Help with testing pressure at home
-
Country4ever
- Posts: 1373
- Joined: Wed Oct 31, 2007 6:22 pm
Help with testing pressure at home
Hi all. I've tested this way before, but forgot about it until I read Tomjax's directions from awhile back. I tied the end of the hose to a ruler, making sure the water level in the container was about 6". Then I put the tube to the bottom of the container and slowly raised it up. Just a warning.......when it starts to bubble, it's like a volcanic eruption....so be careful of furniture, etc., around you. Okay.........so my setting on the Resmed 9 machine is 9.6. It started "erupting" at 3 and 1/2", which is 8.9cm.
Beforehand, I did take off the little tube that goes from the big tube to the mask. Would that have changed the pressure? Is 8.9 close enough to 9.6? (I adjusted it in the past to this setting...so that's why it's not a round number.) Any suggestions? Thanks!
Beforehand, I did take off the little tube that goes from the big tube to the mask. Would that have changed the pressure? Is 8.9 close enough to 9.6? (I adjusted it in the past to this setting...so that's why it's not a round number.) Any suggestions? Thanks!
Re: Help with testing pressure at home
I have taken my machine to a local DME and she tested it with a hydrometer in about 3 minutes...no charge.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: Help with testing pressure at home
I hope not.
A manometer would work better.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: Help with testing pressure at home
A man-o-meter?
-thinking back to prior thread about hero men in thights-
nevermind
I never even thought about testing. Part of me thinks why would I care as long as it works.
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
| Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
Re: Help with testing pressure at home
Yeah.
That thing you (guys) were talking about in the other thread:
babydinosnoreless wrote: ↑Fri Sep 09, 2022 10:02 amI know for some reason you guys feel the need to whip out your measuring tapes ...
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: Help with testing pressure at home
You'll get an approximation or pressure by doing it that way, but to *properly* measure the pressure, the machine should have a vent going on. When I test pressure, I put an oxygen adapter in the hose and attach the manometer to that, and use the mask vent to provide normal venting while pressure is measured.Country4ever wrote: ↑Thu Sep 15, 2022 8:09 amHi all. I've tested this way before, but forgot about it until I read Tomjax's directions from awhile back. I tied the end of the hose to a ruler, making sure the water level in the container was about 6". Then I put the tube to the bottom of the container and slowly raised it up. Just a warning.......when it starts to bubble, it's like a volcanic eruption....so be careful of furniture, etc., around you. Okay.........so my setting on the Resmed 9 machine is 9.6. It started "erupting" at 3 and 1/2", which is 8.9cm.
Beforehand, I did take off the little tube that goes from the big tube to the mask. Would that have changed the pressure? Is 8.9 close enough to 9.6? (I adjusted it in the past to this setting...so that's why it's not a round number.) Any suggestions? Thanks!
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Help with testing pressure at home
OK goofy question but I have to ask.palerider wrote: ↑Thu Sep 15, 2022 2:31 pmYou'll get an approximation or pressure by doing it that way, but to *properly* measure the pressure, the machine should have a vent going on. When I test pressure, I put an oxygen adapter in the hose and attach the manometer to that, and use the mask vent to provide normal venting while pressure is measured.Country4ever wrote: ↑Thu Sep 15, 2022 8:09 amHi all. I've tested this way before, but forgot about it until I read Tomjax's directions from awhile back. I tied the end of the hose to a ruler, making sure the water level in the container was about 6". Then I put the tube to the bottom of the container and slowly raised it up. Just a warning.......when it starts to bubble, it's like a volcanic eruption....so be careful of furniture, etc., around you. Okay.........so my setting on the Resmed 9 machine is 9.6. It started "erupting" at 3 and 1/2", which is 8.9cm.
Beforehand, I did take off the little tube that goes from the big tube to the mask. Would that have changed the pressure? Is 8.9 close enough to 9.6? (I adjusted it in the past to this setting...so that's why it's not a round number.) Any suggestions? Thanks!
What do you do with this information once you have it? What is the relevance? I’m all for gathering data but I just didn’t know what you did with it.
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
| Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
Re: Help with testing pressure at home
Easy peasy lemon squeezy…
http://www.apneaboard.com/forums/Threa ... ew-dollars
Maybe not that easy, but it will work.
http://www.apneaboard.com/forums/Threa ... ew-dollars
Maybe not that easy, but it will work.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: 5-25 PS 4 |
Sometimes it is the very people who no one imagines anything of who do the things that no one imagines
- babydinosnoreless
- Posts: 2375
- Joined: Fri Nov 02, 2018 2:53 pm
Re: Help with testing pressure at home
Sounds kinky.Rubicon wrote: ↑Thu Sep 15, 2022 12:59 pmYeah.
That thing you (guys) were talking about in the other thread:
babydinosnoreless wrote: ↑Fri Sep 09, 2022 10:02 amI know for some reason you guys feel the need to whip out your measuring tapes ...
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
Re: Help with testing pressure at home
I bought a manometer from Amazon. I had other reasons to get one (pellet stove draft), but it works well for checking CPAP pressure.
https://www.amazon.com/gp/product/B0168 ... UTF8&psc=1
Had to cobble up fittings for the hose connection. Tried a range of several pressure settings on the CPAP's (DS2 and AS10) and the reads on the manometer matched very closely to what was set on the CPAP's.
https://www.amazon.com/gp/product/B0168 ... UTF8&psc=1
Had to cobble up fittings for the hose connection. Tried a range of several pressure settings on the CPAP's (DS2 and AS10) and the reads on the manometer matched very closely to what was set on the CPAP's.
Re: Help with testing pressure at home
The struggle is real!babydinosnoreless wrote: ↑Thu Sep 15, 2022 8:07 pmI know for some reason you guys feel the need to whip out your measuring tapes ...


_________________
| Machine: AirSense 11 Autoset |
| Mask: Evora Full Face Mask |
| Additional Comments: I CAN BREATHE! |
Angry much?
Re: Help with testing pressure at home
ROFLMAOBrad S wrote: ↑Fri Sep 16, 2022 7:43 amThe struggle is real!babydinosnoreless wrote: ↑Thu Sep 15, 2022 8:07 pmI know for some reason you guys feel the need to whip out your measuring tapes ...
![]()
Maybe if I ask someone else on the thread who is courteous and helpful I will get a reply.
I am curious to know what one does with the information regarding the pressure testing? Isn't it all sort of about just making adjustments to see what works? Or is this more particular to those on a CPAP rather than an APAP?
Thanks for any helpful responses.
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
| Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
-
Iamstumped
- Posts: 157
- Joined: Tue May 17, 2022 2:56 am
- Location: Kansas City, Mo.
Re: Help with testing pressure at home
I think that it is to just make sure that your machine is putting out the correct pressure, that it should be putting out.
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Re: Help with testing pressure at home
Thanks for taking the time. I can certainly see why that is important for someone on a CPAP, for sure. But with something like i have, with the autoset and the APAP, I can't see the significance.Iamstumped wrote: ↑Fri Sep 16, 2022 11:00 amI think that it is to just make sure that your machine is putting out the correct pressure, that it should be putting out.
That's why I ask, so I can learn.
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
| Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
Re: Help with testing pressure at home
It's crazy and frustrating that you have to qualify your question to get an appropriate response.
This Man-O-Meter thing is really interesting to me. I am really interested in the pressure of things- I am a guy
I am with you though, the pressure of the CPAP might be nice to know, but like you say, if the machine is doing what it is supposed to, then the actual pressure measurement isn't really that important.
_________________
| Machine: AirSense 11 Autoset |
| Mask: Evora Full Face Mask |
| Additional Comments: I CAN BREATHE! |
Angry much?