ResMed Aircurve`10 VAUTO

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
KENSKIP1
Posts: 47
Joined: Thu Jun 02, 2022 7:46 am
Location: Missouri Central

ResMed Aircurve`10 VAUTO

Post by KENSKIP1 » Thu Jun 09, 2022 6:42 am

Good morning. Well I have had the machine being in usage now for 5 nights. However I do not feel any benefits from it at all.
My last 5 AHI score may tell us a bit more. Here is the past readings from last Saturday to present.
7.88 Saturday
1.43 Sunday
.83 Monday
0.50 Tuesday
1.50 Wednesday.
I will mention that the pressure settings are 12 and 8.
Something else, when I am exhaling it seems to be redirecting part of the air back into the mask.
Does the outside barometric pressure have anything to do with the way the machine operates? It seems as if
I am not getting the pressure when inhaling like I received the past few nights ago. How accurate is the flow
meter? It showed around 4 when it was supposed to be at 8. Same for the incoming air. It showed 8 and was
set to be 12.Thanks Ken

_________________
MachineMask

User avatar
Dog Slobber
Posts: 4232
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: ResMed Aircurve`10 VAUTO

Post by Dog Slobber » Thu Jun 09, 2022 7:24 am

KENSKIP1 wrote:
Thu Jun 09, 2022 6:42 am
Good morning. Well I have had the machine being in usage now for 5 nights. However I do not feel any benefits from it at all.
I don't *feel* the benefits of the rosuvastatin, I take.
My last 5 AHI score may tell us a bit more. Here is the past readings from last Saturday to present.
7.88 Saturday
1.43 Sunday
.83 Monday
0.50 Tuesday
1.50 Wednesday.
I will mention that the pressure settings are 12 and 8.
Instead of posting a list of numbers, post OSCAR charts.
Something else, when I am exhaling it seems to be redirecting part of the air back into the mask.
What does this even mean?
Does the outside barometric pressure have anything to do with the way the machine operates?


Not until you hit an altitude of about 8000 feet or 752.6mb/hPa
It seems as if I am not getting the pressure when inhaling like I received the past few nights ago. How accurate is the flow meter? It showed around 4 when it was supposed to be at 8. Same for the incoming air. It showed 8 and was set to be 12.Thanks Ken
The sensors are a lot more accurate than what we, subjectively believe them to be. As we get accustomed to CPAP, the feelings and pressures we felt when new, or the beginning of a session become less noticable. OSCAR graphs would show the devices settings and responses.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65119
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed Aircurve`10 VAUTO

Post by Pugsy » Thu Jun 09, 2022 7:28 am

KENSKIP1 wrote:
Thu Jun 09, 2022 6:42 am
How accurate is the flow
meter? It showed around 4 when it was supposed to be at 8. Same for the incoming air. It showed 8 and was
set to be 12.
Is ramp enabled? If using ramp that would probably explain the difference.

Post an OSCAR detailed report so we can see the details for more comprehensive responses.
Right now all we can do is give you WAGs...wild ass guesses.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
KENSKIP1
Posts: 47
Joined: Thu Jun 02, 2022 7:46 am
Location: Missouri Central

Re: ResMed Aircurve`10 VAUTO

Post by KENSKIP1 » Thu Jun 09, 2022 7:37 am

OSCAR
Usage Statistics

I hope this is what you are referring to. Yes I do have the ramp feature activated 5 minutes.




CPAP Statistics
5 days of CPAP Data, between Sat Jun 4 2022 and Wed Jun 8 2022
Details
Most Recent
Last Week
Last 30 Days
Last 6 Months
Last Year
CPAP Usage
Average Hours per Night
07:19
05:55
05:55
05:55
05:55
Compliance (4 hrs/day)
100%
80%
80%
80%
80%
Therapy Efficacy
AHI
1.50
2.03
2.03
2.03
2.03
Obstructive Index
0.82
1.28
1.28
1.28
1.28
Hypopnea Index
0.41
0.30
0.30
0.30
0.30
Unclassified Apnea Index
0.00
0.00
0.00
0.00
0.00
Clear Airway Index
0.27
0.44
0.44
0.44
0.44
95% Flow Limitation
0.00
0.00
0.00
0.00
0.00
Leak Statistics
Average Leak Rate
6.69
7.55
7.55
7.55
7.55
95% Leak Rate
14.40
18.00
18.00
18.00
18.00
% of time above Leak Rate threshold
0.03%
0.91%
0.91%
0.91%
0.91%
Pressure Statistics
Average Pressure
11.33
11.34
11.34
11.34
11.34
Min Pressure
4.36
4.36
4.36
4.36
4.36
Max Pressure
13.14
14.00
14.00
14.00
14.00
95% Pressure
12.52
12.70
12.70
12.70
12.70
Average EPAP
8.39
8.40
8.40
8.40
8.40
Min EPAP
4.00
4.00
4.00
4.00
4.00
Max EPAP
10.14
11.70
11.70
11.70
11.70



Changes to Machine Settings
First
Last
Days
AHI
FL
Machine
Pressure Relief
Mode
Pressure Settings
Wed Jun 8 2022
Wed Jun 8 2022
1
1.50
0.00
AirCurve 10 VAuto (37094)
None
VPAPauto
PS 2.0 over 10.0-12.0 (cmH2O)
Tue Jun 7 2022
Tue Jun 7 2022
1
0.50
0.00
AirCurve 10 VAuto (37094)
None
VPAPauto
PS 3.0 over 8.0-14.0 (cmH2O)
Mon Jun 6 2022
Mon Jun 6 2022
1
0.83
0.00
AirCurve 10 VAuto (37094)
None
VPAPauto
PS 3.0 over 9.0-14.0 (cmH2O)
Sun Jun 5 2022
Sun Jun 5 2022
1
1.43
0.00
AirCurve 10 VAuto (37094)
None
VPAPauto
PS 2.0 over 10.0-12.0 (cmH2O)
Sat Jun 4 2022
Sat Jun 4 2022
1
7.88
0.00
AirCurve 10 VAuto (37094)
None
VPAPauto
PS 4.0 over 4.0-12.0 (cmH2O)



Machine Information
Brand
Model
Serial
First Use
Last Use


ResMed
AirCurve 10 VAuto (37094)
23212041429
Sat Jun 4 2022
Wed Jun 8 2022




This report was prepared on Thu Jun 9 2022 08:29 by OSCAR 1.3.1+3ca1ee24
OSCAR is free open-source CPAP report software

_________________
MachineMask

User avatar
Respirator99
Posts: 389
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: ResMed Aircurve`10 VAUTO

Post by Respirator99 » Thu Jun 09, 2022 8:11 am

Ken, that report format isn't very helpful. What we should have asked for is a screen shot of the Oscar daily view. We need to see the following graphs: Events, Flow rate; Pressure, Leak rate and Flow limitation. Also need to see the left side-pane with the statistics and machine settings, but make sure the calendar is turned off. To create the screen shot, press the [F12] button on your computer - that will create the image file in your pictures folder. Here's an example of what we need to see:
screenshot-20220609-221014.png
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

User avatar
KENSKIP1
Posts: 47
Joined: Thu Jun 02, 2022 7:46 am
Location: Missouri Central

Re: ResMed Aircurve`10 VAUTO

Post by KENSKIP1 » Thu Jun 09, 2022 10:22 am

But how do I get a "Screen Shot"

_________________
MachineMask

User avatar
Miss Emerita
Posts: 3750
Joined: Sun Nov 04, 2018 8:07 pm

Re: ResMed Aircurve`10 VAUTO

Post by Miss Emerita » Thu Jun 09, 2022 11:00 am

That will work differently on different computers. On my laptop, there's a key marked PrtSc. You can always google "how to print screen on a [fill in your computer make and model]."
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
KENSKIP1
Posts: 47
Joined: Thu Jun 02, 2022 7:46 am
Location: Missouri Central

Re: ResMed Aircurve`10 VAUTO

Post by KENSKIP1 » Thu Jun 09, 2022 11:02 am

Screenshot (1).png
Screenshot (1).png (177.78 KiB) Viewed 906 times

_________________
MachineMask

User avatar
Pugsy
Posts: 65119
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed Aircurve`10 VAUTO

Post by Pugsy » Thu Jun 09, 2022 11:04 am

http://www.apneaboard.com/wiki/index.ph ... ganization organize and how to do screen shots.

Read the above

Pay attention to the organization section

Turn off the calendar
On the right side with the graphs these graphs
Events
FLow rate
Pressure (not mask pressure)
Leak
FLow limitation

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
loggerhead12
Posts: 585
Joined: Fri Sep 01, 2017 4:00 pm

Re: ResMed Aircurve`10 VAUTO

Post by loggerhead12 » Thu Jun 09, 2022 11:39 am

KENSKIP1 wrote:
Thu Jun 09, 2022 10:22 am
But how do I get a "Screen Shot"
In Windows you use the Snipping Tool and on a Mac you use Shift-Command-4.

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: ResMed Aircurve`10 VAUTO

Post by Iamstumped » Thu Jun 09, 2022 11:49 am

KENSKIP1 wrote:
Thu Jun 09, 2022 10:22 am
But how do I get a "Screen Shot"
First of all, download the OSCAR software, if you have not done so.

User avatar
KENSKIP1
Posts: 47
Joined: Thu Jun 02, 2022 7:46 am
Location: Missouri Central

Re: ResMed Aircurve`10 VAUTO

Post by KENSKIP1 » Thu Jun 09, 2022 11:58 am

Iamstumped wrote:
Thu Jun 09, 2022 11:49 am
KENSKIP1 wrote:
Thu Jun 09, 2022 10:22 am
But how do I get a "Screen Shot"
First of all, download the OSCAR software, if you have not done so.
Yes I have the software installed.

_________________
MachineMask

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: ResMed Aircurve`10 VAUTO

Post by Iamstumped » Thu Jun 09, 2022 12:27 pm

Refer to PUGSY's post above.

In the upper left click "view", then click "take screenshot"

User avatar
Miss Emerita
Posts: 3750
Joined: Sun Nov 04, 2018 8:07 pm

Re: ResMed Aircurve`10 VAUTO

Post by Miss Emerita » Thu Jun 09, 2022 1:19 pm

KENSKIP1, you're almost there! Pugsy has explained the additional changes to make before taking another screen shot.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Dog Slobber
Posts: 4232
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: ResMed Aircurve`10 VAUTO

Post by Dog Slobber » Thu Jun 09, 2022 3:25 pm

The anti-bacterial filter setting is on.

Unless you have an inline AB filter, that is likely a mistake and would explain pressures feeling off.
Battery Backup: EcoFlow Delta 2