Evidently NO WIFI

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
golfgame
Posts: 68
Joined: Thu Jul 18, 2019 12:33 pm

Evidently NO WIFI

Post by golfgame » Mon Apr 18, 2022 4:08 am

Evidently over the last few days my ResMed Aircurve10 VAUTO unit has lost its WIFI. I am seeing a yellow circle with a line thru it on the screen. Am I wrong in thinking that too mean that it is not receiving or sending data out? The Remed daily report has shown no use for several days when in fact it has been used.
:?: :?: :?:

User avatar
Respirator99
Posts: 406
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: Evidently NO WIFI

Post by Respirator99 » Mon Apr 18, 2022 4:30 am

Your machine communicates by phone, not wifi. Older generations of cellular phone systems (eg 3G) have been phased out by the operator, so there is no longer any way for it to communicate.

By the way - use the search engine. There have been numerous threads on this topic over the past few weeks.
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

User avatar
LSAT
Posts: 13412
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Evidently NO WIFI

Post by LSAT » Mon Apr 18, 2022 6:03 am

All your data is recorded on the machine's SD Card. You can see all your data on the screen's Sleep Report. If you are not seeing the data, it is not turned on in the settings area. Go into settings...find ESSENTIALS and change it to PLUS,'
Last edited by LSAT on Fri Apr 22, 2022 7:51 am, edited 1 time in total.

User avatar
Conrad
Posts: 704
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Evidently NO WIFI

Post by Conrad » Mon Apr 18, 2022 6:48 am

golfgame wrote:
Mon Apr 18, 2022 4:08 am
Evidently over the last few days my ResMed Aircurve10 VAUTO unit has lost its WIFI. I am seeing a yellow circle with a line thru it on the screen. Am I wrong in thinking that too mean that it is not receiving or sending data out? The Remed daily report has shown no use for several days when in fact it has been used.
:?: :?: :?:
I have the same machine and it's connecting to the cellular network just fine, I just now checked it. Did you change any settings?
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Evidently NO WIFI

Post by Dog Slobber » Mon Apr 18, 2022 6:59 am

Conrad wrote:
Mon Apr 18, 2022 6:48 am
I have the same machine and it's connecting to the cellular network just fine, I just now checked it. Did you change any settings?
You may have an AirCurve 10 VAuto, but they are not identical. The Air 10 line has had three different modems (CDMA, 3G, 4G), as well as numerous firmware versions. You are also (likely) not in the same cell region.

The cell carriers are shutting down older cell services and these topics are popping up all over many CPAP forums.
Battery Backup: EcoFlow Delta 2

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Evidently NO WIFI

Post by Janknitz » Mon Apr 18, 2022 12:44 pm

And if you've been relying on wifi data, you may not have an SD card in your CPAP machine. A card was shipped with your machine but may have been removed by the DME. You can buy an inexpensive data card for going forward. If you need to prove compliance for insurance purposes, you can let your doctor or DME get the data from the card.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

golfgame
Posts: 68
Joined: Thu Jul 18, 2019 12:33 pm

Re: Evidently NO WIFI

Post by golfgame » Mon Apr 18, 2022 1:11 pm

Dog Slobber wrote:
Mon Apr 18, 2022 6:59 am
Conrad wrote:
Mon Apr 18, 2022 6:48 am
I have the same machine and it's connecting to the cellular network just fine, I just now checked it. Did you change any settings?
You may have an AirCurve 10 VAuto, but they are not identical. The Air 10 line has had three different modems (CDMA, 3G, 4G), as well as numerous firmware versions. You are also (likely) not in the same cell region.

The cell carriers are shutting down older cell services and these topics are popping up all over many CPAP forums.
I contacted ResMed and they confirmed that the machine was not sending or receiving data via the cell tower. They said I had to go thru my supplier and the supplier would then contact ResMed for the fix. Kinda stupid I'd say.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Evidently NO WIFI

Post by Dog Slobber » Mon Apr 18, 2022 1:23 pm

golfgame wrote:
Mon Apr 18, 2022 1:11 pm
Dog Slobber wrote:
Mon Apr 18, 2022 6:59 am
Conrad wrote:
Mon Apr 18, 2022 6:48 am
I have the same machine and it's connecting to the cellular network just fine, I just now checked it. Did you change any settings?
You may have an AirCurve 10 VAuto, but they are not identical. The Air 10 line has had three different modems (CDMA, 3G, 4G), as well as numerous firmware versions. You are also (likely) not in the same cell region.

The cell carriers are shutting down older cell services and these topics are popping up all over many CPAP forums.
I contacted ResMed and they confirmed that the machine was not sending or receiving data via the cell tower. They said I had to go thru my supplier and the supplier would then contact ResMed for the fix. Kinda stupid I'd say.
There is no fix, at least not with your current machine.

Your machine uses cell technology that is no longer supported by the cell carrier.

If you want to keep using your existing machine you won't have data upload capability. If you want data upload capability, get a new machine.
Battery Backup: EcoFlow Delta 2

golfgame
Posts: 68
Joined: Thu Jul 18, 2019 12:33 pm

Re: Evidently NO WIFI

Post by golfgame » Mon Apr 18, 2022 1:26 pm

I would still give data via Oscar though, correct?

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Evidently NO WIFI

Post by Pugsy » Mon Apr 18, 2022 1:31 pm

golfgame wrote:
Mon Apr 18, 2022 1:26 pm
I would still give data via Oscar though, correct?

Correct.
Your machine still will write data to the SD card and it still will store the limited data in the machine's internal memory that it always stored.
The only thing that it can't do now is send the data to the cloud via the cellular transmission.

You can still see your sleep report data on the LCD screen if you wish. The sleep report is pretty much what MyAir showed anyway minus the "scores". Do you know about how to get the sleep report? I glance at mine each morning when I turn the machine off. Shows all I need for the short term....AHI and hours used and leak. If I want details I use the SD card and OSCAR software.

The only thing that has changed is the machine's ability to use the cellular network it came with.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.


golfgame
Posts: 68
Joined: Thu Jul 18, 2019 12:33 pm

Re: Evidently NO WIFI

Post by golfgame » Tue Apr 19, 2022 2:20 am

Pugsy wrote:
Mon Apr 18, 2022 1:31 pm
golfgame wrote:
Mon Apr 18, 2022 1:26 pm
I would still give data via Oscar though, correct?

Correct.
Your machine still will write data to the SD card and it still will store the limited data in the machine's internal memory that it always stored.
The only thing that it can't do now is send the data to the cloud via the cellular transmission.

You can still see your sleep report data on the LCD screen if you wish. The sleep report is pretty much what MyAir showed anyway minus the "scores". Do you know about how to get the sleep report? I glance at mine each morning when I turn the machine off. Shows all I need for the short term....AHI and hours used and leak. If I want details I use the SD card and OSCAR software.

The only thing that has changed is the machine's ability to use the cellular network it came with.
Since the machines are no longer forwarding the daily data to whomever, if you know how will that affect Medicare monitoring a patients usage requirements to enable them to receive the required monthly supplies?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Evidently NO WIFI

Post by palerider » Tue Apr 19, 2022 7:55 am

golfgame wrote:
Tue Apr 19, 2022 2:20 am
Since the machines are no longer forwarding the daily data to whomever, if you know how will that affect Medicare monitoring a patients usage requirements to enable them to receive the required monthly supplies?
There's no reason to get supplies on a monthly basis, that's entirely unnecessary and wasteful.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

golfgame
Posts: 68
Joined: Thu Jul 18, 2019 12:33 pm

Re: Evidently NO WIFI

Post by golfgame » Tue Apr 19, 2022 8:06 am

palerider wrote:
Tue Apr 19, 2022 7:55 am
golfgame wrote:
Tue Apr 19, 2022 2:20 am
Since the machines are no longer forwarding the daily data to whomever, if you know how will that affect Medicare monitoring a patients usage requirements to enable them to receive the required monthly supplies?
There's no reason to get supplies on a monthly basis, that's entirely unnecessary and wasteful.
How so, Medicare allows for the supplier to send a 3 month supply of the facemask portion for example. That eliminates the end user from having to place an order monthly. My supplier will not take an order until the 30 or 90 days cycle has completed and then it takes about 10 days for the order to arrive.

User avatar
Pugsy
Posts: 65406
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Evidently NO WIFI

Post by Pugsy » Tue Apr 19, 2022 8:43 am

Your DME can still satisfy any usage requirements from the machine stored hours of use or the SD card data.
So usage hours requirement data is available if the DME needs it but they may not even need it.
If they do it is usually once a year.
Contact your DME directly to figure out what they will need later....if anything.