Wow !! 0 Leaks several months

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Architect
Posts: 168
Joined: Sun Jan 31, 2021 5:09 pm
Location: Central Tx

Wow !! 0 Leaks several months

Post by Architect » Thu Aug 05, 2021 9:59 pm

Hi friends, desired to share last year I switched over to Resmed Airtouch F20 from Qauttro, many issues at first, headgear, leaks, etc... Once I figured out how mask fits on my face properly, I ordered Silent night mask liner and neck padding.

Past several months 0 leaks !!!!!!!!!!!! No Ahi over 1 !! I download data from Resmed Myair past 3 months, don't know how to attach image. I'm still in disbelief.

My little secret, suppose to change liners every week, I change once a month :shock: The trick is, wash or wipe face down, before using mask liner.

Sleep wonderful friends............

clownbell
Posts: 369
Joined: Fri Apr 02, 2021 8:46 pm

Re: Wow !! 0 Leaks several months

Post by clownbell » Thu Aug 05, 2021 10:26 pm

To OP - I use a pillow mask (Brevida) and most nights I get VERY few leaks. Less than 1% of the time over redline. However there are one or two nights per month when large leaks are a problem. Although I normally do not wash my face before bed, do you think washing my nose would help? That may sound like a crazy question but I am serious. Wat do you think?
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask

Architect
Posts: 168
Joined: Sun Jan 31, 2021 5:09 pm
Location: Central Tx

Re: Wow !! 0 Leaks several months

Post by Architect » Thu Aug 05, 2021 11:21 pm

clownbell wrote: ↑
Thu Aug 05, 2021 10:26 pm
To OP - I use a pillow mask (Brevida) and most nights I get VERY few leaks. Less than 1% of the time over redline. However there are one or two nights per month when large leaks are a problem. Although I normally do not wash my face before bed, do you think washing my nose would help? That may sound like a crazy question but I am serious. Wat do you think?
Hi friend, fantastic leak results. TBH I wouldn't know, cause I never used nasal pillows, sorry. Hopefully, someone can chime in that use nasal pillows.

But looking at Brevida mask, and if your face get oily, I would clean around the nose area with wipes, see if that helps. I use unscented, no alcohol baby wipes.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Wow !! 0 Leaks several months

Post by chunkyfrog » Fri Aug 06, 2021 7:30 am

Architect wrote: ↑
Thu Aug 05, 2021 11:21 pm
. . .
unscented, no alcohol baby wipes.
Baby wipes should not EVER contain alcohol.
That stuff STINGS!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Wow !! 0 Leaks several months

Post by palerider » Fri Aug 06, 2021 9:59 am

chunkyfrog wrote: ↑
Fri Aug 06, 2021 7:30 am
Architect wrote: ↑
Thu Aug 05, 2021 11:21 pm
. . .
unscented, no alcohol baby wipes.
Baby wipes should not EVER contain alcohol.
That stuff STINGS!
No, they shouldn't. However, using alcohol on the silicone mask to clean it *IS* a *GOOD* idea, because the alcohol won't hurt the silicone, but it's very effective at removing facial oils.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11501
Joined: Fri Jun 19, 2015 4:36 pm

Re: Wow !! 0 Leaks several months

Post by zonker » Fri Aug 06, 2021 11:01 am

Architect wrote: ↑
Thu Aug 05, 2021 9:59 pm


Past several months 0 leaks !!!!!!!!!!!! No Ahi over 1 !! I download data from Resmed Myair past 3 months, don't know how to attach image. I'm still in disbelief.
feels good, don't it? :D

now my obligatory disclaimer to the newbies-

we don't "chase" zero, but it still feels good when we get it.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Wow !! 0 Leaks several months

Post by Janknitz » Fri Aug 06, 2021 6:11 pm

I'm always suspicious of several 0 leaks in a row . . .
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
amarnina
Posts: 4
Joined: Mon Nov 16, 2020 11:28 am

Re: Wow !! 0 Leaks several months

Post by amarnina » Sat Nov 06, 2021 3:53 pm

For those who have used Silent Night liners and write of 0 leaks - do you tighten your FFM tighter? I used my first SIlent Night liner last night and loved the feel of it going on but upon awakening was presented with an orange/red sad face for SEAL on my Resmed display.

Any tips? What am I missing here? I had always wondered how a porous cloth between my skin and the silicon cushion could actually improve on leaks.

clownbell
Posts: 369
Joined: Fri Apr 02, 2021 8:46 pm

Re: Wow !! 0 Leaks several months

Post by clownbell » Sun Nov 07, 2021 2:08 pm

Janknitz wrote: ↑
Fri Aug 06, 2021 6:11 pm
I'm always suspicious of several 0 leaks in a row . . .
If I understand correctly, OP is using ResMed My Air app, which only reports leaks over redline, so OP could have 23 lpm of leakage and My Air would say 0.
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Wow !! 0 Leaks several months

Post by Pugsy » Sun Nov 07, 2021 3:00 pm

clownbell wrote: ↑
Sun Nov 07, 2021 2:08 pm
using ResMed My Air app, which only reports leaks over redline, so OP could have 23 lpm of leakage and My Air would say 0.
Don't think so. According to this discussion there is also a L/min data point on the MyAir app report.
Maybe it's in a different area than the "score" numbers.
http://www.apneaboard.com/forums/Thread ... ew-formula
This guy is seeing L/min and mentions "usually in the teens" which is of course below 24L/min
Discovered the new 30 day MyAir yesterday. I have written recorded data from MyAir for part of November 2017. Vast majority of leaks were under 24L-Usually in the teens. Now the new 30 day listing gives markedly increased values for those exact same past dates- 30's to 50's! All the other AHI, duration, mask on off are correct. I have informed my sleep physician and am waiting for an answer. Last night I had a "smiley" face on my Resmed LCD screen meaning the leak was under 24L, but the MyAir online report says 25L. The leak function is just erroneous and I hope it doesn't result in untoward effects or changes in patient's therapy/compliance status.
Can someone using MyAir confirm or not? I don't use it myself.

The S9 models were well known to get stuck reporting a certain number for prolonged periods of time.
I hadn't heard of the AirSense/AirCurve models getting stuck but since the fix is easy....just unplug the machine and do a forced reset of the software....that's what I would do.
I would also verify with OSCAR. Anytime I see any sort of weird looking and not "normal" data point I verify with some other software.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Wow !! 0 Leaks several months

Post by Dog Slobber » Sun Nov 07, 2021 4:07 pm

Pugsy wrote: ↑
Sun Nov 07, 2021 3:00 pm
Can someone using MyAir confirm or not? I don't use it myself.
I haven't seen any changes in MyAir, nor noticed any "markedly increased values".

My MyAir leaks aligns well with The OSCAR historical statistics. Having said that OSCAR reported leaks never align exactly with MyAir's. BlueDragon has occasionally commented on occasional differences between OSCAR and ResMed/MyAir, often attributing it to rounding differences, sometimes exasperated by cumulation and multiple sessions. MyAir also only reports in whole numbers.

I've never been concerned about the slight differences between OSCAR and MyAir, but no, there is not a significant difference.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Wow !! 0 Leaks several months

Post by Pugsy » Sun Nov 07, 2021 4:14 pm

Dog Slobber wrote: ↑
Sun Nov 07, 2021 4:07 pm
Pugsy wrote: ↑
Sun Nov 07, 2021 3:00 pm
Can someone using MyAir confirm or not? I don't use it myself.
I haven't seen any changes in MyAir, nor noticed any "markedly increased values".

My MyAir leaks aligns well with The OSCAR historical statistics. Having said that OSCAR reported leaks never align exactly with MyAir's. BlueDragon has occasionally commented on occasional differences between OSCAR and ResMed/MyAir, often attributing it to rounding differences, sometimes exasperated by cumulation and multiple sessions. MyAir also only reports in whole numbers.

I've never been concerned about the slight differences between OSCAR and MyAir, but no, there is not a significant difference.
Does MyAir report leaks in L/min??? Someone said that MyAir only reported leaks when over 24 L/min.
That's what I was wondering. I thought MyAir reported the 95% leak number like the machine does on the LCD screen sleep report.

Now I could see a 0 deduction on the "score" that it gives if someone didn't go into large leak territory...but a 0 deduction isn't the same as 0 leak.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Wow !! 0 Leaks several months

Post by Dog Slobber » Sun Nov 07, 2021 4:41 pm

Pugsy wrote: ↑
Sun Nov 07, 2021 4:14 pm
Does MyAir report leaks in L/min??? Someone said that MyAir only reported leaks when over 24 L/min.
Yes.

MyAir_Leaks.jpg
MyAir_Leaks.jpg (132.52 KiB) Viewed 1822 times
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Wow !! 0 Leaks several months

Post by Pugsy » Sun Nov 07, 2021 5:03 pm

Dog Slobber wrote: ↑
Sun Nov 07, 2021 4:41 pm
Yes.
Thank you.

Since I am using a "used" machine I can't setup up MyAir even if I wanted to (which I don't) as it was once registered to someone else and the hurdles, I would need to jump just to get something I don't want anyway, are more than I want to jump.
I thought it reported the 95% number or maybe that's the max??? Anyway it does report more than 0.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.