Dreamstation 2 preheat & SD card question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
shimon
Posts: 59
Joined: Mon Apr 01, 2013 3:58 pm
Location: Skagit County, WA

Dreamstation 2 preheat & SD card question

Post by shimon » Fri Dec 30, 2022 5:19 pm

Does the Dreamstation 2 have a preheat function for the humidifier so you can warm the water before starting the machine? I usually turn on preheat about 10 minutes before using the machine so the air isn't so cold but I can't figure out how.

Also I'm switching the SD card from old machine (System One REMstar Auto A-Flex) to Dreamstation 2...do I just put the old card in new machine? I've saved all my data from old machine in Oscar. Does the new machine overwrite old data? Do I need to format it first? Should I just use a new card?

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Dreamstation 2 preheat & SD card question

Post by chunkyfrog » Fri Dec 30, 2022 5:37 pm

I know virtually nothing about that company's machines.
And, for obvious reasons, I'm happy that I don't need to know.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
zonker
Posts: 11259
Joined: Fri Jun 19, 2015 4:36 pm

Re: Dreamstation 2 preheat & SD card question

Post by zonker » Fri Dec 30, 2022 6:16 pm

shimon wrote:
Fri Dec 30, 2022 5:19 pm
Does the Dreamstation 2 have a preheat function for the humidifier so you can warm the water before starting the machine?
according to a comment on youtube, it does not.
https://www.youtube.com/watch?v=bQ_qLsAc_x0
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Grumpy48
Posts: 421
Joined: Tue May 12, 2020 7:57 am

Re: Dreamstation 2 preheat & SD card question

Post by Grumpy48 » Fri Dec 30, 2022 6:46 pm

As noted in Zonker's post, I can confirm there is no capability to preheat the water. If you should happen to use tap water perhaps use warm water from the faucet to fill the tank just before using it(?)

If I were to use the SD card that had prior data on it, I think it would be wise to format (FAT32) the card before using it in the new machine. As well you could buy a new card (32gb or less) and just set the old card aside if you ever needed the old data for some reason.

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: Resmed AirSense 10 Auto with Humidifier

User avatar
Rock Star
Posts: 32
Joined: Wed May 02, 2012 10:31 pm

Re: Dreamstation 2 preheat & SD card question

Post by Rock Star » Fri Dec 30, 2022 6:49 pm

shimon wrote:
Fri Dec 30, 2022 5:19 pm
Does the Dreamstation 2 have a preheat function for the humidifier so you can warm the water before starting the machine?
When I first started I would heat the water in my microwave. As time went on I didn't feel the need to.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Not having trouble just joined to read

Grumpy48
Posts: 421
Joined: Tue May 12, 2020 7:57 am

Re: Dreamstation 2 preheat & SD card question

Post by Grumpy48 » Fri Dec 30, 2022 7:01 pm

Rock Star wrote:
Fri Dec 30, 2022 6:49 pm
shimon wrote:
Fri Dec 30, 2022 5:19 pm
Does the Dreamstation 2 have a preheat function for the humidifier so you can warm the water before starting the machine?
When I first started I would heat the water in my microwave. As time went on I didn't feel the need to.
Just do not try to heat the water with this method in the CPAP tub. The Microwave will not like the metal heating plate on the tubs bottom. :wink:

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: Resmed AirSense 10 Auto with Humidifier

User avatar
Rock Star
Posts: 32
Joined: Wed May 02, 2012 10:31 pm

Re: Dreamstation 2 preheat & SD card question

Post by Rock Star » Fri Dec 30, 2022 7:27 pm

Grumpy48 wrote:
Fri Dec 30, 2022 7:01 pm
Just do not try to heat the water with this method in the CPAP tub. The Microwave will not like the metal heating plate on the tubs bottom.
It didn't seem to bother my tub. ymmv

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Not having trouble just joined to read

User avatar
Rock Star
Posts: 32
Joined: Wed May 02, 2012 10:31 pm

Re: Dreamstation 2 preheat & SD card question

Post by Rock Star » Fri Dec 30, 2022 7:38 pm

You can always heat it up in your fav tea cup not hard

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Not having trouble just joined to read

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Dreamstation 2 preheat & SD card question

Post by chunkyfrog » Fri Dec 30, 2022 7:39 pm

Rock Star wrote:
Fri Dec 30, 2022 7:27 pm
Grumpy48 wrote:
Fri Dec 30, 2022 7:01 pm
Just do not try to heat the water with this method in the CPAP tub. The Microwave will not like the metal heating plate on the tubs bottom.
It didn't seem to bother my tub. ymmv
Even a tiny bit of metal can damage the magnetron tube.
Of course, if you have been wanting an excuse to upgrade to a newer oven . . .

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Grumpy48
Posts: 421
Joined: Tue May 12, 2020 7:57 am

Re: Dreamstation 2 preheat & SD card question

Post by Grumpy48 » Fri Dec 30, 2022 7:41 pm

Rock Star wrote:
Fri Dec 30, 2022 7:27 pm
Grumpy48 wrote:
Fri Dec 30, 2022 7:01 pm
Just do not try to heat the water with this method in the CPAP tub. The Microwave will not like the metal heating plate on the tubs bottom.
It didn't seem to bother my tub. ymmv
This video below mirrors my experience when a wire tie from a bread wrapper snuck its way in with some food. The wire tie actually generated enough heat to split a Corelle plate in half at the point of contact.. Maybe the type of metal makes a difference(?)

https://youtu.be/Kt9Bah0rXjQ

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: Resmed AirSense 10 Auto with Humidifier

User avatar
kbh209
Posts: 339
Joined: Wed Dec 30, 2015 9:12 pm

Re: Dreamstation 2 preheat & SD card question

Post by kbh209 » Fri Dec 30, 2022 11:07 pm

shimon wrote:
Fri Dec 30, 2022 5:19 pm


Also I'm switching the SD card from old machine (System One REMstar Auto A-Flex) to Dreamstation 2...do I just put the old card in new machine? I've saved all my data from old machine in Oscar. Does the new machine overwrite old data? Do I need to format it first? Should I just use a new card?
I moved my card from my DreamStation 1, to a temporary ResMed to a DreamStation 2. Occasionally the DS2 will have an issue not reading the card but that's about it.
Phillips Dreamstation 2 Auto CPAP with Humidifier **

User avatar
Okie bipap
Posts: 3566
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Dreamstation 2 preheat & SD card question

Post by Okie bipap » Sat Dec 31, 2022 9:47 am

chunkyfrog wrote:
Fri Dec 30, 2022 7:39 pm
Even a tiny bit of metal can damage the magnetron tube.
Of course, if you have been wanting an excuse to upgrade to a newer oven . . .
The first microwave oven I had was an Amana Radar Range and the inside of the oven was all stainless steel. Now, they say not to put any metal inside of the microwaves.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
Dog Slobber
Posts: 4180
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Dreamstation 2 preheat & SD card question

Post by Dog Slobber » Sat Dec 31, 2022 10:59 am

Okie bipap wrote:
Sat Dec 31, 2022 9:47 am
chunkyfrog wrote:
Fri Dec 30, 2022 7:39 pm
Even a tiny bit of metal can damage the magnetron tube.
Of course, if you have been wanting an excuse to upgrade to a newer oven . . .
The first microwave oven I had was an Amana Radar Range and the inside of the oven was all stainless steel. Now, they say not to put any metal inside of the microwaves.
Metal has always been potentially harmful and harmful to microwaves. But that doesn't mean they can't be made from metal. In fact microwaves are dependent on its sides reflecting back the microwaves to be absorbed by the food.

Thick, smooth metals are almost always OK to be but in a microwave because the waves don't penetrate deep into the metal and most of the waves bounce off (like the sides). Thin, crinkled, and sharp edged metals, like aluminum foil and services can overheat causing fires, spark and release electrons creating heated plasma.

It's easier to warn against placing metal in microwaves, instead of describing the conditions when metal can and cannot be used in a microwave.

BTW, putting distilled water in a microwave is actually very dangerous. Now, nobody here is heating distilled water beyond warm for CPAP use, but using a microwave to boil distilled water. The lack of impurities can cause water to heat beyond the boiling point, then when it's disturbed can cause it to vigorously boil over and explode out of it's container.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Dreamstation 2 preheat & SD card question

Post by chunkyfrog » Sat Dec 31, 2022 11:00 am

I fried a nice Panasonic--just saying . . .
Also had brussels sprouts explode.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
shimon
Posts: 59
Joined: Mon Apr 01, 2013 3:58 pm
Location: Skagit County, WA

Re: Dreamstation 2 preheat & SD card question

Post by shimon » Sat Dec 31, 2022 1:10 pm

zonker wrote:
Fri Dec 30, 2022 6:16 pm
according to a comment on youtube, it does not.
https://www.youtube.com/watch?v=bQ_qLsAc_x0
Grumpy48 wrote:
Fri Dec 30, 2022 6:46 pm
As noted in Zonker's post, I can confirm there is no capability to preheat the water.
kbh209 wrote:
Fri Dec 30, 2022 11:07 pm
I moved my card from my DreamStation 1, to a temporary ResMed to a DreamStation 2. Occasionally the DS2 will have an issue not reading the card but that's about it.
Thanks for these replies.