Desperately Frustrated sleeper seeking help

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
doriluigi28jk92
Posts: 8
Joined: Thu Mar 16, 2023 10:56 pm

Desperately Frustrated sleeper seeking help

Post by doriluigi28jk92 » Thu Mar 16, 2023 11:12 pm

Hi. I’ve been on CPAP/APAP for years.
I had to get a replacement new ResMed Airsense 10 APAP machine on Feb 9 as the motor broke on the other machine. they gave me a new machine.

Since then, I’ve been waking every single day with a dry mouth. I’m also having more events than I’ve ever had in years. Before I got the new machine, no dry mouth, close to 0 events and I was using that machine since 2021. I have no clue what the humidity or tube temp was set on the other machine I returned. How do I fix this issue? No one seems to know when I ask around the community so I’m asking here. Nothing changed except a replacement machine. I’m using the Bleep mask, same mask I’ve been using since 2021. I do know what the sleep settings were on old machine and I changed the new machine to reflect this. This issues started that night so I know it’s the machine settings.

I also have a humidifier in my bedroom and I do use that. I also used that for the last year with the other machine and never encountered any issues with dry mouth and a lot of events per hour.

What do I need to set the tube temp at and the humidity level on the machine? The humidity in my bedroom gets to between 53-59%. It’s Arizona so really dry. I’m about to cry cuz this is so frustrating. I hope someone can please help guide me. Thanks so much

ozij
Posts: 10542
Joined: Fri Mar 18, 2005 11:52 pm

Re: Desperately Frustrated sleeper seeking help

Post by ozij » Fri Mar 17, 2023 7:57 am

Did you have and SD card in your old machine? Did you keep it?
If you did, there's a program that will tell us everything about the settings on your old machine.
Please read the first announcement the forum, and follow the intructions:viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

If you don't have the SD card from the old machine, check if you have one in the new machine. If you don't, buy one, insert it, keep it there when you're sleeping and upload the data as instructed.
The will help us see all present setting, and what's going on with your breathing as you sleep.

We need the data in order to help you figure out what's going on.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Desperately Frustrated sleeper seeking help

Post by Julie » Fri Mar 17, 2023 8:58 am

You could try various settings for the humidifier over a couple of nights each and see what works for you... just be careful not to congest your nose with a very high setting unless it works well for you.


doriluigi28jk92
Posts: 8
Joined: Thu Mar 16, 2023 10:56 pm

Re: Desperately Frustrated sleeper seeking help

Post by doriluigi28jk92 » Fri Mar 17, 2023 4:45 pm

I've already read this and have been using Resmed Airsense for years, but thank you.

doriluigi28jk92
Posts: 8
Joined: Thu Mar 16, 2023 10:56 pm

Re: Desperately Frustrated sleeper seeking help

Post by doriluigi28jk92 » Fri Mar 17, 2023 4:54 pm

The old SD card was in the old machine. I bought a new one. I posted them here. Thanks for reviewing them.

doriluigi28jk92
Posts: 8
Joined: Thu Mar 16, 2023 10:56 pm

Re: Desperately Frustrated sleeper seeking help

Post by doriluigi28jk92 » Fri Mar 17, 2023 4:56 pm

Here is the data. I appreciate anyone looking at it. Thanks.

doriluigi28jk92
Posts: 8
Joined: Thu Mar 16, 2023 10:56 pm

Re: Desperately Frustrated sleeper seeking help

Post by doriluigi28jk92 » Fri Mar 17, 2023 5:07 pm

Here is the data. Would love some guidance, anything, I need to sleep as I feel like I'm losing my mind.
I turned the humidity level up last night to a 4 and woke up with dry throat, headache, and water chamber close to empty (which has never ever happened before this new stupid machine).

Thanks.

User avatar
zonker
Posts: 11488
Joined: Fri Jun 19, 2015 4:36 pm

Re: Desperately Frustrated sleeper seeking help

Post by zonker » Fri Mar 17, 2023 8:45 pm

doriluigi28jk92 wrote:
Fri Mar 17, 2023 5:07 pm
Here is the data. Would love some guidance, anything, I need to sleep as I feel like I'm losing my mind.
I turned the humidity level up last night to a 4 and woke up with dry throat, headache, and water chamber close to empty (which has never ever happened before this new stupid machine).

Thanks.
viewtopic/t183806/Attachment-storage-is ... MENTS.html

and please stay in one thread.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

doriluigi28jk92
Posts: 8
Joined: Thu Mar 16, 2023 10:56 pm

Re: Desperately Frustrated sleeper seeking help

Post by doriluigi28jk92 » Fri Mar 17, 2023 10:41 pm

I don’t follow what you’re saying. I am in one thread.
Where do I post the Oscar data for someone to look at it?
Thx

ozij
Posts: 10542
Joined: Fri Mar 18, 2005 11:52 pm

Re: Desperately Frustrated sleeper seeking help

Post by ozij » Fri Mar 17, 2023 11:23 pm

Zonker is right, you created two threads:

This is the second one, unnecessary, and containing no data:
viewtopic/t186349/Sleep-Deprived--Despe ... -Help.html

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

ozij
Posts: 10542
Joined: Fri Mar 18, 2005 11:52 pm

Re: Desperately Frustrated sleeper seeking help

Post by ozij » Fri Mar 17, 2023 11:30 pm

doriluigi28jk92 wrote:
Fri Mar 17, 2023 4:54 pm
The old SD card was in the old machine. I bought a new one. I posted them here. Thanks for reviewing them.
If you have OSCAR data from both machines, that's great.

If you don't know how to post the images, Zonker will no doubt help you.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Desperately Frustrated sleeper seeking help

Post by chunkyfrog » Fri Mar 17, 2023 11:36 pm

You need to use an image hosting service, as direct attachments are impossible.
This is because the forum sponsors refuse to fix the problem.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
zonker
Posts: 11488
Joined: Fri Jun 19, 2015 4:36 pm

Re: Desperately Frustrated sleeper seeking help

Post by zonker » Sat Mar 18, 2023 11:18 am

ozij wrote:
Fri Mar 17, 2023 11:30 pm
doriluigi28jk92 wrote:
Fri Mar 17, 2023 4:54 pm
The old SD card was in the old machine. I bought a new one. I posted them here. Thanks for reviewing them.
If you have OSCAR data from both machines, that's great.

If you don't know how to post the images, Zonker will no doubt help you.
Image
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

doriluigi28jk92
Posts: 8
Joined: Thu Mar 16, 2023 10:56 pm

Re: Desperately Frustrated sleeper seeking help

Post by doriluigi28jk92 » Sun Mar 19, 2023 11:44 am

Hi, here are the Oscar reports:

https://www.flickr.com/photos/197939421 ... d080r69815

I attached last 3 months and the one from last night. If this doesn't work, I give up. I'm too exhausted to try to figure this out. Thanks to whoever looks at this and lets me know what you see.

Recap:
1. Been using the same Resmed Airsense Auto APAP since 2021. Same Bleep mask, AHI's close to 0 and I was sleeping okay, everything fine.
2. Feb 9, 2023-the motor was going out, my DME gave me a new replacement machine; same model.
3. Since Feb 9th, haven't slept much at all. I'm about to lose my mind.

I stopped using Oscar after Sept 2022 as I was sleeping okay.
I started using it again when I got this new machine Feb 9, 2023.

Last night, humidity in my bedroom was 48%.
I lowered the min and max pressure settings and I was gasping for air, so I increased them back to what they were.
I had the tube temp on 79 and humidity level set to 5, ended up woken up with rainout.
I changed the tube temp to Auto and humidity to 4. That fixed the rainout.
Used Hyperfix mouth tape last night. Woke up dry mouth, dry nose, headache and dry and sore throat.

The night before last, I didn't run my humidifier and slept better, but still AHI's.

thanks.