Read and Decipher Oscar?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CChamp
Posts: 6
Joined: Wed Mar 16, 2022 8:44 am

Read and Decipher Oscar?

Post by CChamp » Fri Mar 18, 2022 8:32 am

Hi all, I have been using CPAP for about 4 nights now and I think I am doing ok but trying to read the Oscar data is really confusing to me. I think that the one break in pressure is when I had to disconnect and get up to turn the heat on. Could someone please let me know if I should change or adjust anything? Thank you very much.


Machine: ResMed Air Sense 11
Mask Type: Full face mask
Mask Make & Model: RESMED F30i
Humidifier: Unsure
CPAP Pressure: 4-16
CPAP Software: OSCAR myAir

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Read and Decipher Oscar?

Post by chunkyfrog » Fri Mar 18, 2022 9:26 am

Time to start sneaking your bottom pressure up to at least 8.
A little bit at a time, a few days each level.
4 cm is too low to start when your 95% pressure is over 10.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65261
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Read and Decipher Oscar?

Post by Pugsy » Fri Mar 18, 2022 9:50 am

95% pressure numbers just mean the person was at OR BELOW that number for 95% of the night.
It's just a number...no more and no less and it doesn't necessarily mean that someone needs to be using that number (or near that number) for the bulk of the night.
95% numbers are easily skewed to the high side by a relatively short period of time at higher pressures.
This is what happened on the above report. Short period of time where more pressure was needed at around 3:30 to 4:15.

Most likely either REM stage sleep or supine sleeping or maybe a combination of both resulted in the OSA being worse and needing a higher pressure to hold the airway open.
The rest of the night things were really well controlled.

Often when we increase that minimum pressure a little bit and give the machine a better head start then the machine might not need to go so high anyway.

I think a minimum pressure of 6 would be a good choice to try and see if it keeps clusters like what is seen above, away before I would be wanting to go to 8 cm minimum. At this point I don't think 8 cm minimum is needed.
I would want to try less first. Might get lucky and with 6 cm minimum that cluster doesn't happen.

I don't ever base pressure needs on 95% numbers when the only time the higher pressure is seen is for a 30 to 45 minute time frame during the night. Now if it were half the night or more...different story.

Besides...only 4 nights on cpap....need to know if that cluster has happened before or happens again in the future.
If a random rare occurrence I wouldn't go making any big changes in anything.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

CChamp
Posts: 6
Joined: Wed Mar 16, 2022 8:44 am

Re: Read and Decipher Oscar?

Post by CChamp » Fri Mar 18, 2022 10:05 am

Thank you both very much.

User avatar
zonker
Posts: 11427
Joined: Fri Jun 19, 2015 4:36 pm

Re: Read and Decipher Oscar?

Post by zonker » Fri Mar 18, 2022 10:55 am

CChamp wrote:
Fri Mar 18, 2022 10:05 am
Thank you both very much.
also, could you stay in this thread for any further therapy questions?

thanks!

and good luck.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg