Z1 Fault 33 Issue and Response from Breas
Z1 Fault 33 Issue and Response from Breas
The following is an email exchange with an orders@Breas person.
Fault 33 - Means the internal battery for memory has stopped working and can no longer save data in its memory.
While the issue will continue to occur in the Z1's due to lack of use, Covid paid a huge factor in this, the new Z2 does not appear to have the same issue.
I just did not remember any instructions on the Z1 instructing one to plug the unit in during periods of extended non-use. Did I miss this point in the provided instructions?
No unfortunately there are no statements in the instructions manual. This was more discovered as with travel coming to a halt there was a lot of non use of the units. So going forward the recommendation is to plug the unit in once a month to ensure consistency.
Fault 33 - Means the internal battery for memory has stopped working and can no longer save data in its memory.
While the issue will continue to occur in the Z1's due to lack of use, Covid paid a huge factor in this, the new Z2 does not appear to have the same issue.
I just did not remember any instructions on the Z1 instructing one to plug the unit in during periods of extended non-use. Did I miss this point in the provided instructions?
No unfortunately there are no statements in the instructions manual. This was more discovered as with travel coming to a halt there was a lot of non use of the units. So going forward the recommendation is to plug the unit in once a month to ensure consistency.
- chunkyfrog
- Posts: 34459
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the abyss that is Nebraska--wish me luck!
Re: Z1 Fault 33 Issue and Response from Breas
Non-rechargeable lithium batteries are often used in these applications.
These batteries have long lifespans, often ranging from 5 to 10 years.
Plugging it in will not help. The lifespan is continuous, and failure is inevitable.
Contact Acbio.com, if you are reluctant to change your battery yourself.
Continuous non-use spares other parts from wear, but not the battery.
These batteries have long lifespans, often ranging from 5 to 10 years.
Plugging it in will not help. The lifespan is continuous, and failure is inevitable.
Contact Acbio.com, if you are reluctant to change your battery yourself.
Continuous non-use spares other parts from wear, but not the battery.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Z1 Fault 33 Issue and Response from Breas
Is the internal battery customer replaceable? If so, are there any instructions on how to do this? I've searched online but cannot find any instructions on this.
- chunkyfrog
- Posts: 34459
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the abyss that is Nebraska--wish me luck!
Re: Z1 Fault 33 Issue and Response from Breas
It depends on the customer.
How into DIY are you?
The process requires skill with tools, possibly soldering.
Some posters have done-or attempted it.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Z1 Fault 33 Issue and Response from Breas
I haven’t looked but ifixit.com would be the first place I would look.
_________________
Machine: AirSense 11 Autoset |
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
- chunkyfrog
- Posts: 34459
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the abyss that is Nebraska--wish me luck!
Re: Z1 Fault 33 Issue and Response from Breas
Thank you, Lynn.
I had forgotten about that site.
I had forgotten about that site.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Z1 Fault 33 Issue and Response from Breas
Sorry. I just looked and they don’t have that machine though they do have others.
_________________
Machine: AirSense 11 Autoset |
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
Re: Z1 Fault 33 Issue and Response from Breas
HI.
Just got the error 33 on the z1
Has anyone found out how to replace the battery?
It's a decent travel machine. I'd rather fix it than to lug a larger one around & I won't be spending more money on another travel machine.
I assumed (yea...I know) that it wouldn't wear out from lack of use.
Just got the error 33 on the z1
Has anyone found out how to replace the battery?
It's a decent travel machine. I'd rather fix it than to lug a larger one around & I won't be spending more money on another travel machine.
I assumed (yea...I know) that it wouldn't wear out from lack of use.
Re: Z1 Fault 33 Issue and Response from Breas
My husband is having the same issue. I found this site with instructions on how to replace the battery, but it isn’t an easy DIY project: https://www.apneaboard.com/forums/Threa ... eplacement