Fluctuating pulse

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Applecheeks
Posts: 86
Joined: Fri Dec 24, 2021 10:07 am

Re: Fluctuating pulse

Post by Applecheeks » Tue Jun 07, 2022 9:38 pm

Chris33022 wrote:
Tue Jun 07, 2022 6:20 pm
Applecheeks wrote:
Tue Jun 07, 2022 5:59 pm
So, it is probably useful to know for sure the details on how Wellu calculates pulse rates and the interval used for each data point. ( my suspicion is 4 seconds - but that actually seems like overkill)
Is the following helpful, by any chance?
<< Measurement interval: 1 second (in-app intervals 4secs) >>
Yes, I saw that in the O2 ring Tech Specifications (at their website).
The only problem is that in the one second measurement interval there will be only 1 detected pulse when the actual pulse is 60 or less. In order to calculate a pulse rate --- you need at least two detected pulses and the interval between the two pulses. (see below)

img010.jpg

User avatar
Chris33022
Posts: 39
Joined: Wed Mar 30, 2022 5:46 pm

Re: Fluctuating pulse

Post by Chris33022 » Tue Jun 07, 2022 9:44 pm

dataq1 wrote:
Tue Jun 07, 2022 8:56 pm
P.S. Also, this guy on YouTube seems quite smart and talks about HRV in relation to sleep and the Oura Ring: https://www.youtube.com/watch?v=AK47cav1i28
But I agree we don't have to talk more about Heart Rate Variability in this thread if it's too off-topic.
(Oh, I also wanted to add that when I'm stressed, my HRV during sleep is always dramatically lower than when I'm not stressed.)

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Fluctuating pulse

Post by Rubicon » Wed Jun 08, 2022 4:26 am

Chris33022 wrote:
Tue Jun 07, 2022 9:44 pm
But I agree we don't have to talk more about Heart Rate Variability in this thread if it's too off-topic.
HRV is precisely on-topic if it's used in the right context.

But you guys are talking about rMSSD which is not appropriate in re: OP's concerns. They need to specifically look at LF which requires a 1-2 minute window.

That said, if the OP is not in NSR (and the AF point is highly relevant) then all talk about HRV is out the window.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Fluctuating pulse

Post by Rubicon » Wed Jun 08, 2022 4:34 am

Cause you get the AF doing something like this and HRV is meaningless:

Image
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
Pugsy
Posts: 65384
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Fluctuating pulse

Post by Pugsy » Wed Jun 08, 2022 11:35 am

I think that pulse rate variability and HRV are indeed "on topic" and I see no need for a separate thread just to rehash what has already been brought forward here.

I feel it will just result in duplication of posts and questions with same results.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Fluctuating pulse

Post by dataq1 » Wed Jun 08, 2022 12:36 pm

Rubicon wrote:
Wed Jun 08, 2022 4:34 am
Cause you get the AF doing something like this and HRV is meaningless:
By "AF" are you referring to AFib ?
Would the HRV also be meaningless in PVCs and other ectopic arrhythmias?
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
Applecheeks
Posts: 86
Joined: Fri Dec 24, 2021 10:07 am

Re: Fluctuating pulse

Post by Applecheeks » Wed Jun 08, 2022 8:57 pm

Regarding the sampling rate and the reporting intervals for the Wellue O2ring.
For what it is worth: I wrote to the company's tech service department and here is what their answer was:
The Sampling rate is 150HZ. The interface is refreshed in real time, also 4s point records the last data (that is the 600th single data, not the average data)
For the sake of completeness, this is what I asked:
For the O2 saturation reading, how many individual measurements have been taken of SpO2 before a measurement is reported? If more than one measurement is used, are the individual measurements averaged?

On pulse rate, the natural assumption is that the O2 ring detects and counts the pulses over a 4 second interval and calculates a pulse rate from that. Is that assumption correct?

Finally, regarding pulse rate and O2 display/report only, is the data reported completely independent from prior 4 second display/reports for each parameter?