Resmed Events vs AHI

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
SleepDeep
Posts: 23
Joined: Sun Apr 02, 2006 8:45 pm
Location: Kelowna, BC Canada

Resmed Events vs AHI

Post by SleepDeep » Wed Feb 02, 2022 12:14 am

My ResMed S10 Auto is displaying 1.0 Events with a 1.8 AHI. That would seem impossible since the AHI is included in the event count and they are both calculated as an hourly number. Am I missing something?
ResMed AirSense 10 Autoset
P30i Mask

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed Events vs AHI

Post by palerider » Wed Feb 02, 2022 12:53 am

SleepDeep wrote: ↑
Wed Feb 02, 2022 12:14 am
My ResMed S10 Auto is displaying 1.0 Events with a 1.8 AHI. That would seem impossible since the AHI is included in the event count and they are both calculated as an hourly number. Am I missing something?
First, there's no such thing as a Resmed S10. What does it actually say on the machine?

Second, that's not at all impossible, since that AHI is an average over a period of time.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Resmed Events vs AHI

Post by Miss Emerita » Wed Feb 02, 2022 5:21 pm

I don't recall seeing an event count displayed on the machine itself. Where are you seeing it, and what exactly does it say?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

SleepDeep
Posts: 23
Joined: Sun Apr 02, 2006 8:45 pm
Location: Kelowna, BC Canada

Re: Resmed Events vs AHI

Post by SleepDeep » Wed Feb 02, 2022 6:30 pm

Sorry to introduce some confusion with my typo.
The machine in an Airsense 10 Auto

Yes I understand the AHI is an average, but wouldn't the Events per hour, of which AHI is an included event be an average as well? That is why I dont see how the Events per hour can be less than the AHI. Should be greater than or equal to.

As I indicated, I had 1.0 Events per hour and 1.8 AHI.

Miss Emerita, Events per Hour is the second line of the sleep report display.
ResMed AirSense 10 Autoset
P30i Mask

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed Events vs AHI

Post by Pugsy » Wed Feb 02, 2022 6:43 pm

SleepDeep wrote: ↑
Wed Feb 02, 2022 6:30 pm
Miss Emerita, Events per Hour is the second line of the sleep report display.
Which sleep report display?
The one from just last night or the one that has the default parameter of 30 days for the data reporting?

There is a sleep report just from last night and there is a sleep report with the default 30 day/one month average (can be changed to 1 day, 1 week and maybe something else but defaults back to one month if there is any sort of power disruption).

If it is the 30 day report...then the overall average could very well be different and would explain the obvious difference.

See page 18....down at the bottom
https://www.respshop.com/manuals/ResMed ... %20her.pdf

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

SleepDeep
Posts: 23
Joined: Sun Apr 02, 2006 8:45 pm
Location: Kelowna, BC Canada

Re: Resmed Events vs AHI

Post by SleepDeep » Wed Feb 02, 2022 8:02 pm

That was the problem. I had it set to daily reporting cycles, then I unplugged the unit when cleaning it. So it re-set to a 30 day number. Sure confused me. Thanks.
ResMed AirSense 10 Autoset
P30i Mask

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed Events vs AHI

Post by Pugsy » Wed Feb 02, 2022 8:17 pm

Any little power disruption will cause the re-set to the default.
Even a quick brown out...so just remember that if things start looking weird. Could happen when you aren't at home and wouldn't realize it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

SleepDeep
Posts: 23
Joined: Sun Apr 02, 2006 8:45 pm
Location: Kelowna, BC Canada

Re: Resmed Events vs AHI

Post by SleepDeep » Wed Feb 02, 2022 8:40 pm

Thanks. Bad design. Odd since other settings are permanently saved. :oops:
ResMed AirSense 10 Autoset
P30i Mask

User avatar
Pugsy
Posts: 65413
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed Events vs AHI

Post by Pugsy » Wed Feb 02, 2022 8:59 pm

Well....at least the pressure changes and other settings hold/stick with a power outage. As far as I know the default to a month for that data is all that changes.
I used to worry about it but now all I look at is the sleep report on the main screen that just shows AHI from last night and I don't ever go into that second deeper area anyway...so I don't bother changing it from the one month. We get brown outs quite often where I live and it was driving me crazy.

If I want to see more than what is available on the initial sleep report I just use OSCAR software anyway.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Resmed Events vs AHI

Post by Miss Emerita » Thu Feb 03, 2022 12:13 am

Aha, mystery solved by Pugsy Holmes!
Oscar software is available at https://www.sleepfiles.com/OSCAR/