Airsense 10 screen keeps turning on

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Airsense 10 screen keeps turning on

Post by ILoveFlowers » Thu Oct 28, 2021 1:39 pm

I noticed during the night that the screen on my machine keeps turning on for brief periods of time, the light wakes me up. I woke up and got my glasses to see what the problem is, it's just displaying a message that the heated tubing is connected. Anyone else have this happen? Is it malfunctioning? Is there some setting to stop it from doing this?
AirSense 10 AutoSet For Her
Airfit P10 mask

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Airsense 10 screen keeps turning on

Post by Okie bipap » Thu Oct 28, 2021 2:12 pm

Check your hose to make sure it is pushed in securely. It may be breaking and remaking electrical contact during the night. The only time I have ever seen that message is when I first plugged in the hose.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
Dog Slobber
Posts: 4325
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Airsense 10 screen keeps turning on

Post by Dog Slobber » Thu Oct 28, 2021 2:49 pm

Try one or more of the following:
  1. Software glitch
    Reset device by unplugging for a minute and restarting
  2. Heat Hose not seated properly
    Unplug and plug back in firmly
    Check/clean contacts
  3. Heated Hose - wearing out
    Replace with your spare heated hose. You do have a spare don't you?
  4. Heated Hose circuitry problem
    Repair or use non-heated hoses
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11437
Joined: Fri Jun 19, 2015 4:36 pm

Re: Airsense 10 screen keeps turning on

Post by zonker » Thu Oct 28, 2021 3:35 pm

Dog Slobber wrote:
Thu Oct 28, 2021 2:49 pm


[*]Heated Hose - wearing out
Replace with your spare heated hose. You do have a spare don't you?