New large leaks after 18 months of normal leaks... (w/ chinstrap)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
vintage
Posts: 33
Joined: Sun Aug 21, 2022 12:15 pm

New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by vintage » Sun Aug 21, 2022 12:29 pm

I've used an Resmed 10 for about 18 months with no unusual leakage. My Resmed iOS app would generally be 100/100 points per night. But for the last 4 months it's started dipping to as low as 88 (e.g. 8/20 for leaks).

I have the nasal mask (AirFit n20) and I've used a chin strap since about month two.

Here's my Oscar data for last night:

Image

You can see in June my average leak rate went up 68% from May. This month it's almost double what it was in May.

Image

Min/max pressure is 5/10.

Anecdotally I was in bed for 9 hours I took the mask off around hour 7. I feel destroyed today - feels like my pre-CPAP days :(

If anybody has ideas about how to get to the bottom of what might be going on so I can reduce leaks I'd be super grateful.

Thanks in advance!

User avatar
zonker
Posts: 11369
Joined: Fri Jun 19, 2015 4:36 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by zonker » Sun Aug 21, 2022 4:04 pm

vintage wrote:
Sun Aug 21, 2022 12:29 pm


I have the nasal mask (AirFit n20) and I've used a chin strap since about month two.

ahoy!

might be that your chinstrap has hit it's end-of-life and has become too stretched out. i'd suggest a new one or use a firm foam cervical collar.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

vintage
Posts: 33
Joined: Sun Aug 21, 2022 12:15 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by vintage » Sun Aug 21, 2022 4:50 pm

might be that your chinstrap has hit it's end-of-life and has become too stretched out. i'd suggest a new one or use a firm foam cervical collar.


That is highly probable - good call. So the theory is that I'd be seeing mouth leak due to the stretched chinstrap? Is my Oscar chart consistent with mouth leak? Not sure how to read it.

clownbell
Posts: 369
Joined: Fri Apr 02, 2021 8:46 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by clownbell » Sun Aug 21, 2022 6:21 pm

How old is your mask cushion? The silicone can harden up and become deformed/cracked over time, which can result in leaks.
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask

User avatar
zonker
Posts: 11369
Joined: Fri Jun 19, 2015 4:36 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by zonker » Sun Aug 21, 2022 6:23 pm

vintage wrote:
Sun Aug 21, 2022 4:50 pm
might be that your chinstrap has hit it's end-of-life and has become too stretched out. i'd suggest a new one or use a firm foam cervical collar.


That is highly probable - good call. So the theory is that I'd be seeing mouth leak due to the stretched chinstrap? Is my Oscar chart consistent with mouth leak? Not sure how to read it.
i'm not so good at reading charts myself. but take heart-someone will be along shortly who can.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

vintage
Posts: 33
Joined: Sun Aug 21, 2022 12:15 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by vintage » Mon Aug 22, 2022 5:21 am

> How old is your mask cushion? The silicone can harden up and become deformed/cracked over time, which can result in leaks.

I've had this happen before without realizing and it was the source of much frustration. Thankfully I know to avoid this now! The cushion is about a month old and in good shape.

> i'm not so good at reading charts myself. but take heart-someone will be along shortly who can.

:lol: thanks!

Update - used a new chinstrap last night, here's my Oscar stats:

Image

Anecdotally feel like a got a good sleep. Large Leak time is down to 4.22% from 15.34% too. Hopefully it was the chinstrap, though would love any insights from reading the Oscar tea leaves (or a pointer on where I can learn to do so!)

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by palerider » Mon Aug 22, 2022 1:23 pm

vintage wrote:
Sun Aug 21, 2022 12:29 pm
Here's my Oscar data for last night:

Image

You can see in June my average leak rate went up 68% from May. This month it's almost double what it was in May.

Image
You need to use something besides dropbox to post the files, because that isn't working. Most people recommend imgur.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by palerider » Mon Aug 22, 2022 1:26 pm

vintage wrote:
Mon Aug 22, 2022 5:21 am
though would love any insights from reading the Oscar tea leaves (or a pointer on where I can learn to do so!)
Your min pressure should be more like 8.4, it's too low for you.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

vintage
Posts: 33
Joined: Sun Aug 21, 2022 12:15 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by vintage » Mon Aug 22, 2022 1:35 pm

> Your min pressure should be more like 8.4, it's too low for you.

Can you help me understand how you came to that conclusion?

User avatar
zonker
Posts: 11369
Joined: Fri Jun 19, 2015 4:36 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by zonker » Mon Aug 22, 2022 3:59 pm

vintage wrote:
Mon Aug 22, 2022 1:35 pm
> Your min pressure should be more like 8.4, it's too low for you.

Can you help me understand how you came to that conclusion?
i'll take a crack at it and palerider can correct any of my errors.

see the chart that is labeled "pressure"? it has a green line those goes up and down, up and down and so forth. this chart has numbers on the left hand side. this tells you what pressure the machine putting out for each minute that you have the mask on.

the pressure starts at 5, of course, because that's what the machine is set to start at. but then, the pressure slowly goes up as it begins to react to the events happening while you sleep. it rises to 9, then falls back. it ten raises again until in reaches 10. it can't go any further than that because the machine is capped at a maximum of 10.

and so it goes throughout your sleep, up and down. BUT it never goes back to 5. the lowest it gets is just under 7. if you set the minimum at the recommended 8.4, your machine will have a better chance at doing it's job in reacting to the events.

is that clear?

oh and btw. if you want to get someone's attention when you are replying, use the quote feature. there are a lot of users and a lot of posts. not everyone reads every post for many and sundry reasons.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

vintage
Posts: 33
Joined: Sun Aug 21, 2022 12:15 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by vintage » Mon Aug 22, 2022 4:07 pm

i'll take a crack at it and palerider can correct any of my errors.

see the chart that is labeled "pressure"? it has a green line those goes up and down, up and down and so forth. this chart has numbers on the left hand side. this tells you what pressure the machine putting out for each minute that you have the mask on.

the pressure starts at 5, of course, because that's what the machine is set to start at. but then, the pressure slowly goes up as it begins to react to the events happening while you sleep. it rises to 9, then falls back. it ten raises again until in reaches 10. it can't go any further than that because the machine is capped at a maximum of 10.

and so it goes throughout your sleep, up and down. BUT it never goes back to 5. the lowest it gets is just under 7. if you set the minimum at the recommended 8.4, your machine will have a better chance at doing it's job in reacting to the events.

is that clear?

oh and btw. if you want to get someone's attention when you are replying, use the quote feature. there are a lot of users and a lot of posts. not everyone reads every post for many and sundry reasons.

good luck!
Thanks! This is super helpful. When I look at the chart it seems like it bottoms out at 6. The average seems like 8.4. Wouldn't 8.4 be more than necessary? Based on the chart I'd guess that 6 would be more appropriate. I'm a newbie to tweaking this stuff!

User avatar
zonker
Posts: 11369
Joined: Fri Jun 19, 2015 4:36 pm

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by zonker » Mon Aug 22, 2022 4:12 pm

vintage wrote:
Mon Aug 22, 2022 4:07 pm

Thanks! This is super helpful. When I look at the chart it seems like it bottoms out at 6. The average seems like 8.4. Wouldn't 8.4 be more than necessary? Based on the chart I'd guess that 6 would be more appropriate. I'm a newbie to tweaking this stuff!
look again.

that green pressure line shows you are spending more time above 8.4 than below it.

btw, close on the quote feature, but no cigar. :D

when you see this message, you'll see two symbols ! and ". choose the second. that will quote me in full AND give me a notification.

continued good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by Okie bipap » Mon Aug 22, 2022 4:14 pm

It's the minimum pressure that does the most to prevent apneas. Yes, your minimum pressure waws six, but the machine never stayed there for any length of time. Almost as soon as it dropped to six, it immediately went back up to around 8.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by palerider » Mon Aug 22, 2022 4:15 pm

zonker wrote:
Mon Aug 22, 2022 3:59 pm
vintage wrote:
Mon Aug 22, 2022 1:35 pm
> Your min pressure should be more like 8.4, it's too low for you.

Can you help me understand how you came to that conclusion?
... an excellent explanation...
Oh, man, you make me proud. That's SPOT ON! I do love people that pay attention. :)

I might add that raising the max pressure would be good too, since 10 is too low. (yeah, as usual I'd say to set it to 20 and forget it).
zonker wrote:
Mon Aug 22, 2022 3:59 pm
oh and btw. if you want to get someone's attention when you are replying, use the quote feature. there are a lot of users and a lot of posts. not everyone reads every post for many and sundry reasons.
Yeah, I just stumbled back on your reply. :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Mon Aug 22, 2022 4:18 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New large leaks after 18 months of normal leaks... (w/ chinstrap)

Post by palerider » Mon Aug 22, 2022 4:16 pm

vintage wrote:
Mon Aug 22, 2022 4:07 pm
Thanks! This is super helpful. When I look at the chart it seems like it bottoms out at 6. The average seems like 8.4. Wouldn't 8.4 be more than necessary? Based on the chart I'd guess that 6 would be more appropriate. I'm a newbie to tweaking this stuff!
Where it "bottoms out" is too low, because when it 'bottoms out', it immediately jumps back up because you had more breathing events.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto