Airsense10 battery turns off after 3 minutes

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
hsnopi
Posts: 4
Joined: Tue Apr 19, 2022 2:02 pm

Airsense10 battery turns off after 3 minutes

Post by hsnopi » Tue Apr 19, 2022 2:07 pm

Hello all,
I have the Airsense 10 auto-start (327028) and the converter (370003). I know the equipment works. I have a eco-worthy 12v 8ah (96wh) deep cycle LiFePO4 battery. charged it reads 13.3v. It works for the kids scooters and go carts. I plug it in to the cpap and it runs for about 3 minutes and turns off. heater is off, in airplane mode. setting is 9. The chart says it pulls about .85amps.

Why is it turning off?
Thanks!
Hsnopi

User avatar
Dog Slobber
Posts: 4161
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Airsense10 battery turns off after 3 minutes

Post by Dog Slobber » Tue Apr 19, 2022 2:54 pm

By heater are you referring to the heated hose, or the humidifier?

Ensure the humidifier is also off.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense10 battery turns off after 3 minutes

Post by palerider » Tue Apr 19, 2022 3:01 pm

hsnopi wrote:
Tue Apr 19, 2022 2:07 pm
Why is it turning off?
Thanks!
Hsnopi
magic 8 ball says:
If you hook the converter to your car's battery, how long does the machine run for?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

hsnopi
Posts: 4
Joined: Tue Apr 19, 2022 2:02 pm

Re: Airsense10 battery turns off after 3 minutes

Post by hsnopi » Tue Apr 19, 2022 3:07 pm

Hello,
The humidifier is off. Just a slimline hose. No conditioned hose. I hooked it up to the car and it ran for 15 minutes before I turned it off. Same for house power.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense10 battery turns off after 3 minutes

Post by palerider » Tue Apr 19, 2022 4:01 pm

hsnopi wrote:
Tue Apr 19, 2022 3:07 pm
Hello,
The humidifier is off. Just a slimline hose. No conditioned hose. I hooked it up to the car and it ran for 15 minutes before I turned it off. Same for house power.
The problem then, is your battery, Q.E.D.

Maybe get another one and put 'em in series.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Grumpy48
Posts: 412
Joined: Tue May 12, 2020 7:57 am

Re: Airsense10 battery turns off after 3 minutes

Post by Grumpy48 » Tue Apr 19, 2022 4:51 pm

palerider wrote:
Tue Apr 19, 2022 4:01 pm
hsnopi wrote:
Tue Apr 19, 2022 3:07 pm
Hello,
The humidifier is off. Just a slimline hose. No conditioned hose. I hooked it up to the car and it ran for 15 minutes before I turned it off. Same for house power.
The problem then, is your battery, Q.E.D.

Maybe get another one and put 'em in series.
For 24 volts????

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: Resmed AirSense 10 Auto with Humidifier

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense10 battery turns off after 3 minutes

Post by palerider » Tue Apr 19, 2022 9:13 pm

Grumpy48 wrote:
Tue Apr 19, 2022 4:51 pm
palerider wrote:
Tue Apr 19, 2022 4:01 pm
hsnopi wrote:
Tue Apr 19, 2022 3:07 pm
Hello,
The humidifier is off. Just a slimline hose. No conditioned hose. I hooked it up to the car and it ran for 15 minutes before I turned it off. Same for house power.
The problem then, is your battery, Q.E.D.

Maybe get another one and put 'em in series.
For 24 volts????
Well, yes, the Resmed converts are rated for 12-24 vdc.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

hsnopi
Posts: 4
Joined: Tue Apr 19, 2022 2:02 pm

Re: Airsense10 battery turns off after 3 minutes

Post by hsnopi » Wed Apr 20, 2022 7:01 am

I'm a little confused though about one thing. The chart shows the amp draw at .85 (roughly) but the power brick says 3.75a. is 3.75 a max rating? Trying to calculate actual draw rate and battery run time.

User avatar
Dog Slobber
Posts: 4161
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Airsense10 battery turns off after 3 minutes

Post by Dog Slobber » Wed Apr 20, 2022 7:19 am

hsnopi wrote:
Wed Apr 20, 2022 7:01 am
I'm a little confused though about one thing. The chart shows the amp draw at .85 (roughly) but the power brick says 3.75a. is 3.75 a max rating? Trying to calculate actual draw rate and battery run time.
It's .85 without using the humidifier, heated hose and pressure set at 9. The power brick has to be capable of providing enough amperage using those features and higher pressure.

User avatar
SleepGeek
Posts: 1055
Joined: Mon Feb 09, 2015 2:56 pm

Re: Airsense10 battery turns off after 3 minutes

Post by SleepGeek » Wed Apr 20, 2022 6:34 pm

hsnopi wrote:
Tue Apr 19, 2022 2:07 pm
Hello all,
I have the Airsense 10 auto-start (327028) and the converter (370003). I know the equipment works. I have a eco-worthy 12v 8ah (96wh) deep cycle LiFePO4 battery. charged it reads 13.3v. It works for the kids scooters and go carts. I plug it in to the cpap and it runs for about 3 minutes and turns off. heater is off, in airplane mode. setting is 9. The chart says it pulls about .85amps.
If you have a voltmeter - measure the voltage while the cpap is powered on.
hsnopi wrote:
Tue Apr 19, 2022 2:07 pm
Why is it turning off?
I suspect the voltage is dropping - that is what they will do when the voltage is low, they shut OFF>
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: What exactly do you think you're adding to this thread?

hsnopi
Posts: 4
Joined: Tue Apr 19, 2022 2:02 pm

Re: Airsense10 battery turns off after 3 minutes

Post by hsnopi » Thu Apr 21, 2022 6:39 am

Thanks for all the help. I watched the voltmeter. While it is sitting the battery is 13.4v but when the cpap is on it drops to 8.5. Even though it is an 8Ah battery it still drops that low. I'll play with it more later.

User avatar
SleepGeek
Posts: 1055
Joined: Mon Feb 09, 2015 2:56 pm

Re: Airsense10 battery turns off after 3 minutes

Post by SleepGeek » Thu Apr 21, 2022 6:54 am

hsnopi wrote:
Thu Apr 21, 2022 6:39 am
I watched the voltmeter. While it is sitting the battery is 13.4v but when the cpap is on it drops to 8.5. Even though it is an 8Ah battery it still drops that low.
Time for a new battery.
Is this one old? Or just been discharged too low too many times?
8ah is really not that big. Depends on how long you want it to last a 35ah would be better.
Sometimes two are better so you can charge one while still using the other.

Once the power goes out there is no time for planning this.

Now we know what happens when the voltage goes too low - one guy thought the cpap would catch on fire? lol
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: What exactly do you think you're adding to this thread?

User avatar
loggerhead12
Posts: 584
Joined: Fri Sep 01, 2017 4:00 pm

Re: Airsense10 battery turns off after 3 minutes

Post by loggerhead12 » Thu Apr 21, 2022 10:28 am

hsnopi wrote:
Thu Apr 21, 2022 6:39 am
While it is sitting the battery is 13.4v but when the cpap is on it drops to 8.5. Even though it is an 8Ah battery it still drops that low. I'll play with it more later.
No need to play any more. That is the textbook example of what a battery does under load when it's bad.