Settings suggestion -- partner still hearing snoring

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
cpaperic
Posts: 3
Joined: Sun Apr 09, 2023 10:21 am

Settings suggestion -- partner still hearing snoring

Post by cpaperic » Sun Apr 09, 2023 10:41 am

Resmed Autosense 10 with P30i nasal pillow.

I've been using a CPAP since 2020/2021 for mild/moderate sleep apnea. I don't remember what my initial sleep lab results are but I got the sense the sleep techs there didn't know what they were doing so I've been on an experimentation journey since. Quit on and off since then when it didn't seem to be helping. First nasal mask I tried hurt my noise. Tried a full face mask and it seemed to be damaging the bridge of my nose. Finally things clicked into place as I found a mask that seemed to work for me in January 2023 (P30i). However, my spouse still complains of snoring sometimes (sometimes chainsaw-esque, sometimes like a large lion or velociraptor purring), and is awoken by leaks 1-2x a night. The snoring seems to coincide somewhat with leakage, best I can tell from the data and her recollections. I had my settings at Min 8 max 12 with EPR3, and pressure rarely seemed to ever even reach 12, but I bumped max to 15 last night just in case. Does anything stand out to you on this chart? Any settings I should change?

Image

Here's a chart from the night before, and she noted my weird snoring at about 7:15, which was a little bit after an event:

Image

Bonus -- she took a recording of the snoring:

http://sndup.net/w8by

Thanks for any tips.

User avatar
Pugsy
Posts: 65301
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Settings suggestion -- partner still hearing snoring

Post by Pugsy » Sun Apr 09, 2023 11:49 am

The minimum pressure setting is your most critical setting especially when someone is still snoring too much.
Snores mean the airway is still trying to collapse a bit. Now it may not fully collapse to the point of apnea events materializing enough to get flagged.

You need to better PREVENT the impending collapse in the first place instead of trying to fix it with more max pressure.
Try a small increase in the minimum pressure....usually doesn't take much in situations like yours where everything else looks pretty good.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Settings suggestion -- partner still hearing snoring

Post by Miss Emerita » Sun Apr 09, 2023 2:44 pm

Well, it does sound like a bit of soft snoring, but the panels on your charts say that for at least 99.5% of the time you’re not snoring at all. If you look at your snore graph, do you see any flags?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

cpaperic
Posts: 3
Joined: Sun Apr 09, 2023 10:21 am

Re: Settings suggestion -- partner still hearing snoring

Post by cpaperic » Sun Apr 09, 2023 3:23 pm

Miss Emerita wrote:
Sun Apr 09, 2023 2:44 pm
Well, it does sound like a bit of soft snoring, but the panels on your charts say that for at least 99.5% of the time you’re not snoring at all. If you look at your snore graph, do you see any flags?
The snore graph shows snoring a little after the apnea event.

Image

cpaperic
Posts: 3
Joined: Sun Apr 09, 2023 10:21 am

Re: Settings suggestion -- partner still hearing snoring

Post by cpaperic » Sun Apr 09, 2023 3:25 pm

Pugsy wrote:
Sun Apr 09, 2023 11:49 am
The minimum pressure setting is your most critical setting especially when someone is still snoring too much.
Snores mean the airway is still trying to collapse a bit. Now it may not fully collapse to the point of apnea events materializing enough to get flagged.

You need to better PREVENT the impending collapse in the first place instead of trying to fix it with more max pressure.
Try a small increase in the minimum pressure....usually doesn't take much in situations like yours where everything else looks pretty good.
Thank you, I will try this and report back in a week or so.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Settings suggestion -- partner still hearing snoring

Post by Miss Emerita » Sun Apr 09, 2023 5:26 pm

Tiny bit of snoring, well after the OA. But do try Pugsy’s suggestion. If it works, I’m sure your wife will thank you.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Settings suggestion -- partner still hearing snoring

Post by palerider » Sun Apr 09, 2023 5:54 pm

The *standard* response to 'snoring' is 'more pressure'.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.