is it likely I have sleep apnea based on these results?
What does this OSCAR data mean?
-
alienrayray
- Posts: 2
- Joined: Wed Dec 22, 2021 5:23 pm
What does this OSCAR data mean?
- Attachments
-
- ap.png (134.36 KiB) Viewed 5701 times
-
- ap2.png (20.29 KiB) Viewed 5701 times
Re: What does this OSCAR data mean?
Unless you tell us you have a lot of nasal congestion going on I would say that there is a very strong probability that you have some sort of sleep disordered breathing going on and obstructive sleep apnea would be at the top of the list.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: What does this OSCAR data mean?
You have to realize that without the machine reducing the events, your AHI (Events per hour) would have been much higher.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: What does this OSCAR data mean?
yes.alienrayray wrote: ↑Wed Dec 22, 2021 5:27 pmhttps://ibb.co/ZLQ2gzL
https://ibb.co/dBnRDf1
is it likely I have sleep apnea based on these results?
and welcome to the zoo!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
alienrayray
- Posts: 2
- Joined: Wed Dec 22, 2021 5:23 pm
Re: What does this OSCAR data mean?
thank you for the unfortunate news and kind welcome. do you think there is enough leaking to be concerned about my mask fit? should I raise the maximum pressure since i am often near the upper limit?
Re: What does this OSCAR data mean?
Not enough leak to worry about unless it wakes you up.
The machine can easily compensate for leaks up to at least 24 L/min and actually does a fair job up to a little over 30 L/min.
Yes, go ahead and increase the maximum. Obviously the machine wants to go upwards and is essentially setting a new minimum anyway. I would just set it to 20 cm and see where it goes. It won't go up without a good reason.
The machine can easily compensate for leaks up to at least 24 L/min and actually does a fair job up to a little over 30 L/min.
Yes, go ahead and increase the maximum. Obviously the machine wants to go upwards and is essentially setting a new minimum anyway. I would just set it to 20 cm and see where it goes. It won't go up without a good reason.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.