Can artificial light damage mask?

Post by TheClerk » Mon Oct 25, 2021 2:46 pm

Sorry for the silly question but, can artificial light (i.e. a lamp) damage parts of a CPAP mask?

Here is what happened: Yesterday I went to sleep and put on my Resmed P10 that I have been using for a few days short of a full year.

Soon enough I had to take off my mask as I was not getting enough air and, even worse, exhaling was quite a chore.

Tried again for a few times and, then decided to try and check and clean the cushion and diffuser.

I realized that the diffuser part was a bit black. At first I though it was full of dust, although it didn't make sense. It's not dust; it really look like if it had been heated.

Lucky enough I had purchased a spare P10 a while back (the insurance only covers new stuff every 12 months).

This morning, as I was doing my usual routine of taking off my mask, and putting it on my night table, I realized that, for the past week, I have been turning my table lamp on upon waking up (it's darker than it used to in the morning). I left the lamp on for the entire day.

The way I put my mask on the desk, with the headstrap dangling from the side, the diffuser is directly in sight of the light, about a foot above.

So, I'm really curious; is such an exposure enough to "burn" the P10 diffuser? I can't think this is normal "wear"...

Thanks in advance.

User avatar
Pugsy
Posts: 65250
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can artificial light damage mask?

Post by Pugsy » Mon Oct 25, 2021 2:50 pm

No....your lamp isn't the cause of the dark discolorations on the P10 vent diffuser area.
I had that happen a while back and my P10 was never exposed to artificial or natural light.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
TheClerk
Posts: 32
Joined: Sat Oct 31, 2020 6:46 am
Location: Québec, Canada

Re: Can artificial light damage mask?

Post by TheClerk » Mon Oct 25, 2021 3:03 pm

Ok, thanks... weird that I had not noticed any degradation before, at least not to the point I wasn't able to breathe with it anymore.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can artificial light damage mask?

Post by palerider » Mon Oct 25, 2021 3:10 pm

TheClerk wrote:
Mon Oct 25, 2021 3:03 pm
Ok, thanks... weird that I had not noticed any degradation before, at least not to the point I wasn't able to breathe with it anymore.
You're supposed to clean the mesh on a regular basis, with dishwashing soap and an old toothbrush.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11421
Joined: Fri Jun 19, 2015 4:36 pm

Re: Can artificial light damage mask?

Post by zonker » Mon Oct 25, 2021 4:08 pm

TheClerk wrote:
Mon Oct 25, 2021 3:03 pm
Ok, thanks... weird that I had not noticed any degradation before, at least not to the point I wasn't able to breathe with it anymore.
to add to what palerider said, a water flosser works very well for this.

(tip of the hat to forum friend chunkyfrog)

just be sure you use it in the shower as that pulsing water can bounce off the mesh and go everywhere!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
TheClerk
Posts: 32
Joined: Sat Oct 31, 2020 6:46 am
Location: Québec, Canada

Re: Can artificial light damage mask?

Post by TheClerk » Mon Oct 25, 2021 5:00 pm

palerider wrote:
Mon Oct 25, 2021 3:10 pm
TheClerk wrote:
Mon Oct 25, 2021 3:03 pm
Ok, thanks... weird that I had not noticed any degradation before, at least not to the point I wasn't able to breathe with it anymore.
You're supposed to clean the mesh on a regular basis, with dishwashing soap and an old toothbrush.

I used to simply wash it by hand and under running water with dish soap; didn't realize it needed a tootbrush action or something like it.

I will try that and see if I can get it usable for a little longer. Although I had a backup and I can renew my stuff later this week... Just need to decide on replacement. I like the P10 but, might want to try the N20, N30 or Evora..

Thanks everyone; I know more now than yesterday!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can artificial light damage mask?

Post by palerider » Mon Oct 25, 2021 6:21 pm

TheClerk wrote:
Mon Oct 25, 2021 5:00 pm
palerider wrote:
Mon Oct 25, 2021 3:10 pm
TheClerk wrote:
Mon Oct 25, 2021 3:03 pm
Ok, thanks... weird that I had not noticed any degradation before, at least not to the point I wasn't able to breathe with it anymore.
You're supposed to clean the mesh on a regular basis, with dishwashing soap and an old toothbrush.
I used to simply wash it by hand and under running water with dish soap; didn't realize it needed a tootbrush action or something like it.
https://www.youtube.com/watch?v=MylY6a6KXMY

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
TheClerk
Posts: 32
Joined: Sat Oct 31, 2020 6:46 am
Location: Québec, Canada

Re: Can artificial light damage mask?

Post by TheClerk » Mon Oct 25, 2021 6:32 pm

palerider wrote:
Mon Oct 25, 2021 6:21 pm
TheClerk wrote:
Mon Oct 25, 2021 5:00 pm
palerider wrote:
Mon Oct 25, 2021 3:10 pm
TheClerk wrote:
Mon Oct 25, 2021 3:03 pm
Ok, thanks... weird that I had not noticed any degradation before, at least not to the point I wasn't able to breathe with it anymore.
You're supposed to clean the mesh on a regular basis, with dishwashing soap and an old toothbrush.
I used to simply wash it by hand and under running water with dish soap; didn't realize it needed a tootbrush action or something like it.
https://www.youtube.com/watch?v=MylY6a6KXMY
Thanks again!

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Can artificial light damage mask?

Post by chunkyfrog » Tue Oct 26, 2021 11:37 am

The water flosser works quite well, as does a fizzy denture tablet--
---But it takes a lot of rinsing to get the minty smell out.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: Can artificial light damage mask?

Post by jdm2857 » Tue Oct 26, 2021 3:56 pm

If you happen to have one, an ultrasonic cleaner is great for cleaning things like mask parts.Just fill with water, add a drop or two of dishwashing liquid, drop in the parts and turn it on. A quick rinse afterwards and you're done.

Really vibrates the dirt out of holes and crevices.

I happen to have a cheap one purchased many years ago. I think that Harbor Freight has one for ~$80.
jeff

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can artificial light damage mask?

Post by palerider » Tue Oct 26, 2021 4:03 pm

jdm2857 wrote:
Tue Oct 26, 2021 3:56 pm
If you happen to have one, an ultrasonic cleaner is great for cleaning things like mask parts.Just fill with water, add a drop or two of dishwashing liquid, drop in the parts and turn it on. A quick rinse afterwards and you're done.

Really vibrates the dirt out of holes and crevices.

I happen to have a cheap one purchased many years ago. I think that Harbor Freight has one for ~$80.
The *ONLY* "cpap cleaner' device that actually cleans anything.

However, don't buy the one that looks like someone took the harbor freight one and doubled (or more) the price after slapping "CPAP" on it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Can artificial light damage mask?

Post by chunkyfrog » Thu Oct 28, 2021 9:03 am

You KNOW somebody has done that.
Like unscented baby wipes--same thing as "cpap wipes", newly relabeled,
and re-priced!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can artificial light damage mask?

Post by palerider » Thu Oct 28, 2021 12:16 pm

chunkyfrog wrote:
Thu Oct 28, 2021 9:03 am
You KNOW somebody has done that.
Like unscented baby wipes--same thing as "cpap wipes", newly relabeled,
and re-priced!
Oh yes.
There's one for "only" 380 on amazon, or the smaller one for $289.

The one at Harbor Freight for $80 would probably work just as well. :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Can artificial light damage mask?

Post by chunkyfrog » Fri Oct 29, 2021 12:26 pm

Ya gotta know your target market, aka "mark". :mrgreen:
Extrapolating that OSA can negatively impact intelligence . . .
Conveniently forgetting that treatment can minimize the effects.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her