Is my cpap working correctly resmed airsense 10
-
oledad1978
- Posts: 25
- Joined: Thu Oct 28, 2021 12:24 pm
Re: Is my cpap working correctly resmed airsense 10
Whats the problem and i dint have a laptop to access oscar or sd card inserted
- Miss Emerita
- Posts: 3766
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Is my cpap working correctly resmed airsense 10
It is just fine if you don’t feel resistance. So if that’s what you were worried about, don’t worry.oledad1978 wrote: ↑Fri Oct 29, 2021 8:26 amI do not feel hardly any resistancr when exhaling but ive been using my cpap for a few months
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: Is my cpap working correctly resmed airsense 10
If you'd discover the wonders of punctuation, your posts would tend to make more sense.oledad1978 wrote: ↑Fri Oct 29, 2021 7:54 pmWhats the problem and i dint have a laptop to access oscar or sd card inserted
The "problem" is that only about 1% of CPAP users are properly treated with settings like you have on your machine.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
oledad1978
- Posts: 25
- Joined: Thu Oct 28, 2021 12:24 pm
Re: Is my cpap working correctly resmed airsense 10
........
Moderator note....potty mouth profanity not allowed here...and got deleted....hence the dots above.
Made me cringe and I am no angel with the potty mouth syndrome.
Moderator note....potty mouth profanity not allowed here...and got deleted....hence the dots above.
Made me cringe and I am no angel with the potty mouth syndrome.
-
oledad1978
- Posts: 25
- Joined: Thu Oct 28, 2021 12:24 pm
Re: Is my cpap working correctly resmed airsense 10
Ahhhh no potty mouth but u allow grammar police this is not english class read it slower i guess but the rest of u not on the grammar and punctuation police thank you for the advice
Re: Is my cpap working correctly resmed airsense 10
Read the user agreement you had to agree to to join the forum.oledad1978 wrote: ↑Sat Oct 30, 2021 3:14 pmAhhhh no potty mouth but u allow grammar police this is not english class read it slower i guess but the rest of u not on the grammar and punctuation police thank you for the advice
Profanity is not allowed. A link to it is at the bottom of the main forum page.
There's no restriction about grammar police though...and the suggestions by the grammar police were made so that people could understand what your problem is/was.
It's really hard for people to answer a question when they don't understand the question because someone isn't clear with their communication.
The grammar police was just trying to help you.
Fat chance on that occurring now.
Have a nice day.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Is my cpap working correctly resmed airsense 10
Thanks for making it clear that you should be on the ignore list.oledad1978 wrote: ↑Sat Oct 30, 2021 3:14 pmAhhhh no potty mouth but u allow grammar police this is not english class read it slower i guess but the rest of u not on the grammar and punctuation police thank you for the advice
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
oledad1978
- Posts: 25
- Joined: Thu Oct 28, 2021 12:24 pm
Re: Is my cpap working correctly resmed airsense 10
No no thank you
-
oledad1978
- Posts: 25
- Joined: Thu Oct 28, 2021 12:24 pm
Re: Is my cpap working correctly resmed airsense 10
All u had to do was ask about what u wanted to know im on here seeking help then u point out grammar and only question i was asked to begin with was max and min pressure and i said what it was then u said bunch of other stuff but i hope u can be more specific with other people seeking advice i hope neither of u r real medical professionals so with that being said i will not be back on here everyone that gave advice without making me feel stupid thank u all and good luck with all your cpap problems
Re: Is my cpap working correctly resmed airsense 10
ok, so let's go back to this. i think miss e and pugsy may be on to something here.oledad1978 wrote: ↑Fri Oct 29, 2021 8:26 amI do not feel hardly any resistancr when exhaling but ive been using my cpap for a few months
when you first started on the machine, are you saying you felt resistance on exhaling but that you don't now?
it's possible that your body has grown accustomed to the ebb and flow of breathing with a machine. that is, it feels more natural to you.
if so, this is quite a common occurrence and shouldn't bother you.
what is your current ahi? that can be read on the machine without having to mess with anything. if your ahi is acceptable, i don't see any problems. if your ahi is high, however, that's not good. we could help you with that, but you are reluctant to make changes yourself.
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
oledad1978
- Posts: 25
- Joined: Thu Oct 28, 2021 12:24 pm
Re: Is my cpap working correctly resmed airsense 10
No i started making changes my ahi was 6.8 and then i reset my data but last nights ahi 9.3 but only used it 3 hrs and then i changed tje setting from cpap to autoset then then my pressure went from 10 to 11.2
Re: Is my cpap working correctly resmed airsense 10
i'm confused. you said here-oledad1978 wrote: ↑Sat Oct 30, 2021 5:14 pmNo i started making changes my ahi was 6.8 and then i reset my data but last nights ahi 9.3 but only used it 3 hrs and then i changed tje setting from cpap to autoset then then my pressure went from 10 to 11.2
are YOU making changes or is the doctor/dme?oledad1978 wrote: ↑Thu Oct 28, 2021 10:34 pmOh no not allowed to change settings have to take it back to the provider to get it adjusted but i can access the clinical settings
btw, i would strongly urge you to download the software program oscar and start using it. if you do not have an sd card slot in your laptop, i would suggest you get an sd card reader which will connect to your laptop via usb. we can't just keep making guesses about what is going on. we need some data so we can help out properly.
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Is my cpap working correctly resmed airsense 10
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!

--
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
--
Re: Is my cpap working correctly resmed airsense 10
Oh Oh, I think I made the list!
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
- Attachments
-
- Francis-1.jpg (44.58 KiB) Viewed 505 times
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!

--
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
--
Re: Is my cpap working correctly resmed airsense 10

_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!

--
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
--