Variable Breathing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Variable Breathing

Post by CorgiWalker » Mon Aug 30, 2021 6:06 pm

So, my AHI’s have returned to an average of 9.32. While investigating, I found the data channel of Variable breathing. On my worst AHI night 16.07, the variable breathing was 54.53% and periodic breathing was 11.2%. I am beginning to suspect that these worse episodes coincide with my AFIB. I see my sleep nurse again in September. She just ups the pressure which increases my leaks, which wakes me. Any suggestions of resources to read or doctors to pursue? Cardiologist nor sleep nurse were very concerned last year, and AHI’s are now worse.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Variable Breathing

Post by Miss Emerita » Mon Aug 30, 2021 8:29 pm

Could you post an Oscar chart so we can take a look?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Re: Variable Breathing

Post by CorgiWalker » Mon Aug 30, 2021 8:37 pm

Sure. I will follow the instructions tomorrow for that.

CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Re: Variable Breathing

Post by CorgiWalker » Tue Aug 31, 2021 9:12 am

This is a picture of a really high night and an average night. I am using an Airtouch F20 with the recalled DreamStation 1. My upper pressure is set at 15. I usually only reach this pressure briefly as in the average night picture. And, I usually wake and press the ramp button to bring it back to my starting pressure of 9.0. I was not able to turn off the pie chart.
Attachments
HighNight.png
HighNight.png (237.32 KiB) Viewed 2126 times
AvgNight.png
AvgNight.png (240.95 KiB) Viewed 2126 times

User avatar
Pugsy
Posts: 65287
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Variable Breathing

Post by Pugsy » Tue Aug 31, 2021 9:20 am

How much time do you spend sleeping on your back?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Re: Variable Breathing

Post by CorgiWalker » Tue Aug 31, 2021 9:48 am

I have no idea. I go to sleep on my side and wake on my side. I try not to sleep on my back. Only once in a blue moon do I wake on my back, and that is usually because it hurts.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Variable Breathing

Post by Miss Emerita » Tue Aug 31, 2021 7:10 pm

Pugsy asks about back-sleeping because you have pronounced clustering of obstructive events. Other culprits can include tucking your chin down toward your chest and REM sleep.

The timing doesn’t look quite right for REM. I suggest you try wearing a soft cervical collar at night, adjusted so that it will keep your chin from tucking down. If that works, great; if it doesn’t, we can contributors think about what’s going on with those clusters.

Is that something you feel you could try?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Re: Variable Breathing

Post by CorgiWalker » Tue Aug 31, 2021 7:23 pm

I will dig out my soft collar and try it. I just might be tucking. I also wondered about my AFIB. But, I can see with the collar.

User avatar
Pugsy
Posts: 65287
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Variable Breathing

Post by Pugsy » Tue Aug 31, 2021 7:27 pm

Any special reason for limiting max to 15?

There are times when you obviously need more pressure but the machine can't go there.

Usual culprits
Supine sleeping
REM stage sleep
sometimes airway position alteration by tucking the chin.

I don't think it is REM....no one gets the amount of REM someone would need to explain away the Aug 27 night.

Quit worrying about variable/PB breathing and worry about the OA clusters/hyponea clusters.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Re: Variable Breathing

Post by CorgiWalker » Tue Sep 21, 2021 12:53 pm

Solution Found: (I think.)
Ya’ll did it! I ordered a vest that doesn’t let you sleep on your back. It arrived 5 days ago. I have used it for 4 nights. My average of 7.2 for 30 days with clusters of events through the night is now 2.0 for 4 nights with a high of 2.3, a low of 1.7, and no clusters. Apparently, I was turning on my back without realizing it. This has got to be better for my AFIB. I thank you whole heartedly! (Yes, pun intended.)
Dreamstation 1, Airtouch F20

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Variable Breathing

Post by Miss Emerita » Tue Sep 21, 2021 2:06 pm

Oh, hey, congratulations on the great experimental problem-solving!
Oscar software is available at https://www.sleepfiles.com/OSCAR/