Should I buy a used Resmed AirCurve 10 ASV for $200?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
vilnius
Posts: 19
Joined: Mon Sep 20, 2021 5:01 pm

Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by vilnius » Tue Sep 21, 2021 3:01 pm

I have a Fisher & Paykel Icon+ Auto CPAP machine that I bought used for $100 about a month ago. However, I just realized that it's not supported by OSCAR. It also doesn't feel as good as the machine I tried at my sleep study did. The air feels more restricted. I got diagnosed with mild sleep apnea (8 AHIs on my side, 12 on my back). I'm a newb to the world of CPAPs and OSCAR.

The guy I bought the F&P from is selling a Resmed AirCurve 10 ASV. I could trade in my F&P and $100 to get the AirCurve 10. Should I do it? What should I look for when testing the device? He says the machine is in excellent shape and has low hours.

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by JLROhio » Tue Sep 21, 2021 3:17 pm

I'm not an expert in your question (but they'll be along shortly)...

Some questions that I have initially though, as I read your comments are:

- Does the F&P have some sort of EPR setting that the ResMed has?

When I initially started using my ResMed I felt that I could not exhale (and thus had trouble with a full breath) and felt restricted.
Once I turned on the EPR...breathing was so much better and I've enjoyed my ResMed for years!

- The seller indicated "low hours" to you. That's a subjective term (such as I have a low mileage Corvette here - only 150k miles!)

Do you know the actual hours on the machine? If not, ask and post....that way, others can better evaluate a great price based upon the hours.


Good Luck!
Last edited by JLROhio on Tue Sep 21, 2021 4:14 pm, edited 1 time in total.
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
zonker
Posts: 11341
Joined: Fri Jun 19, 2015 4:36 pm

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by zonker » Tue Sep 21, 2021 3:30 pm

vilnius wrote:
Tue Sep 21, 2021 3:01 pm
I have a Fisher & Paykel Icon+ Auto CPAP machine that I bought used for $100 about a month ago. However, I just realized that it's not supported by OSCAR. It also doesn't feel as good as the machine I tried at my sleep study did. The air feels more restricted. I got diagnosed with mild sleep apnea (8 AHIs on my side, 12 on my back). I'm a newb to the world of CPAPs and OSCAR.

The guy I bought the F&P from is selling a Resmed AirCurve 10 ASV. I could trade in my F&P and $100 to get the AirCurve 10. Should I do it? What should I look for when testing the device? He says the machine is in excellent shape and has low hours.
is this person near by? take a picture of the screen display where it shows run hours. post that here and the experts can tell you if they are truly low run hours.

more than anything else, you're going to need a machine supported by oscar.

if this works out, this will be a hell of a deal for you all the way 'round!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by chunkyfrog » Tue Sep 21, 2021 6:11 pm

Yes, get the run hours.
"Low" is very subjective.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by palerider » Tue Sep 21, 2021 6:26 pm

vilnius wrote:
Tue Sep 21, 2021 3:01 pm
I have a Fisher & Paykel Icon+ Auto CPAP machine that I bought used for $100 about a month ago. However, I just realized that it's not supported by OSCAR. It also doesn't feel as good as the machine I tried at my sleep study did. The air feels more restricted. I got diagnosed with mild sleep apnea (8 AHIs on my side, 12 on my back). I'm a newb to the world of CPAPs and OSCAR.

The guy I bought the F&P from is selling a Resmed AirCurve 10 ASV. I could trade in my F&P and $100 to get the AirCurve 10. Should I do it? What should I look for when testing the device? He says the machine is in excellent shape and has low hours.
An ASV can be challenging to sleep with until you get used to it, (unless you drop right off to sleep) but that said, that's an absolutely KILLER deal (3 grand, or more, new), unless it's got a ton of hours on it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

vilnius
Posts: 19
Joined: Mon Sep 20, 2021 5:01 pm

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by vilnius » Tue Sep 21, 2021 9:18 pm

Thanks everybody for the responses. She's here.

Image

5174 run hrs. That's about 2 years of daily use.

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by JLROhio » Tue Sep 21, 2021 9:32 pm

Congrats...

Sound like a well-spent $200 for a $3k machine.
Figuring the hours (2 years of use), you got a great deal IMO.

Most here don't seem to care much for machines over 2-3k hours but you were in a unique spot, with a unique (albeit very expensive) situation.

Don't forget to turn on EPR!

You do know how to get into the "Clinician Menu" ??? Hold down that big round button and the Home button together for about 5 seconds to access the Clinician Menu (I'm assuming it's the same for all ResMed 10 machines)
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by JLROhio » Tue Sep 21, 2021 9:38 pm

vilnius wrote:
Tue Sep 21, 2021 9:18 pm
Thanks everybody for the responses. She's here.
5174 run hrs. That's about 2 years of daily use.
To get a Clinician Guide e-mailed to you:

- Send an e-mail to: apneaboard@gmail.com
- in the Subject line, type: "Setup Manual" (without the quotes)
- in the body of the e-mail, just type: ResMed AirCurve 10 ASV

The above is per: https://www.apneaboard.com/adjust-cpap- ... nual#THREE
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by palerider » Tue Sep 21, 2021 10:04 pm

JLROhio wrote:
Tue Sep 21, 2021 9:32 pm
Don't forget to turn on EPR!
ASVs don't have EPR.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by JLROhio » Tue Sep 21, 2021 10:15 pm

palerider wrote:
Tue Sep 21, 2021 10:04 pm
JLROhio wrote:
Tue Sep 21, 2021 9:32 pm
Don't forget to turn on EPR!
ASVs don't have EPR.
Well I can't confirm, as I don't have one, nor a Clinician guide to check...but according to this datasheet from Resmed ----> https://resmedwebinars.com/assets/uploa ... arison.pdf

The Resmed aircurve 10 ASV does have "Easy-Breathe waveform with EPR" ??

Maybe it's built into the machine somehow but I digress...
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

vilnius
Posts: 19
Joined: Mon Sep 20, 2021 5:01 pm

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by vilnius » Tue Sep 21, 2021 10:17 pm

JLROhio wrote:
Tue Sep 21, 2021 9:32 pm
Congrats...

Sound like a well-spent $200 for a $3k machine.
Figuring the hours (2 years of use), you got a great deal IMO.

Most here don't seem to care much for machines over 2-3k hours but you were in a unique spot, with a unique (albeit very expensive) situation.

Don't forget to turn on EPR!

You do know how to get into the "Clinician Menu" ??? Hold down that big round button and the Home button together for about 5 seconds to access the Clinician Menu (I'm assuming it's the same for all ResMed 10 machines)
Thanks! Yep, palerider is correct, this one doesn't have an EPR :( (per: http://www.apneaboard.com/forums/Thread ... #pid243774 and I don't see it in the menus).

I just requested a clinician guide. Thanks for that info.

vilnius
Posts: 19
Joined: Mon Sep 20, 2021 5:01 pm

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by vilnius » Tue Sep 21, 2021 10:20 pm

Question: is ASV ok for me? I don't have a prescription and I didn't get a titration done. I did get the sleep study but never went back (cause it's expensive and my insurance recently expired and I am between jobs).

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by Pugsy » Tue Sep 21, 2021 10:47 pm

PS on the ASV works like EPR meaning it will drop the pressure on exhale but there is no EPR as a separate setting choice on any of the bilevel machines like there is on the AirSense models.

The ASV machine is likely more machine than you need but it can be tamed a bit to make it a bit easier to adjust to.
I used an ASV for over 2 years just because I liked it .....and I have never had a problem with central apneas.

So....we can help you sort out the settings.

While ASV is often thought of as for only central apneas it is an incorrect thought. It can be used very effectively for just Obstructive apnea as well. I did it. Not all that difficult.

If I could find another ASV in my area for 200 bucks I would snatch it up in a heartbeat...even with 5 K hours on it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by palerider » Tue Sep 21, 2021 11:29 pm

JLROhio wrote:
Tue Sep 21, 2021 10:15 pm
palerider wrote:
Tue Sep 21, 2021 10:04 pm
JLROhio wrote:
Tue Sep 21, 2021 9:32 pm
Don't forget to turn on EPR!
ASVs don't have EPR.
Well I can't confirm, as I don't have one, nor a Clinician guide to check...but according to this datasheet from Resmed ----> https://resmedwebinars.com/assets/uploa ... arison.pdf

The Resmed aircurve 10 ASV does have "Easy-Breathe waveform with EPR" ??

Maybe it's built into the machine somehow but I digress...
Well, I DO have one, and I DO have the manual, I also have a AC10 VAuto sitting next to the bed, and neither it, nor the ASV, nor ANY Resmed bilevel device has "EPR", no matter what that comparison put after "Easy-Breathe waveform" because they are already *bilevel* devices.

EPR is basically a limited bilevel, up to 3cm PS, whereas a true bilevel can go much further.

Having EPR on a bilevel machine would just be... useless.

Sending someone on a wild goose chase, trying to turn on a non-existent 'feature' isn't something that should just be ignored.

PS, "EPR" is not anywhere in the ASV, or VAuto clinical manuals.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Should I buy a used Resmed AirCurve 10 ASV for $200?

Post by palerider » Tue Sep 21, 2021 11:32 pm

Pugsy wrote:
Tue Sep 21, 2021 10:47 pm
PS on the ASV works like EPR
Pugsy knows this, I'm just clarifying for the OP.

PS on the ASV works 'like' EPR, only exactly opposite.

EPR is a *drop* from base set pressure, PS is an *increase* from the base pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.