ResMed AirSense 10 making noise
- JerseyBiker
- Posts: 15
- Joined: Thu May 01, 2014 5:25 am
- Location: South Jersey Shore
ResMed AirSense 10 making noise
I have owned my ResMed AirSense 10 since July 2020, that's almost 2 and a half years. It's my second, and by far the best CPAP machine I owned. Up until 2 or 3 weeks ago, It was practically silent. Now it makes for the lack of a better term, a whistling noise. It's constant but more pronounced with each inhale. Other than the noise, it still works great. My concern is that it will get worst with time, and completely break down. Nothing seems loose. I only use distilled water. Has anyone else experienced anything like this with their ResMed AirSense 10? If so what was your solution? Thanks
_________________
| Machine: AirSense 10 AutoSet with Heated Humidifer + P10 Nasal Pillow Mask Bundle |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
- PeaceLoveAndPizza
- Posts: 2
- Joined: Mon Dec 26, 2022 8:43 am
Re: ResMed AirSense 10 making noise
I had something similar happen recently and it was due to the water chamber not being seated properly.JerseyBiker wrote: ↑Thu Jan 05, 2023 2:56 pmI have owned my ResMed AirSense 10 since July 2020, that's almost 2 and a half years. It's my second, and by far the best CPAP machine I owned. Up until 2 or 3 weeks ago, It was practically silent. Now it makes for the lack of a better term, a whistling noise. It's constant but more pronounced with each inhale. Other than the noise, it still works great. My concern is that it will get worst with time, and completely break down. Nothing seems loose. I only use distilled water. Has anyone else experienced anything like this with their ResMed AirSense 10? If so what was your solution? Thanks
_________________
| Machine: AirStart™ 10 Auto CPAP with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: 5-25 PS 4 |
Re: ResMed AirSense 10 making noise
Sounds like one of two things to me:JerseyBiker wrote: ↑Thu Jan 05, 2023 2:56 pmI have owned my ResMed AirSense 10 since July 2020, that's almost 2 and a half years. It's my second, and by far the best CPAP machine I owned. Up until 2 or 3 weeks ago, It was practically silent. Now it makes for the lack of a better term, a whistling noise. It's constant but more pronounced with each inhale.
1) The humidifier tank may either have a bad gasket, not be closed properly, or not be fully inserted into the machine.
2) The air filter may have gotten partially sucked into the air intake of the blower unit.
Check those two things and see if there is anything that doesn't look quite right. If the noise continues after making sure the problem is not these two things, let us know.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls