Airsense 11 - Central Apneas While Awake

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
greyhound
Posts: 471
Joined: Fri Nov 11, 2005 1:44 pm
Location: Pennsylvania

Airsense 11 - Central Apneas While Awake

Post by greyhound » Thu Apr 06, 2023 2:55 pm

Long-time CPAP user who recently upgraded (?) from a Dreamstation 2 Auto to a Resmed AirSense 11 Auto. Although central apneas were never a problem in the past, the Airsense has been reporting a large numper of centrals - most of which "ocurred" while I am certain I was awake and unable to fall back to sleep. The new machine is very comfortable and my settings (10.6-14) are consistent with what they were with my Dreamstation 2 and earlier Respironics models.

Is there something I'm missing about Resmed machines or is my machine (I've had it for about a week) likely to be defective? I've tried, but was unable, to attach a representative Oscar daily report.

_________________
Machine: AirSense 11 Autoset
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Oscar software Previous Masks: Airfit P10 Nasal Pillow, Swift FX Nasal Pillow, Comfort Curve, Opus, Mirage Swift II

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Airsense 11 - Central Apneas While Awake

Post by zonker » Thu Apr 06, 2023 3:20 pm

greyhound wrote:
Thu Apr 06, 2023 2:55 pm
I've tried, but was unable, to attach a representative Oscar daily report.
you have to go to an image hosting site and then post the link here in the forum.

please see-
viewtopic/t183806/Attachment-storage-is ... MENTS.html

if you need any help, just ask.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Airsense 11 - Central Apneas While Awake

Post by Miss Emerita » Thu Apr 06, 2023 3:26 pm

No brand of machine can tell whether you're awake or not. When we're awake, it's common to pause between breaths or to hold our breath (e.g., while concentrating). Any pause or breath-holding that lasts for 10 seconds or longer will be flagged by any machine as a CA.

That's not to say the machines are the same. The algorithms for ResMed machines are a little more aggressive than those for PR machines. PR machines don't raise pressure as quickly in response to flow limitations, snores, and even obstructive events. But if all the differences you're noticing concern time when you're awake, that shouldn't matter.

Are you using EPR, and did you use Flex with your PR machine?

The forum site maxes out regularly in its attachment capacity. Try using an image-hosting site like Imgur for your Oscar screenshot; then provide a link.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
ChicagoGranny
Posts: 15533
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Airsense 11 - Central Apneas While Awake

Post by ChicagoGranny » Thu Apr 06, 2023 4:36 pm

Miss Emerita wrote:
Thu Apr 06, 2023 3:26 pm
No brand of machine can tell whether you're awake or not.
I love this juxtaposition - viewtopic/t186479/It-knows-when-you-are ... l-you.html :D

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Airsense 11 - Central Apneas While Awake

Post by Miss Emerita » Thu Apr 06, 2023 7:49 pm

Yes, things come up in clusters, don’t they?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

greyhound
Posts: 471
Joined: Fri Nov 11, 2005 1:44 pm
Location: Pennsylvania

Re: Airsense 11 - Central Apneas While Awake

Post by greyhound » Thu Apr 06, 2023 7:55 pm

Miss Emerita wrote:
Thu Apr 06, 2023 7:49 pm
Yes, things come up in clusters, don’t they?
Except the huge cluster came over a two hour period when I was awake. I have trouble believing my resting breathing could be interpreted as central apneas.

_________________
Machine: AirSense 11 Autoset
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Oscar software Previous Masks: Airfit P10 Nasal Pillow, Swift FX Nasal Pillow, Comfort Curve, Opus, Mirage Swift II

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Airsense 11 - Central Apneas While Awake

Post by Miss Emerita » Thu Apr 06, 2023 8:04 pm

greyhound wrote:
Thu Apr 06, 2023 7:55 pm
Miss Emerita wrote:
Thu Apr 06, 2023 7:49 pm
Yes, things come up in clusters, don’t they?
Except the huge cluster came over a two hour period when I was awake. I have trouble believing my resting breathing could be interpreted as central apneas.
Let’s take a look at an Oscar chart. Please also include a zoomed in image that captures a number of your waking CAs in maybe a 12 minute stretch.

Just so you know, the things I was referring to were topics in threads that come up at the same time, as Chicago Granny was pointing out.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 11 - Central Apneas While Awake

Post by Pugsy » Thu Apr 06, 2023 8:35 pm

greyhound wrote:
Thu Apr 06, 2023 7:55 pm
I have trouble believing my resting breathing could be interpreted as central apneas.

Believe it. It happens all the time. Hold your breath for 10 seconds...that's essentially a 10 second central apnea.
No air is moving because you are holding your breath. We pause our breathing all the time while awake and don't realize we are doing it.

Below is an example of a truckload of central apneas all flagged while I was awake and I know for sure I was awake.
I was battling a big old insomnia monster and didn't even remotely fall asleep until after 1:30.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 11 - Central Apneas While Awake

Post by palerider » Thu Apr 06, 2023 11:28 pm

greyhound wrote:
Thu Apr 06, 2023 2:55 pm
Is there something I'm missing about Resmed machines or is my machine (I've had it for about a week) likely to be defective?
No, what you're unaware of is that awake breathing is notoriously irregular, and your machine is just letting you know about it.
Just ignore it.
greyhound wrote:
Thu Apr 06, 2023 2:55 pm
I've tried, but was unable, to attach a representative Oscar daily report.
If only you'd read the pinned posts, like this one: viewtopic/t183806/Attachment-storage-is ... 2b9231ac11

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

greyhound
Posts: 471
Joined: Fri Nov 11, 2005 1:44 pm
Location: Pennsylvania

Re: Airsense 11 - Central Apneas While Awake

Post by greyhound » Fri Apr 07, 2023 11:24 am

palerider wrote:
Thu Apr 06, 2023 11:28 pm
greyhound wrote:
Thu Apr 06, 2023 2:55 pm
Is there something I'm missing about Resmed machines or is my machine (I've had it for about a week) likely to be defective?
No, what you're unaware of is that awake breathing is notoriously irregular, and your machine is just letting you know about it.
Just ignore it.

Your experience is similar to Pugsy's and, I'm sure, many others. Resmed's software must diagnose centrals much more frequently than the three Respironics machines I've had in the past. With Respironics, I rarely had centrals, but now - awake or asleep - I've got a lot (relatively) of centrals and few OAs or Hs
greyhound wrote:
Thu Apr 06, 2023 2:55 pm
I've tried, but was unable, to attach a representative Oscar daily report.
If only you'd read the pinned posts, like this one: viewtopic/t183806/Attachment-storage-is ... 2b9231ac11
Guilty as charged. As a long-time cpaptalk member (albeit absent for many years), I assumed, incorrectly, that the sticky notes didn't apply to me. I did read them after the first reply to my original post, but was reluctant to give my phone number and upload images to a site I knew nothing about.

_________________
Machine: AirSense 11 Autoset
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Oscar software Previous Masks: Airfit P10 Nasal Pillow, Swift FX Nasal Pillow, Comfort Curve, Opus, Mirage Swift II

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 11 - Central Apneas While Awake

Post by palerider » Fri Apr 07, 2023 1:53 pm

greyhound wrote:
Fri Apr 07, 2023 11:24 am
but was reluctant to give my phone number and upload images to a site I knew nothing about.
We're a generally trustworthy group ;)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

greyhound
Posts: 471
Joined: Fri Nov 11, 2005 1:44 pm
Location: Pennsylvania

Re: Airsense 11 - Central Apneas While Awake

Post by greyhound » Fri Apr 07, 2023 3:01 pm

I wasn’t referring to CPAP talk.com (I’ve been a member since 2005) or CPAP.com. I was referring to the image repository which I had never heard of before.

_________________
Machine: AirSense 11 Autoset
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Oscar software Previous Masks: Airfit P10 Nasal Pillow, Swift FX Nasal Pillow, Comfort Curve, Opus, Mirage Swift II

User avatar
robysue1
Posts: 1311
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: Airsense 11 - Central Apneas While Awake

Post by robysue1 » Fri Apr 07, 2023 6:23 pm

greyhound wrote:
Fri Apr 07, 2023 3:01 pm
I wasn’t referring to CPAP talk.com (I’ve been a member since 2005) or CPAP.com. I was referring to the image repository which I had never heard of before.
imgur.com should allow you to register for free without giving it any phone number. And imgur.com is what a lot of us use for posting images over here.

Do note: imgur does have a "pay" account as well as a freebie. I use the freebie and the main difference as far as I can tell is pop-up ads when you are doing things like uploading images. So far there's been nothing to indicate there is a quota for what I can upload over there. And copying the bbcode link for an image over there and pasting it over here is really easy.
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 11 - Central Apneas While Awake

Post by palerider » Fri Apr 07, 2023 10:17 pm

greyhound wrote:
Fri Apr 07, 2023 3:01 pm
I wasn’t referring to CPAP talk.com (I’ve been a member since 2005) or CPAP.com. I was referring to the image repository which I had never heard of before.
I never gave imgur anything other than my email address.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
ChicagoGranny
Posts: 15533
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Airsense 11 - Central Apneas While Awake

Post by ChicagoGranny » Sat Apr 08, 2023 6:28 am

Imgur had so many problems with spam that they now require a phone number for verification of new accounts.