Help with testing pressure at home

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Country4ever
Posts: 1373
Joined: Wed Oct 31, 2007 6:22 pm

Help with testing pressure at home

Post by Country4ever » Thu Sep 15, 2022 8:09 am

Hi all. I've tested this way before, but forgot about it until I read Tomjax's directions from awhile back. I tied the end of the hose to a ruler, making sure the water level in the container was about 6". Then I put the tube to the bottom of the container and slowly raised it up. Just a warning.......when it starts to bubble, it's like a volcanic eruption....so be careful of furniture, etc., around you. Okay.........so my setting on the Resmed 9 machine is 9.6. It started "erupting" at 3 and 1/2", which is 8.9cm.
Beforehand, I did take off the little tube that goes from the big tube to the mask. Would that have changed the pressure? Is 8.9 close enough to 9.6? (I adjusted it in the past to this setting...so that's why it's not a round number.) Any suggestions? Thanks!

_________________
MachineMask
Additional Comments: Resmed HumidAir 11 for humidifier

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Help with testing pressure at home

Post by LSAT » Thu Sep 15, 2022 8:47 am

I have taken my machine to a local DME and she tested it with a hydrometer in about 3 minutes...no charge.

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Help with testing pressure at home

Post by Rubicon » Thu Sep 15, 2022 9:11 am

LSAT wrote:
Thu Sep 15, 2022 8:47 am
I have taken my machine to a local DME and she tested it with a hydrometer...
I hope not.

A manometer would work better.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: Help with testing pressure at home

Post by lynninnj » Thu Sep 15, 2022 11:50 am

Rubicon wrote:
Thu Sep 15, 2022 9:11 am
LSAT wrote:
Thu Sep 15, 2022 8:47 am
I have taken my machine to a local DME and she tested it with a hydrometer...
I hope not.

A manometer would work better.
A man-o-meter?

-thinking back to prior thread about hero men in thights-

nevermind

I never even thought about testing. Part of me thinks why would I care as long as it works.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Help with testing pressure at home

Post by Rubicon » Thu Sep 15, 2022 12:59 pm

lynninnj wrote:
Thu Sep 15, 2022 11:50 am

A man-o-meter?
Yeah.

That thing you (guys) were talking about in the other thread:
babydinosnoreless wrote:
Fri Sep 09, 2022 10:02 am
I know for some reason you guys feel the need to whip out your measuring tapes ...
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Help with testing pressure at home

Post by palerider » Thu Sep 15, 2022 2:31 pm

Country4ever wrote:
Thu Sep 15, 2022 8:09 am
Hi all. I've tested this way before, but forgot about it until I read Tomjax's directions from awhile back. I tied the end of the hose to a ruler, making sure the water level in the container was about 6". Then I put the tube to the bottom of the container and slowly raised it up. Just a warning.......when it starts to bubble, it's like a volcanic eruption....so be careful of furniture, etc., around you. Okay.........so my setting on the Resmed 9 machine is 9.6. It started "erupting" at 3 and 1/2", which is 8.9cm.
Beforehand, I did take off the little tube that goes from the big tube to the mask. Would that have changed the pressure? Is 8.9 close enough to 9.6? (I adjusted it in the past to this setting...so that's why it's not a round number.) Any suggestions? Thanks!
You'll get an approximation or pressure by doing it that way, but to *properly* measure the pressure, the machine should have a vent going on. When I test pressure, I put an oxygen adapter in the hose and attach the manometer to that, and use the mask vent to provide normal venting while pressure is measured.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: Help with testing pressure at home

Post by lynninnj » Thu Sep 15, 2022 3:06 pm

palerider wrote:
Thu Sep 15, 2022 2:31 pm
Country4ever wrote:
Thu Sep 15, 2022 8:09 am
Hi all. I've tested this way before, but forgot about it until I read Tomjax's directions from awhile back. I tied the end of the hose to a ruler, making sure the water level in the container was about 6". Then I put the tube to the bottom of the container and slowly raised it up. Just a warning.......when it starts to bubble, it's like a volcanic eruption....so be careful of furniture, etc., around you. Okay.........so my setting on the Resmed 9 machine is 9.6. It started "erupting" at 3 and 1/2", which is 8.9cm.
Beforehand, I did take off the little tube that goes from the big tube to the mask. Would that have changed the pressure? Is 8.9 close enough to 9.6? (I adjusted it in the past to this setting...so that's why it's not a round number.) Any suggestions? Thanks!
You'll get an approximation or pressure by doing it that way, but to *properly* measure the pressure, the machine should have a vent going on. When I test pressure, I put an oxygen adapter in the hose and attach the manometer to that, and use the mask vent to provide normal venting while pressure is measured.
OK goofy question but I have to ask.

What do you do with this information once you have it? What is the relevance? I’m all for gathering data but I just didn’t know what you did with it.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

User avatar
Jlfinkels
Posts: 468
Joined: Sun Jun 20, 2021 3:51 pm

Re: Help with testing pressure at home

Post by Jlfinkels » Thu Sep 15, 2022 5:17 pm

Easy peasy lemon squeezy…

http://www.apneaboard.com/forums/Threa ... ew-dollars

Maybe not that easy, but it will work.
Sometimes it is the very people who no one imagines anything of who do the things that no one imagines

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: Help with testing pressure at home

Post by babydinosnoreless » Thu Sep 15, 2022 8:07 pm

Rubicon wrote:
Thu Sep 15, 2022 12:59 pm
lynninnj wrote:
Thu Sep 15, 2022 11:50 am

A man-o-meter?
Yeah.

That thing you (guys) were talking about in the other thread:
babydinosnoreless wrote:
Fri Sep 09, 2022 10:02 am
I know for some reason you guys feel the need to whip out your measuring tapes ...
Sounds kinky. :lol: :lol:

Grumpy48
Posts: 462
Joined: Tue May 12, 2020 7:57 am

Re: Help with testing pressure at home

Post by Grumpy48 » Fri Sep 16, 2022 6:35 am

I bought a manometer from Amazon. I had other reasons to get one (pellet stove draft), but it works well for checking CPAP pressure.

https://www.amazon.com/gp/product/B0168 ... UTF8&psc=1

Had to cobble up fittings for the hose connection. Tried a range of several pressure settings on the CPAP's (DS2 and AS10) and the reads on the manometer matched very closely to what was set on the CPAP's.

User avatar
Brad S
Posts: 135
Joined: Mon Jul 25, 2022 10:48 am
Location: Peoples Republik of Kolorado

Re: Help with testing pressure at home

Post by Brad S » Fri Sep 16, 2022 7:43 am

babydinosnoreless wrote:
Thu Sep 15, 2022 8:07 pm
I know for some reason you guys feel the need to whip out your measuring tapes ...
The struggle is real!

Image

Image

_________________
Machine: AirSense 11 Autoset
Mask: Evora Full Face Mask
Additional Comments: I CAN BREATHE!
Angry much?

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: Help with testing pressure at home

Post by lynninnj » Fri Sep 16, 2022 10:42 am

Brad S wrote:
Fri Sep 16, 2022 7:43 am
babydinosnoreless wrote:
Thu Sep 15, 2022 8:07 pm
I know for some reason you guys feel the need to whip out your measuring tapes ...
The struggle is real!

Image

Image
ROFLMAO

Maybe if I ask someone else on the thread who is courteous and helpful I will get a reply.

I am curious to know what one does with the information regarding the pressure testing? Isn't it all sort of about just making adjustments to see what works? Or is this more particular to those on a CPAP rather than an APAP?

Thanks for any helpful responses.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: Help with testing pressure at home

Post by Iamstumped » Fri Sep 16, 2022 11:00 am

I think that it is to just make sure that your machine is putting out the correct pressure, that it should be putting out.

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: Help with testing pressure at home

Post by lynninnj » Fri Sep 16, 2022 11:05 am

Iamstumped wrote:
Fri Sep 16, 2022 11:00 am
I think that it is to just make sure that your machine is putting out the correct pressure, that it should be putting out.
Thanks for taking the time. I can certainly see why that is important for someone on a CPAP, for sure. But with something like i have, with the autoset and the APAP, I can't see the significance.

That's why I ask, so I can learn.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

User avatar
Brad S
Posts: 135
Joined: Mon Jul 25, 2022 10:48 am
Location: Peoples Republik of Kolorado

Re: Help with testing pressure at home

Post by Brad S » Fri Sep 16, 2022 11:18 am

lynninnj wrote:
Fri Sep 16, 2022 11:05 am
Thanks for taking the time. I can certainly see why that is important for someone on a CPAP, for sure. But with something like i have, with the autoset and the APAP, I can't see the significance.
That's why I ask, so I can learn.
It's crazy and frustrating that you have to qualify your question to get an appropriate response.
This Man-O-Meter thing is really interesting to me. I am really interested in the pressure of things- I am a guy :wink: I have looked in the past for a way to measure the output of an oxygen concentrator. What I found was $1,500 or more. I am not that interested. But this might be a way to affordably figure this out.
I am with you though, the pressure of the CPAP might be nice to know, but like you say, if the machine is doing what it is supposed to, then the actual pressure measurement isn't really that important.

_________________
Machine: AirSense 11 Autoset
Mask: Evora Full Face Mask
Additional Comments: I CAN BREATHE!
Angry much?