss of air. This is gonna take some getting used to ... What's the average amount of time before things start working?
_________________
| Machine: AirSense 11 Autoset |
| Mask: DreamWear Silicone Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Heat and humidity set on high end. Switched from nasal cradle style to pillows. Added mouth tape 4/27 |
Last edited by Trainer4346 on Tue Apr 18, 2023 2:41 pm, edited 5 times in total.
Diagnosed with 28 events per hour, first night with CPAP 4/12/23. Setting is 10.
- loggerhead12
- Posts: 585
- Joined: Fri Sep 01, 2017 4:00 pm
Re: Super frustrated with dreamwear nasal and airsense 11
Both of those things are good when you're trying to sleep.Trainer4346 wrote: ↑Wed Apr 12, 2023 8:20 pmThere's no power light and it's so quiet I'm not even sure.
You'll know it's on when it blows air up your nose.
Re: Super frustrated with dreamwear nasal and airsense 11
Pull the mask just a bit away from your nose: If the machine is on, you'll feel a blast of air as the machine compensates for the leak you just created.Trainer4346 wrote: ↑Wed Apr 12, 2023 8:20 pmThe other question I have is how do you know it's on? There's no power light and it's so quiet I'm not even sure.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
Re: Super frustrated with dreamwear nasal and airsense 11
Touch the screen. If the machine is on, the display will show the current pressure.
- ChicagoGranny
- Posts: 15523
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Updated: Super frustrated with dreamwear nasal and airsense 11
You will have to solve the mouth-opening problem. A full face mask (covers nose and mouth) is one solution. Others - taping mouth shut, wearing a cervical collar, wearing a chinstrap.Trainer4346 wrote: ↑Wed Apr 12, 2023 8:20 pmI think as soon as I open my mouth I wake up from the hiss of air.
When your mouth opens while using a nasal-only mask, the therapeutic pressure is likely to be lost.
It's a statistic that means nothing to any individual case.Trainer4346 wrote: ↑Wed Apr 12, 2023 8:20 pmWhat's the average amount of time before things start working?