NEW Machine

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
zoophaugus
Posts: 10
Joined: Tue Sep 02, 2014 4:19 pm
Location: Greater Toronto Area (GTA) Ontario, Canada

NEW Machine

Post by zoophaugus » Wed Nov 03, 2021 10:33 am

Hi all,
I am in the market for a new auto bilevel machine. My machine is fairly old - An S9 VPAP auto.
Is it possible to get into clinician setup on these newer machines to make adjustments your self like on the older series of machines?
I remember reading long time ago all the new machines would be wirelessly connected to the provider with RX info and controlled on provider end.
Is this true?
If so is there a work around?
I like to be able to make adjustments on my own.

Im looking at philps respironic or resmed machine leaning towards PR as last time i checked their machines run on 12 volts. I go camping some times and can bring a marine battery to power the machine.. nice and easy no inverter required.. :)

Any advise would be greatly appreciated.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: PR-System One (60 Series) BiPAP Auto SV Advance

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: NEW Machine

Post by Miss Emerita » Wed Nov 03, 2021 11:56 am

I warmly recommend the ResMed VAuto. Yes, it is easy to change the settings yourself. You can put the machine in airplane mode if you don't want data to be sent anywhere by wireless, though your provider may actually not care whether you change the settings yourself.

The PR bilevel is under a big recall effort right now, due to potential health problems with the sound abatement foam inside them.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zoophaugus
Posts: 10
Joined: Tue Sep 02, 2014 4:19 pm
Location: Greater Toronto Area (GTA) Ontario, Canada

Re: NEW Machine

Post by zoophaugus » Thu Nov 04, 2021 8:44 am

Thank you for the reply. It appears as though all the PR machines are sold out any ways. My main concern was being able to take control of the machine and you answered my question.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: PR-System One (60 Series) BiPAP Auto SV Advance

User avatar
babydinosnoreless
Posts: 2372
Joined: Fri Nov 02, 2018 2:53 pm

Re: NEW Machine

Post by babydinosnoreless » Thu Nov 04, 2021 8:50 am

I've heard the new dreamstation 2 does not work with the oscar software. I have an aircurve 10 vauto which does and I can get in and change stuff myself.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: NEW Machine

Post by chunkyfrog » Thu Nov 04, 2021 8:56 am

babydinosnoreless wrote:
Thu Nov 04, 2021 8:50 am
I've heard the new dreamstation 2 does not work with the oscar software. I have an aircurve 10 vauto which does and I can get in and change stuff myself.
Our advice--stick with Resmed machines.
It looks like respironics intends to withhold info/control from the patient.
Avoid their stuff.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
babydinosnoreless
Posts: 2372
Joined: Fri Nov 02, 2018 2:53 pm

Re: NEW Machine

Post by babydinosnoreless » Thu Nov 04, 2021 8:57 am

chunkyfrog wrote:
Thu Nov 04, 2021 8:56 am
babydinosnoreless wrote:
Thu Nov 04, 2021 8:50 am
I've heard the new dreamstation 2 does not work with the oscar software. I have an aircurve 10 vauto which does and I can get in and change stuff myself.
Our advice--stick with Resmed machines.
It looks like respironics intends to withhold info/control from the patient.
Avoid their stuff.
Agreed!

User avatar
zoophaugus
Posts: 10
Joined: Tue Sep 02, 2014 4:19 pm
Location: Greater Toronto Area (GTA) Ontario, Canada

Re: NEW Machine

Post by zoophaugus » Thu Nov 04, 2021 11:47 am

Just to be sure of the function this machine,

(ResMed AirCurve™ 10 VAuto w/HumidAir™ BiLevel)

****Has a fixed pressure support with an auto ranging lower and upper limit of base pressure? Is this correct? ***** (Example Base pressure 8 - 16 with pressure support of 4)

This is how my current S9 Vpap auto works

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: PR-System One (60 Series) BiPAP Auto SV Advance

User avatar
Pugsy
Posts: 65258
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: NEW Machine

Post by Pugsy » Thu Nov 04, 2021 12:18 pm

zoophaugus wrote:
Thu Nov 04, 2021 11:47 am
Just to be sure of the function this machine,

(ResMed AirCurve™ 10 VAuto w/HumidAir™ BiLevel)

****Has a fixed pressure support with an auto ranging lower and upper limit of base pressure? Is this correct? ***** (Example Base pressure 8 - 16 with pressure support of 4)

This is how my current S9 Vpap auto works
Correct. The AirCurve 10 VAuto will work just like your current S9 VPap auto works.
I have used the S9 VPAP Auto and currently use the AirCurve 10 VAuto.

I know the lure of 12 volt with Respironics is there but even if the DreamStation 2 wasn't so difficult to get data or change your own settings.....Respironics hasn't yet even released a DreamStation 2 Auto BiPap model. The machine you would need isn't even available yet.

Be happy with the ResMed AirCurve 10 VAuto...it's a great machine and essentially does exactly what the S9 does only better. If you want to go camping and need to go with 12 volt...there are ways to get it done. Not quite as convenient and just a bit more costly but will be worth it in the long run.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zoophaugus
Posts: 10
Joined: Tue Sep 02, 2014 4:19 pm
Location: Greater Toronto Area (GTA) Ontario, Canada

Re: NEW Machine

Post by zoophaugus » Thu Nov 04, 2021 12:35 pm

Again thanks for the reply.

I do have have a PR system1 Auto SV Advance in the closet which can be used for camping duties. I found it was a little too aggressive for me and preferred the S9 Vpap Auto and its more basic functionality to that of the servo ventilator.

The convenience of the 12 volts was great.. just bring a heavy duty marine battery and your good to go for days.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: PR-System One (60 Series) BiPAP Auto SV Advance

User avatar
Pugsy
Posts: 65258
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: NEW Machine

Post by Pugsy » Thu Nov 04, 2021 12:58 pm

You can dumb the Respironics ASV down a bit and make it a bit less aggressive if you wish but please be aware that it likely is part of the recent Respironics recall.

https://www.cpaptalk.com/viewtopic/t182 ... ation.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zoophaugus
Posts: 10
Joined: Tue Sep 02, 2014 4:19 pm
Location: Greater Toronto Area (GTA) Ontario, Canada

Re: NEW Machine

Post by zoophaugus » Fri Dec 03, 2021 3:48 pm

Well I jumped through all the hoops with insurance for the last small amount they would pay for and it's all approved. Our provincial government insurance covers most at 70%.
So there some relief on my end but now I am on a waiting list for this new machine with no idea of when it would arrive.
A few days ago my Resmed Vpap auto starts making a strange motor noise. It's constant and gets louder with each breath. The noise is bothering my wife on the other side of our king size bed as well as me right beside the machine. Sounds like a bearing on the motor is gone.

So I had no choice other than to pull out my trusty PR SystemONE auto Sv advance.

Funny thing the Resmed has only 7k hours on it and the bearing is gone meanwhile I've dragged the PR machine around the globe with me numerous times which includes it being in a backpack while I've crossed rivers on bamboo bridges and it has 14k hours and it still works great.
Go figure huh.
No sure if I should worry about this sound abatement foam. I don't recall ever seeing anything in the water tank. I've never used any of those cleaning things on the machine and for the most part it's always been used in climate controlled sleep settings aside from a few trips to various place in the world. The machine is under recall as I've checked and registered it. But I'm not holding my breath and waiting on anything from PR.
I'm not sure what the big deal is . Any chunks or particles would first land within the water in the tank and Then have to get the lung mucus and the lungs would clean themselves as normal. Off gassing? I'm sure that's all done long ago before 14k hours on the machine.

Any way I have a basic S9 Cpap machine. Can I take the motor out of that and put into S9 VPAP Auto? Has any one done this before?
Any instructions?

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: PR-System One (60 Series) BiPAP Auto SV Advance

User avatar
zoophaugus
Posts: 10
Joined: Tue Sep 02, 2014 4:19 pm
Location: Greater Toronto Area (GTA) Ontario, Canada

Re: NEW Machine

Post by zoophaugus » Fri Dec 03, 2021 7:26 pm

Forget about my last 2 questions. I came a cross the thread titled "Resmed S9 Elite Tear down with pictures".
Instead of swapping motors I just swapped the circuit board and top cover from the VPAP Auto to the S9 Elite. No more motor noise. Technically I did swap motors..
Easy peasy and took about an hour.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: PR-System One (60 Series) BiPAP Auto SV Advance

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: NEW Machine

Post by chunkyfrog » Fri Dec 03, 2021 9:07 pm

Well done!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her