Flow Limitation
Flow Limitation
A quick question about flow limitation events. I just watched a youtube video recommended by someone here which explains all the common respiratory sleep events, including flow limitation. The video specifically referred to a "chair like" shape to the breathing pattern during a flow limitation event. My Devilbiss/Intellipap machine show me having frequent flow limitations but the wave pattern is nothing like the chair shape. Below are two screenshots, the first a detail from my Oscar data last night, the second from the video. I strongly suspect that my CPAP machine is flagging false flow limitation events. Any thoughts please?
_________________
| Mask: AirFit™ N10 Nasal CPAP Mask with Headgear |
| Additional Comments: My machine is sold in Europe as Devilbliss Blue Auto Plus Automatic |
- Attachments
-
- Screen Shot 2021-11-02 at 16.45.38.png (94.38 KiB) Viewed 4351 times
-
- Screen Shot 2021-11-02 at 16.55.29.png (182.73 KiB) Viewed 4351 times
- BlueDragon
- Posts: 563
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: Flow Limitation
Your OSCAR chart shows typical flow limitations -- the inhalation portion of your breath has a flat top, suggesting that inhalation rate was limited. "Normal" breath pattern shows a rounded top.
The "chair" diagram seems to be less common, at least in my experience. It suggests that inhalation started at full volume and then was restricted. I think usually it is restricted in the first place so you don't get the "back of the chair" part.
The "chair" diagram seems to be less common, at least in my experience. It suggests that inhalation started at full volume and then was restricted. I think usually it is restricted in the first place so you don't get the "back of the chair" part.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
Last edited by BlueDragon on Tue Nov 02, 2021 11:04 am, edited 1 time in total.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
- Miss Emerita
- Posts: 3783
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Flow Limitation
MMcG, the images aren't there. Could you try again? Meanwhile, here are images of flow limited inspiration:
On the far left is a normal trace; then we see the variety of flow-limited breaths. I think of them as various combinations of dents, humps, and plateaus. My machine will also sometimes flag small-sized inhalations as flow-limited.
On the far left is a normal trace; then we see the variety of flow-limited breaths. I think of them as various combinations of dents, humps, and plateaus. My machine will also sometimes flag small-sized inhalations as flow-limited.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
- Dog Slobber
- Posts: 4332
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Flow Limitation
When asking for comments related to some other resource (video, webpage, post, etc) you're better off linking to the resource so we can also see exactly what was said, the context and the credibility of the resource.
While a chair-like trace is a classic look for flow limited breathing, it's *not* the only pattern. A flattened top instead of a rounded peak is also an indicator of flow limitations.
I would not consider your flow limitations to be false positives.
While a chair-like trace is a classic look for flow limited breathing, it's *not* the only pattern. A flattened top instead of a rounded peak is also an indicator of flow limitations.
I would not consider your flow limitations to be false positives.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
- Miss Emerita
- Posts: 3783
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Flow Limitation
Apologies: Pugsy pointed out she could see the attachments, plus a comment from Blue Dragon. I looked again, and there there were! Plus now a comment from DS.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: Flow Limitation
Thank you. Maybe I should look at increasing my minimum pressure then? Pugsy here in reply to someone recommended the youtube video. Here's the link:- https://www.youtube.com/watch?v=-gie2dhqP2cDog Slobber wrote: ↑Tue Nov 02, 2021 11:09 amWhen asking for comments related to some other resource (video, webpage, post, etc) you're better off linking to the resource so we can also see exactly what was said, the context and the credibility of the resource.
While a chair-like trace is a classic look for flow limited breathing, it's *not* the only pattern. A flattened top instead of a rounded peak is also an indicator of flow limitations.
I would not consider your flow limitations to be false positives.
_________________
| Mask: AirFit™ N10 Nasal CPAP Mask with Headgear |
| Additional Comments: My machine is sold in Europe as Devilbliss Blue Auto Plus Automatic |
Re: Flow Limitation
Looks like I need to try to limit them then. Thanks.BlueDragon wrote: ↑Tue Nov 02, 2021 11:00 amYour OSCAR chart shows typical flow limitations -- the inhalation portion of your breath has a flat top, suggesting that inhalation rate was limited. "Normal" breath pattern shows a rounded top.
The "chair" diagram seems to be less common, at least in my experience. It suggests that inhalation started at full volume and then was restricted. I think usually it is restricted in the first place so you don't get the "back of the chair" part.
_________________
| Mask: AirFit™ N10 Nasal CPAP Mask with Headgear |
| Additional Comments: My machine is sold in Europe as Devilbliss Blue Auto Plus Automatic |
Re: Flow Limitation
Looks like mine are of the far right type. Thanks.Miss Emerita wrote: ↑Tue Nov 02, 2021 11:02 amMMcG, the images aren't there. Could you try again? Meanwhile, here are images of flow limited inspiration:
Flow images for FL.jpg
On the far left is a normal trace; then we see the variety of flow-limited breaths. I think of them as various combinations of dents, humps, and plateaus. My machine will also sometimes flag small-sized inhalations as flow-limited.
_________________
| Mask: AirFit™ N10 Nasal CPAP Mask with Headgear |
| Additional Comments: My machine is sold in Europe as Devilbliss Blue Auto Plus Automatic |
- BlueDragon
- Posts: 563
- Joined: Sun May 27, 2018 10:20 am
- Location: Sonoran Desert
Re: Flow Limitation
DeVilbiss BLUE machines report flow limits at three levels: Mild, Moderate, and Severe. Flow limits classified as "severe" are flagged as events and show up in your Events graph. OSCAR graphs all three levels of flow limits in the Flow Limits graph. So you should see a FL event only when the graph shows a Severe flow limit. (Mild, moderate, and severe flow limits have graph values of 1, 2, or 3 respectively.)
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020. |
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.
- Dog Slobber
- Posts: 4332
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Flow Limitation
I'm familiar with the video.MMcG wrote: ↑Tue Nov 02, 2021 11:54 amThank you. Maybe I should look at increasing my minimum pressure then? Pugsy here in reply to someone recommended the youtube video. Here's the link:- https://www.youtube.com/watch?v=-gie2dhqP2cDog Slobber wrote: ↑Tue Nov 02, 2021 11:09 amWhen asking for comments related to some other resource (video, webpage, post, etc) you're better off linking to the resource so we can also see exactly what was said, the context and the credibility of the resource.
While a chair-like trace is a classic look for flow limited breathing, it's *not* the only pattern. A flattened top instead of a rounded peak is also an indicator of flow limitations.
I would not consider your flow limitations to be false positives.
And yes, you interpretted it incorrectly. The video did reference a Chair-Like shape, but also referenced a flattening. You can't interpret that as, it must or always is chair-like.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Flow Limitation
The image presented by Miss Emerita is found in the article at: https://www.sciencedirect.com/science/a ... 516#bbib30Dog Slobber wrote: ↑Tue Nov 02, 2021 11:09 amWhen asking for comments related to some other resource (video, webpage, post, etc) you're better off linking to the resource so we can also see exactly what was said, the context and the credibility of the resource.
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
Re: Flow Limitation
Another source to consider is:
http://www.apneaboard.com/wiki/index.ph ... mitations authored by the OSCAR team.
The Expiratory mouth breathing with normal inspiratory flow looks like the reverse chair trace.
I'm going to go out on a limb here but I'd bet that the expiratory mouth breathing with normal inspiration is what might expect if the patient were inhaling thru the nose and exhaling thru the mouth while using a nasal mask
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
Re: Flow Limitation
See the thread I started on Oct 19 2021 titled "Pressure Response on AirSense 11"
about half way down the page we get into a discussion of flow limitations.
I've also sent you a new private message
about half way down the page we get into a discussion of flow limitations.
I've also sent you a new private message
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
Re: Flow Limitation
Will do. Thanks.
_________________
| Mask: AirFit™ N10 Nasal CPAP Mask with Headgear |
| Additional Comments: My machine is sold in Europe as Devilbliss Blue Auto Plus Automatic |
Re: Flow Limitation
Yes, I use a nasal pillow and I know from the wave pattern that I am breathing in through the nose and exhaling through the mouth. This varies a lot. Some nights it can only be for a few minutes in total, but on other occasions I've kept up for forty-five minutes or more and done that more than once during the night. I have EPR on my Devilbiss/Intellipap machine but my sleep technician has recommended against using it. I think his logic is that if I can adapt and outgrow the exhale puffing (I am doing a lot less of it after 7-8 weeks on CPAP therapy), it will be better in the long run than to become reliant on EPR. I can see logically how it might discourage exhale puffing, but in the example on "apnea board", it suggests it also helps with flow limitation too. I'm not clear as the why this would be so, because surely EPR does nothing for inhaling? If I knew why it might solve both problems, I'd discuss it with my technician and give it a try.dataq1 wrote: ↑Tue Nov 02, 2021 1:23 pmAnother source to consider is:
http://www.apneaboard.com/wiki/index.ph ... mitations authored by the OSCAR team.
The Expiratory mouth breathing with normal inspiratory flow looks like the reverse chair trace.
I'm going to go out on a limb here but I'd bet that the expiratory mouth breathing with normal inspiration is what might expect if the patient were inhaling thru the nose and exhaling thru the mouth while using a nasal mask
_________________
| Mask: AirFit™ N10 Nasal CPAP Mask with Headgear |
| Additional Comments: My machine is sold in Europe as Devilbliss Blue Auto Plus Automatic |