no detail data in Oscar

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Pinhead
Posts: 86
Joined: Sat Apr 19, 2008 5:14 am
Location: The Villages Fl

no detail data in Oscar

Post by Pinhead » Sat Sep 17, 2022 12:14 pm

The truth I don't up load my data very often, but noted my AHI is geting higher then it has been. I look at the screen and if things look ok I dont go much further, in fact I haven't downloaded information in almost a year. I have been talking to others on Nextdoor and decided I should look at my data closer. I downloaded the current software and data today. Im getting only summary data and no detail. Anyone know what I'm doing wrong? I 've never had this issue before.

Pinhead
Posts: 86
Joined: Sat Apr 19, 2008 5:14 am
Location: The Villages Fl

Re: no detail data in Oscar

Post by Pinhead » Sat Sep 17, 2022 12:18 pm

It says data turned off. How do I turn it back on

clownbell
Posts: 369
Joined: Fri Apr 02, 2021 8:46 pm

Re: no detail data in Oscar

Post by clownbell » Sat Sep 17, 2022 12:24 pm

Is that msg on the machi e? Or is it on OSCAR?
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask

Pinhead
Posts: 86
Joined: Sat Apr 19, 2008 5:14 am
Location: The Villages Fl

Re: no detail data in Oscar

Post by Pinhead » Sat Sep 17, 2022 2:34 pm

The message is in Oscar. I ll see if i can get a screen shot

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: no detail data in Oscar

Post by Pugsy » Sat Sep 17, 2022 2:42 pm

Was the SD card in the machine during the night when the machine was in use?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Pinhead
Posts: 86
Joined: Sat Apr 19, 2008 5:14 am
Location: The Villages Fl

Re: no detail data in Oscar

Post by Pinhead » Sat Sep 17, 2022 2:52 pm

I hope this works , I can now see today's information but not Prior days

Pinhead
Posts: 86
Joined: Sat Apr 19, 2008 5:14 am
Location: The Villages Fl

Re: no detail data in Oscar

Post by Pinhead » Sat Sep 17, 2022 2:53 pm

Yes Pugsy the card was in the machine

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: no detail data in Oscar

Post by Pugsy » Sat Sep 17, 2022 3:03 pm

If you aren't seeing detailed data on prior days then maybe the card was in the machine but not seated fully and that's why no details are available.
I have done it myself...insert the SD card and not confirm that the machine recognizes it and I guess I didn't push it in totally or bumped it and caused it to back out just a little. Not enough to notice but out enough that the machine can't write to the SD card.

If you have details now with same card we can't blame the card.

Just make sure when you put the SD card back in your machine that the machine recognizes it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Pinhead
Posts: 86
Joined: Sat Apr 19, 2008 5:14 am
Location: The Villages Fl

Re: no detail data in Oscar

Post by Pinhead » Sat Sep 17, 2022 3:17 pm

Pugsy I just signe up for Imgur if see if this works. https://imgur.com/EmG3I8m what im concerned with is most of my events appear to be centrals, last night about 5 events per hour almost all are clear airway
.https://imgur.com/EmG3I8m

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: no detail data in Oscar

Post by palerider » Sat Sep 17, 2022 3:35 pm

Pinhead wrote:
Sat Sep 17, 2022 3:17 pm
Pugsy I just signe up for Imgur if see if this works. https://imgur.com/EmG3I8m what im concerned with is most of my events appear to be centrals, last night about 5 events per hour almost all are clear airway
.https://imgur.com/EmG3I8m
Until your centrals are approaching 10/hr then just ignore them.

a few centrals happen here and there, but they're nothing to be concerned about unless you're having a bunch of 'em all together.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Pinhead
Posts: 86
Joined: Sat Apr 19, 2008 5:14 am
Location: The Villages Fl

Re: no detail data in Oscar

Post by Pinhead » Sun Sep 18, 2022 5:32 am

Thanks I never know when to start worrying about them

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: no detail data in Oscar

Post by Pugsy » Sun Sep 18, 2022 6:27 am

There's a lot of "it depends" when it comes to when we might be a little worried about central apneas.
It depends on how many and it depends on when they got flagged and it depends on the question were we asleep or not.

If not asleep...doesn't matter how many we see other than it reflects poor sleep itself. A symptom of a problem more than a cause of a problem.

Here's one of mine. Initially a bit ugly looking but this night I had a major visit from the insomnia monster and I simply couldn't fall asleep.
All these "centrals" are nothing more than awake pauses in breathing and the machine thinks that they are central apneas. I never went to sleep until after the second cluster of "centrals".

For help in figuring out awake vs asleep flagged breathing you can watch the videos here
http://freecpapadvice.com/sleepyhead-free-software

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: no detail data in Oscar

Post by lynninnj » Sun Sep 18, 2022 10:17 am

Pinhead wrote:
Sat Sep 17, 2022 2:52 pm
I hope this works , I can now see today's information but not Prior days
I have done this a few times myself. I didn't have the card fitted in the slot properly. Now I pay close attention to the screen when I load it up and watch for blue lights flashing near the card slot when I put it in. It takes time for the SD error message to show up.

I also had a POS walmart card in there that broke after a few weeks, so getting a good SD card is essential if you will be taking it out a fair bit to follow along. The POS card just stopped working and the blue light never blinked when I put it in and machine stopped recognizing it.

HTH

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: no detail data in Oscar

Post by lynninnj » Sun Sep 18, 2022 10:19 am

palerider wrote:
Sat Sep 17, 2022 3:35 pm
Pinhead wrote:
Sat Sep 17, 2022 3:17 pm
Pugsy I just signe up for Imgur if see if this works. https://imgur.com/EmG3I8m what im concerned with is most of my events appear to be centrals, last night about 5 events per hour almost all are clear airway
.https://imgur.com/EmG3I8m
Until your centrals are approaching 10/hr then just ignore them.

a few centrals happen here and there, but they're nothing to be concerned about unless you're having a bunch of 'em all together.
Oh, you mean like this?

https://i.imgur.com/7j6khri.png

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: no detail data in Oscar

Post by Pugsy » Sun Sep 18, 2022 10:25 am

lynninnj wrote:
Sun Sep 18, 2022 10:19 am
palerider wrote:
Sat Sep 17, 2022 3:35 pm
Pinhead wrote:
Sat Sep 17, 2022 3:17 pm
Pugsy I just signe up for Imgur if see if this works. https://imgur.com/EmG3I8m what im concerned with is most of my events appear to be centrals, last night about 5 events per hour almost all are clear airway
.https://imgur.com/EmG3I8m
Until your centrals are approaching 10/hr then just ignore them.

a few centrals happen here and there, but they're nothing to be concerned about unless you're having a bunch of 'em all together.
Oh, you mean like this?

https://i.imgur.com/7j6khri.png
One thing missing on the report from the other person you posted their report...
Anyone notice that the machine didn't seem to flag that flow rate pattern as CSR????? Unless the snippet just doesn't show it or it's turned off....and why I don't like snippets like that.

And one report all by itself like that isn't the end of the world either....doctors look for repetition or trends or patterns before they push the panic button....if they push at all.

One or two nights with ugly central index isn't the end of the world. Step back...take a deep breath and look a the big picture too.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.