French newbie struggling with an Oscar reporting in French

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Albatros
Posts: 141
Joined: Sat Nov 05, 2022 11:33 am
Location: Le Vesinet Paris France

French newbie struggling with an Oscar reporting in French

Post by Albatros » Sat Dec 03, 2022 9:22 am

I am a NEWBIE with 6 weeks in CPAP therapy.
I read you assiduously.
I am French and I have a Resmed Airsense 11, with a persenal profile and parameters in French.
Big problem: my (first) ODCAR reporting is in French. The acronyms are in french, which is very annoying.
Is it possible to solve this problem ?

Thank you for your help.

ozij
Posts: 10474
Joined: Fri Mar 18, 2005 11:52 pm

Re: French newbie struggling with an Oscar reporting in French

Post by ozij » Sat Dec 03, 2022 9:30 am

Upload and image so we can see the acronym's you're seeing.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

MrPaapoo
Posts: 3
Joined: Sat Dec 03, 2022 9:35 am

Re: French newbie struggling with an Oscar reporting in French

Post by MrPaapoo » Sat Dec 03, 2022 9:43 am

a simple google search (oscar apnea change language) would have given you the requested answer. In the Aide menu, simply select "Changer de langue", then select Français.
Bonne chance. :)

Albatros
Posts: 141
Joined: Sat Nov 05, 2022 11:33 am
Location: Le Vesinet Paris France

Re: French newbie struggling with an Oscar reporting in French

Post by Albatros » Sat Dec 03, 2022 9:50 am

A few images in french:
Capture d'écran_20221203_164001.png
Capture d'écran_20221203_164001.png (44.93 KiB) Viewed 1450 times
Capture d'écran_20221203_164001.png
Capture d'écran_20221203_164001.png (44.93 KiB) Viewed 1450 times
Thank you
Attachments
Capture d'écran_20221203_164723.png
Capture d'écran_20221203_164723.png (33.43 KiB) Viewed 1450 times
Capture d'écran_20221203_164723.png
Capture d'écran_20221203_164723.png (33.43 KiB) Viewed 1450 times

Albatros
Posts: 141
Joined: Sat Nov 05, 2022 11:33 am
Location: Le Vesinet Paris France

Re: French newbie struggling with an Oscar reporting in French

Post by Albatros » Sat Dec 03, 2022 10:32 am

MrPaapoo wrote:
Sat Dec 03, 2022 9:43 am
a simple google search (oscar apnea change language) would have given you the requested answer. In the Aide menu, simply select "Changer de langue", then select Français.
Bonne chance. :)
You were right ! It was in the Aide menu.

A great thank you. You saved my day ...

User avatar
Miss Emerita
Posts: 3758
Joined: Sun Nov 04, 2018 8:07 pm

Re: French newbie struggling with an Oscar reporting in French

Post by Miss Emerita » Sat Dec 03, 2022 11:29 am

I realize you didn't ask about this, but I can't resist wondering why your pressure is capped at 11 and why you aren't using EPR. If you don't want to wander down the settings path, no worries!
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: French newbie struggling with an Oscar reporting in French

Post by palerider » Sat Dec 03, 2022 4:59 pm

Albatros wrote:
Sat Dec 03, 2022 9:22 am
I am a NEWBIE with 6 weeks in CPAP therapy.
I read you assiduously.
I am French and I have a Resmed Airsense 11, with a persenal profile and parameters in French.
Big problem: my (first) ODCAR reporting is in French. The acronyms are in french, which is very annoying.
Is it possible to solve this problem ?
I'm a little confused, the way I read your question, it implied to me that your profile was *already* in French, and the acronyms were the confusing area.

Did you get your questions answered?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: French newbie struggling with an Oscar reporting in French

Post by palerider » Sat Dec 03, 2022 5:00 pm

Miss Emerita wrote:
Sat Dec 03, 2022 11:29 am
I realize you didn't ask about this, but I can't resist wondering why your pressure is capped at 11 and why you aren't using EPR. If you don't want to wander down the settings path, no worries!
I'm guessing there are incompetent doctors/sleep techs even in France. :D

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Albatros
Posts: 141
Joined: Sat Nov 05, 2022 11:33 am
Location: Le Vesinet Paris France

Re: French newbie struggling with an Oscar reporting in French

Post by Albatros » Sun Dec 04, 2022 2:56 am

Miss Emerita wrote:
Sat Dec 03, 2022 11:29 am
by Miss Emerita » Sat Dec 03, 2022 6:29 pm

I realize you didn't ask about this, but I can't resist wondering why your pressure is capped at 11 and why you aren't using EPR. If you don't want to wander down the settings path, no worries!

When I look at my 1st Oscar report:
-I notice on my graph Flow limitation, that (outside of apnea zones) my % is low, 0.25 or less, permanently. Does this mean that I have a structural problem of blocked breathing?

-Concerning the pressure level of 11. I put it up (vs. my doctor's 10) a bit randomly. It turns out that the pressure is stuck near 11. Should I change? Increase it,

-I note that I do not have a LEAKS graph. Is thereba reason for that ?


THANK YOU
Attachments
Capture d'écran_20221204_095037.png
Capture d'écran_20221204_095037.png (40.29 KiB) Viewed 1383 times
Capture d'écran_20221204_095100.png
Capture d'écran_20221204_095100.png (36.01 KiB) Viewed 1383 times

Albatros
Posts: 141
Joined: Sat Nov 05, 2022 11:33 am
Location: Le Vesinet Paris France

Re: French newbie struggling with an Oscar reporting in French

Post by Albatros » Sun Dec 04, 2022 8:25 am

I stopped EPR assuming that it was useless for me, as I do not have a problem with the air coming from the machine.
A mistake ?

User avatar
Dog Slobber
Posts: 4261
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: French newbie struggling with an Oscar reporting in French

Post by Dog Slobber » Sun Dec 04, 2022 9:43 am

Albatros wrote:
Sun Dec 04, 2022 8:25 am
I stopped EPR assuming that it was useless for me, as I do not have a problem with the air coming from the machine.
A mistake ?
No. For many, EPR is purely a comfort issue. Some claim, that EPR can help reduce Flow Limitations. Personally, I think that its value in reducing Flow Limitations is over stated.
Battery Backup: EcoFlow Delta 2

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: French newbie struggling with an Oscar reporting in French

Post by D.H. » Sun Dec 04, 2022 11:26 am

If you're more comfortable with English (or the English acronyms), you have the option to reinstall OSCAR in English. There are actually separate options for US English and UK English, but these differences are no big deal, at least not for native English speakers.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Dog Slobber
Posts: 4261
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: French newbie struggling with an Oscar reporting in French

Post by Dog Slobber » Sun Dec 04, 2022 11:40 am

D.H. wrote:
Sun Dec 04, 2022 11:26 am
If you're more comfortable with English (or the English acronyms), you have the option to reinstall OSCAR in English. There are actually separate options for US English and UK English, but these differences are no big deal, at least not for native English speakers.

DH's Resume wrote:...
Skills:
  • All though I'm a native US English speaker, and haven't been formally trained in UK English. Using UK English is no big deal for me.
...
Battery Backup: EcoFlow Delta 2

Albatros
Posts: 141
Joined: Sat Nov 05, 2022 11:33 am
Location: Le Vesinet Paris France

Re: French newbie struggling with an Oscar reporting in French

Post by Albatros » Sun Dec 04, 2022 11:47 am

I was advised to go in the Aide of Oscar. And I just had to click on "Language change". And it was in english !

User avatar
Miss Emerita
Posts: 3758
Joined: Sun Nov 04, 2018 8:07 pm

Re: French newbie struggling with an Oscar reporting in French

Post by Miss Emerita » Sun Dec 04, 2022 12:15 pm

Albatros, click on the button in the lower right corner of the daily chart to make sure the leak graph is checked. Looking at the numbers in the right-hand panel, you probably don't have a leak problem, but you might as well have the data visualized for you.

I would recommend setting EPR to 3. As Dog Slobber suggests, that may make using your machine more comfortable for you, and it will probably reduce your flow limitations and thus the pressure swings that you're seeing. Also as DS suggests, reducing FLs might not make much difference to the quality of your sleep -- but then again it might.

If you decide to add EPR, please raise your minimum by a corresponding amount. This will keep your exhalation pressure steady.

And as an experiment, try raising your maximum to 20, just to see what happens. The machine won't go there unless it "sees" a need to. If that sounds too radical, try raising it to, say, 15.

Your events show some clustering, which can sometimes mean a positional problem: either sleeping on your back, or sleeping with your head tucked down toward your chest. If you use a high pillow, or several pillows, try using a single, fairly low and firm pillow to see how that works.
Oscar software is available at https://www.sleepfiles.com/OSCAR/