Re: On the verge of giving up

Post by Pugsy » Thu Dec 02, 2021 9:52 am

Please post a recent night detailed report from OSCAR showing the high AHI night.
Just one detailed report for now.
OSCAR https://www.sleepfiles.com/OSCAR/
http://www.apneaboard.com/wiki/index.ph ... stallation
http://www.apneaboard.com/wiki/index.ph ... rpretation


While getting all that set up..
Exactly what settings are you using right now? AHI last night? Settings?
Mode....
Minimum
Maximum
EPR on or off full time or ramp only...if on...at what setting.

Did you ever have a sleep study using the cpap machine and mask to determine pressures needed?
Was your diagnostic sleep study done in a sleep lab or home study?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

wano
Posts: 6
Joined: Thu Dec 02, 2021 8:34 am

Re: On the verge of giving up

Post by wano » Thu Dec 02, 2021 1:47 pm

[deleted]
Last edited by wano on Thu Dec 02, 2021 3:14 pm, edited 2 times in total.

wano
Posts: 6
Joined: Thu Dec 02, 2021 8:34 am

Re: On the verge of giving up

Post by wano » Thu Dec 02, 2021 2:01 pm

[deleted]
Last edited by wano on Thu Dec 02, 2021 3:14 pm, edited 2 times in total.

wano
Posts: 6
Joined: Thu Dec 02, 2021 8:34 am

Re: On the verge of giving up

Post by wano » Thu Dec 02, 2021 2:11 pm

[deleted]
Last edited by wano on Thu Dec 02, 2021 3:15 pm, edited 1 time in total.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: On the verge of giving up

Post by palerider » Thu Dec 02, 2021 2:50 pm


_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

wano
Posts: 6
Joined: Thu Dec 02, 2021 8:34 am

Re: On the verge of giving up

Post by wano » Thu Dec 02, 2021 3:01 pm

[deleted]
Last edited by wano on Thu Dec 02, 2021 3:14 pm, edited 1 time in total.

User avatar
Pugsy
Posts: 65122
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: On the verge of giving up

Post by Pugsy » Thu Dec 02, 2021 3:02 pm

So....more questions....please take the time to answer all of them

1...Do you have a copy of the original sleep study report? If not get a copy.
2...Was the sleep study done in a lab or at home?
3...How was it determined that you needed this setting? What other settings were tried?
4...Have you ever tried a pressure setting with EPR (exhale pressure relief) turned off? If so, can you share a detailed report from that night?
5...Have all your reports showed a high CA/central apnea count?
6...Were you awake very much the night of this report or did you sleep soundly?
7...Do you take any medications? If so, what.

I would like to try something simple to start off with that should be easy to do.
Leave it in fixed cpap mode for now.
Change pressure to 8 or 9 and turn EPR totally off.
If you simply can't breathe with EPR off...change it to ramp only and make use of the ramp feature.

I want to see if your centrals reduce in number with EPR off and with the lower pressure if the OAs increase or not.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65122
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: On the verge of giving up

Post by Pugsy » Thu Dec 02, 2021 3:04 pm

The AHI graph on the right is useless. Put the flow rate graph in it's place.
I have some examples here
viewtopic/t158560/How-to-post-images-for-review.html

Are those leaks waking you up?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

wano
Posts: 6
Joined: Thu Dec 02, 2021 8:34 am

Re: On the verge of giving up

Post by wano » Thu Dec 02, 2021 3:09 pm

[deleted]

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: On the verge of giving up

Post by palerider » Thu Dec 02, 2021 3:18 pm

Well, that was weird.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65122
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: On the verge of giving up

Post by Pugsy » Thu Dec 02, 2021 3:27 pm

palerider wrote:
Thu Dec 02, 2021 3:18 pm
Well, that was weird.
:shock: :shock: :shock: What the hell?
I guess I wasted my time....oh well...not the first time and won't be the last.

I guess he really wants to quit and not get help....wanted us to tell him "there, there, it's too hard so you can just quit"...fat chance of that happening.
I wish people who come here to whine and say they are quitting to just quit and not waste our time...we don't need to hear it.

I will remove the entire thread later after everyone gets a chance to have their say about things.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Janknitz
Posts: 8511
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: On the verge of giving up

Post by Janknitz » Thu Dec 02, 2021 3:42 pm

I often start my answers to people like that as follows: "Do you just want to vent or do you want help? If you want help, then tell us . . ." Some people DO want help (I did when I posted my initial whining) and some just want to vent and complain and hope they get "permission" to quit. Sometimes THEY don't know which they want until we ask.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
zonker
Posts: 11339
Joined: Fri Jun 19, 2015 4:36 pm

Re: On the verge of giving up

Post by zonker » Thu Dec 02, 2021 4:46 pm

Image
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
CyberSpiff
Posts: 59
Joined: Tue Sep 21, 2021 1:47 pm

Re: On the verge of giving up

Post by CyberSpiff » Thu Dec 02, 2021 5:14 pm

It’s like me Dad used to say. “What the hell are you doing in the loo for so long? Get out of there and give someone else a chance.”

(Nod to late great Marty Feldman)
There is no right or wrong side of the road. There is only the road.