I Stopped Leak Puffing With Nasal Mask

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

I Stopped Leak Puffing With Nasal Mask

Post by Jerry69 » Thu Aug 19, 2021 12:55 pm

Straw to prevent puffing-small.jpg
Straw to prevent puffing-small.jpg (32.51 KiB) Viewed 3695 times
Slits in end of straw.jpg
Slits in end of straw.jpg (43.17 KiB) Viewed 3695 times
I was troubled by air filling my mouth and puffing between my lips which disturbed my sleep. I fixed it. No more puffing. I cut a soda straw about in half and cut some slits about 1/2" long in one end. I put that end into my mouth between my jaw and rear teeth and let the straw protrude at the corner of my mouth. I had no puffing last night. I experienced none before I went to sleep and was not awakened a single time. (It took a long time for me to go to sleep because I was waiting to see if a puff would occur and I was excited at the possibility of a fix.)

I tried this straw because I reasoned that if the pressure in my mouth was relieved by passage out of my mouth through the straw before it could build up and produce a puff, that the problem would be eliminated. IT WORKS. TRY IT. Refer to it as JERRY'S NO PUFF FIX.

_________________
MachineMask
Last edited by Jerry69 on Thu Aug 19, 2021 5:02 pm, edited 2 times in total.
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

1speechpick
Posts: 130
Joined: Wed Jun 09, 2021 12:16 pm

Re: I Stopped Leak Puffing With Nasal Mask

Post by 1speechpick » Thu Aug 19, 2021 2:07 pm

Jerry69: Do you have a photo of your "straw fix"? I'm having trouble visualizing the fix from the description.
ResMed AirSense 10 For Her with humidifier and heated hose
Resmed Airfit N20


"The best view comes after the hardest climb"

User avatar
ChicagoGranny
Posts: 15532
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Post by ChicagoGranny » Thu Aug 19, 2021 5:08 pm

Straw for Hitting the Hay

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: I Stopped Leak Puffing With Nasal Mask

Post by Jerry69 » Thu Aug 19, 2021 5:09 pm

See my original post. I added pictures.

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

GrumpyHere
Posts: 523
Joined: Sat Feb 13, 2021 2:40 am

Re: I Stopped Leak Puffing With Nasal Mask

Post by GrumpyHere » Thu Aug 19, 2021 11:35 pm

Jerry69 wrote:
Thu Aug 19, 2021 12:55 pm
IT WORKS. TRY IT. Refer to it as JERRY'S NO PUFF FIX.
Jerry's Alvins-Away

Jerry's Chipmunk Pancaker

Jerry's Chipmunk-B-Gone

Jerry's Chipmunk Tunnel

:mrgreen: :mrgreen: :mrgreen:
ResMed Lumis Tx

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: I Stopped Leak Puffing With Nasal Mask

Post by squid13 » Fri Aug 20, 2021 8:27 am

What was your leak rate?

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

User avatar
Pugsy
Posts: 65399
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I Stopped Leak Puffing With Nasal Mask

Post by Pugsy » Fri Aug 20, 2021 8:45 am

squid13 wrote:
Fri Aug 20, 2021 8:27 am
What was your leak rate?
My thoughts as well...
This would be my only concern with using the DIY stent to release the pressure. It's probably not going to be a huge thing but I would want to verify just to make sure.

It is essentially creating a known mouth leak....which may or may not be something worth worrying about.
I know from personal experience that not all mouth leak is massive and in some situations it can be a very minor amount...but I would watch the leak line like a hawk using the straw for a pressure release at least at the beginning for a week or two.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ChicagoGranny
Posts: 15532
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: I Stopped Leak Puffing With Nasal Mask

Post by ChicagoGranny » Fri Aug 20, 2021 9:50 am

This reminds me of the little mouth vent on Somnifix.

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: I Stopped Leak Puffing With Nasal Mask

Post by Jerry69 » Fri Aug 20, 2021 8:40 pm

GrumpyHere wrote:
Thu Aug 19, 2021 11:35 pm
Jerry69 wrote:
Thu Aug 19, 2021 12:55 pm
IT WORKS. TRY IT. Refer to it as JERRY'S NO PUFF FIX.
Jerry's Alvins-Away

Jerry's Chipmunk Pancaker

Jerry's Chipmunk-B-Gone

Jerry's Chipmunk Tunnel

:mrgreen: :mrgreen: :mrgreen:
I like them. I thought about a name for a few. I like yours.

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: I Stopped Leak Puffing With Nasal Mask

Post by Jerry69 » Fri Aug 20, 2021 9:36 pm

ChicagoGranny wrote:
Fri Aug 20, 2021 9:50 am
This reminds me of the little mouth vent on Somnifix.
I went to the website for Somnifix and bought one after reading this: "Yes, you can use PureSleep if you breathe through your mouth during sleep as there is a lot of space between the upper and lower components of the device to allow air to freely pass." That's the point of the straw: to allow air that has already leaked into the mouth to pass out of the mouth before it builds up pressure, expands the cheeks, and finally puffs through the lips.

The Somnifix website explains what causes snoring: "During sleep, the muscles surrounding the upper airway relax, allowing the base of the tongue and surrounding soft tissue to move back, against the rear of the upper airway. Air rushing through this narrowed opening causes the soft tissue to “flutter” like a flag in a breeze, producing the snoring sound.

PureSleep is molded to your teeth, allowing it to firmly but gently hold your lower teeth forward of their normal position, relative to your upper teeth. This moves the jaw forward, along with the base of the tongue and surrounding soft tissue.

By holding your jaw forward, the base of the tongue and surrounding soft tissue are also moved forward, away from the back of the upper airway. This widens the upper airway, allowing improved airflow. Since air no longer must rush through the airway, the soft tissue is far less likely to flutter, so there the sound of snoring is therefore dramatically reduced or entirely eliminated." I think this is also when air leaks into the mouth.

Anyhoo, I'm going to try this Somnifix device to see if it is a better way to allow air to escape the mouth than my straw. Certainly more expensive. Maybe it will prevent air leakage into the mouth. The straw can't do that.

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: I Stopped Leak Puffing With Nasal Mask

Post by Jerry69 » Fri Aug 20, 2021 9:58 pm

Pugsy wrote:
Fri Aug 20, 2021 8:45 am
squid13 wrote:
Fri Aug 20, 2021 8:27 am
What was your leak rate?
My thoughts as well...
This would be my only concern with using the DIY stent to release the pressure. It's probably not going to be a huge thing but I would want to verify just to make sure.

It is essentially creating a known mouth leak....which may or may not be something worth worrying about.
I know from personal experience that not all mouth leak is massive and in some situations it can be a very minor amount...but I would watch the leak line like a hawk using the straw for a pressure release at least at the beginning for a week or two.
The air must escape/leak from the nasal airway before it can be "leaked" out of the mouth, either through a straw or by puffing. The straw doesn't disturb sleep as does puffing. The straw does nothing if their is no leak from the nasal airway, and it doesn't promote this nasal airway leakage.

I checked my leak rate for the two nights I have used the straw, It was very low: 95% leak rate liters/min: Monday, 13; Tuesday, 13; Wed (1st night with straw), 12; Thursday (2nd night with straw), 9. So, the leak rate is not promoted by the straw. I wouldn't expect it to be. The leak has to occur at the base of the tongue before air will enter the mouth. Air flow through the straw or between the lips with puffing only occurs if there is a leak from the airway first.

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
ChicagoGranny
Posts: 15532
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: I Stopped Leak Puffing With Nasal Mask

Post by ChicagoGranny » Sat Aug 21, 2021 5:28 am

Jerry69 wrote:
Fri Aug 20, 2021 9:36 pm
I'm going to try this Somnifix device
Somnifix is not a device. Somnifix is nothing like the PureSleep device.

Did you misgoogle?

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: I Stopped Leak Puffing With Nasal Mask

Post by Jerry69 » Sat Aug 21, 2021 1:34 pm

ChicagoGranny wrote:
Sat Aug 21, 2021 5:28 am
Jerry69 wrote:
Fri Aug 20, 2021 9:36 pm
I'm going to try this Somnifix device
Somnifix is not a device. Somnifix is nothing like the PureSleep device.

Did you misgoogle?
Please give me a link to what you are talking about.

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
Jerry69
Posts: 334
Joined: Fri Nov 11, 2005 7:10 am
Location: The Villages, Florida

Re: I Stopped Leak Puffing With Nasal Mask

Post by Jerry69 » Sat Aug 21, 2021 1:42 pm

I'm sorry to report that the soda straw-fix let me down last night. I puffed a couple of times and found that I was getting too much saliva causing me to have to swallow a lot. Maybe I'll try mouth tape. I hope I don't leak because I'll explode if I leak into my mouth. :D

_________________
MachineMask
Un-treated AHI = 9.5
Titrated prssure: 6 cm
Ave. AHI after therapy = 0.5
Ave. Snore Index = <10
Current pressure = 9 cm

User avatar
zonker
Posts: 11490
Joined: Fri Jun 19, 2015 4:36 pm

Re: I Stopped Leak Puffing With Nasal Mask

Post by zonker » Sat Aug 21, 2021 7:39 pm

Jerry69 wrote:
Sat Aug 21, 2021 1:34 pm
ChicagoGranny wrote:
Sat Aug 21, 2021 5:28 am
Jerry69 wrote:
Fri Aug 20, 2021 9:36 pm
I'm going to try this Somnifix device
Somnifix is not a device. Somnifix is nothing like the PureSleep device.

Did you misgoogle?
Please give me a link to what you are talking about.
lmgtfy-
https://somnifix.com/products/mouth-str ... sleep-aids
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg