Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by palerider » Wed Oct 05, 2022 6:30 am

I changed the filter (using another hypoallergenic thin filter) in my VAuto a couple days ago, and the machine started acting weird.

It wouldn't shut off like it normally would, the fan would run at a lower speed, and I could feel some vibrating in the machine. Additionally, it wouldn't deliver air like normal after it got into that strange new behavior pattern, when I'd inhale, I could feel and hear a vibration in the airflow. A couple times it was making the wheezing noises that so many have complained about.

What I finally figured out is that the new filter was getting sucked down onto the opening, so it was causing a lot more airflow resistance. I popped open the filter cover, while it was acting weird, noticed the filter was sucked into the opening, pulled the filter out, and the airflow picked up to where it always was before when I'd take the mask off, and then smart start shut off the machine, just like it is supposed to do.

playing with it more, I found that if I put the filter in the square opening of the filter cover door, it would get sucked into the intake, If however I sat the filter in the opening of the housing, and then closed the door, it would pinch the bottom edge of the filter in place, and the filter would stay, and the machine would work fine.

I guess pictures are in order.

How I thought I should put in a new filter:
Image

What happened:
Image

What worked:
Image

There needs to be a little more support around the actual air inlet when using just the thin hypoallergenic filters, but pinching the filter in the bottom of the door seems to be working for the time being :)

There needs to be a little more support around the inner opening,

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Conrad
Posts: 670
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by Conrad » Wed Oct 05, 2022 9:09 am

I have the same machine that you do. I've never had an issue with the filter like you describe. I always place the filter on the door of the compartment when replacing it and have never had an issue.
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by raisedfist » Wed Oct 05, 2022 9:23 am

I read a post about this recently...the hypoallergenic filter getting sucked in. I find my humidifier chamber gets stuck sometimes where I have to yank harder than I'd like to get it out to clean it.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by chunkyfrog » Wed Oct 05, 2022 9:33 am

Is the filter OEM, or third party?
I don't know what mine are, but never have that problem.
Knock on wood . . .
A brilliant solution, nonetheless!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Dog Slobber
Posts: 4244
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by Dog Slobber » Wed Oct 05, 2022 9:47 am

Thanks for this, we'll add it to the list of things to check for when noisey or odd behaving machines.

Given the HA filter is thinner, more pliable and less porous it would be more prone to getting sucked in. Well done.
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by palerider » Wed Oct 05, 2022 11:21 am

chunkyfrog wrote:
Wed Oct 05, 2022 9:33 am
Is the filter OEM, or third party?
I don't know what mine are, but never have that problem.
Knock on wood . . .
A brilliant solution, nonetheless!
Might be third party. The weird thing is that the previous one from the same package had no problem.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by palerider » Wed Oct 05, 2022 11:24 am

Conrad wrote:
Wed Oct 05, 2022 9:09 am
I have the same machine that you do. I've never had an issue with the filter like you describe. I always place the filter on the door of the compartment when replacing it and have never had an issue.
The point, as it were, wasn't that this WOULD happen, but that it *CAN* happen, and if the machine starts making unexpected noises (a complaint that gets heard, no pun intended, check the filter.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11336
Joined: Fri Jun 19, 2015 4:36 pm

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by zonker » Wed Oct 05, 2022 12:11 pm

thanks for the tip. i don't use hypoallergenic filters and usually put the filter against the port.

so far so good.

now, if i can just remember the tip to pass on when someone complains of noise.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
lazarus
Posts: 1419
Joined: Sat Mar 12, 2022 3:21 pm

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by lazarus » Wed Oct 05, 2022 12:30 pm

PR:

I thought that you were once in the double-filtering camp.

No longer so?

viewtopic/t157518/Doubling-up-on-ResMed ... -idea.html
Last edited by lazarus on Wed Oct 05, 2022 12:37 pm, edited 1 time in total.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by palerider » Wed Oct 05, 2022 12:36 pm

lazarus wrote:
Wed Oct 05, 2022 12:30 pm
PR:

I thought that you were once in the double-filtering camp.

No longer so?
I usually do, but didn't have a coarse filter handy the last two times I changed.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
lazarus
Posts: 1419
Joined: Sat Mar 12, 2022 3:21 pm

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by lazarus » Wed Oct 05, 2022 12:39 pm

palerider wrote:
Wed Oct 05, 2022 12:36 pm
lazarus wrote:
Wed Oct 05, 2022 12:30 pm
PR:

I thought that you were once in the double-filtering camp.

No longer so?
I usually do, but didn't have a coarse filter handy the last two times I changed.
I would make a joke about filtering out coarseness, but I'm no longer paid by the joke here.
Last edited by lazarus on Wed Oct 05, 2022 12:40 pm, edited 1 time in total.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by palerider » Wed Oct 05, 2022 12:40 pm

lazarus wrote:
Wed Oct 05, 2022 12:39 pm
palerider wrote:
Wed Oct 05, 2022 12:36 pm
lazarus wrote:
Wed Oct 05, 2022 12:30 pm
PR:

I thought that you were once in the double-filtering camp.

No longer so?
I usually do, but didn't have a coarse filter handy the last two times I changed.
I would make a joke about a filtering out coarseness, but I'm no longer paid by the joke here.
Someone get this guy back on the payroll!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by chunkyfrog » Wed Oct 05, 2022 2:21 pm

Yeah, we need the yuks; and we have no taste . . .

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
lazarus
Posts: 1419
Joined: Sat Mar 12, 2022 3:21 pm

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by lazarus » Wed Oct 05, 2022 3:03 pm

palerider wrote:
Wed Oct 05, 2022 12:40 pm
payroll
I was told this year's entire budget went toward finding Zonker some new gifs.
chunkyfrog wrote:
Wed Oct 05, 2022 2:21 pm
no taste . . .
StarKist doesn't want frogs with good taste . . .

All in all, a very helpful thread on possible problems with filters.

In my case, I never change them often enough.

Something else I'm waay too lazy to put a reminder for inside my phone's calendar.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Strange behavior with Aircurve 10 VAuto, noise/won't shut off automatically

Post by chunkyfrog » Wed Oct 05, 2022 3:58 pm

My strategy is my spiral notebook.
I draw up a manual spreadsheet, with 21 lines for 21 days. (fits the page)
Day of week/month is followed by hours, AHI, pressure, leak, CI, and a few others,
as requested by my doctor.
At the end of each page, I swap pillows and filters.
Medicare allows for 2 a month, so I am a bit ahead.
Good for when I run out of credit--already out of money . . .

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her